If a new IFUs version is released, how is it ensured, in a realistic way, that the surgeon really takes notice of the changes in the IFUs (bearing in mind that only the device, without the IFUs, is presented to the surgeon during the surgical procedure)?
 
Elsmar Forum Sponsor

yodon

Staff member
Super Moderator
#2
Other than training, it's probably not possible. You could try to get confirmation but that may be difficult.

I think this is one of the reasons why, in the harmonized 14971, you cannot reduce risk through labeling alone.
 
M

MIREGMGR

#3
Is the nature of your device that you do in-services or other in-person training?

If the change is patient-risk- or outcome-significant, don't let anyone use it for whom you don't have evidence of completed training.
 

Edugar

Involved In Discussions
#4
Thanks for your comments.

The device is an implant, widely used in a specific surgical speciality, and it does not need special training. When the product is introduced in the hospital, the sales rep. explains specific details. Afterwards, the experienced surgeons transmit their experience to the new surgeons.

Almost always the changes of the IFUs do not represent patient risk issues. But obviously it is interesting that IFUs changes reach the surgeons; and the habitude to check periodically the IFUs is not generally extended.
 

pkost

Trusted Information Resource
#5
If the change is due to a risk to the patient, you should consider issueing a field safety corrective action - FSCA do not have to just be recalls, they can be for informational purposes as well i.e. IFU updates. Certainly within the UK the MHRA distributes to hospitals and the hospital should communicate to the relevent parties.

Additionally, as advised in previous posts, most implantables I've worked with have sales reps attend theatre and advise the surgeons on appropriate techniques...The sales reps will know all the surgeons using your products, should have a good relationship with them and should be in a position to communicate changes to them.
 
Thread starter Similar threads Forum Replies Date
N Is Notified Body approval necessary for small changes to IVD IFU? CE Marking (Conformité Européene) / CB Scheme 7
E Changes to IFU (Instructions for Use) on a CE Marked Medical Device ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
K No IFU, where the requirements would go? EU Medical Device Regulations 5
P IFU Translation Requirements for EU EU Medical Device Regulations 2
V Preparing the IFU in compliance with MDR 745, Chapter III Elsmar Xenforo Forum Software Instructions and Help 2
E Commission Regulation (EU) No. 207/2012 E-IFU Applicability EU Medical Device Regulations 3
K REF symbol with product code in the IFU EU Medical Device Regulations 1
F Control of medical device IFU content Other Medical Device and Orthopedic Related Topics 3
B Does EN ISO 15223-1:2016 include the graphic symbols to be added to software and IFU? Other Medical Device Related Standards 9
K REF code and Consult IFU in Shipping carton boxes. EU Medical Device Regulations 1
S Providing IFU (Instructions for Use) in multiple languages EU Medical Device Regulations 0
S IFU on web page - Consequences if the most recent one is not uploaded EU Medical Device Regulations 3
I 2017/745 (MDR) - Instruction for use (IFU) - manufacturer's website EU Medical Device Regulations 11
H IFU in different language other than English EU Medical Device Regulations 5
QIE Electronic IFU - new project, need input Other Medical Device and Orthopedic Related Topics 0
Z IFU (Instructions for Use) Symbol - 7010:M002 - can it be in black? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Include IFU in Class II Medical Device Marketing Brochure? US Food and Drug Administration (FDA) 2
M EU Regulations or Standards for instruction for use (IFU ) EU Medical Device Regulations 3
P IVDD - IFU related requirements - Type of information/statement expected EU Medical Device Regulations 1
K Label - IFU requirement, maybe? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
W Electrosurgical devices/surgery IFU contraindications EU Medical Device Regulations 2
N Updating class I device IFU with instructions to use class II cleared device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Q Product Instruction for Use Manual (IFU) - Obligation to list the standard used 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Ajit Basrur Incoming Inspection for Instructions for Use (IFU) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
T Product Instructions (IFU) - Placement Requirements - ISO 13485:2003/2016 Other ISO and International Standards and European Regulations 2
D 207/2012 Update for MDR (?) - Instructions for Use (IFU) EU Medical Device Regulations 2
B IFU and deviation 7 in ISO 14971 Annex ZA ISO 14971 - Medical Device Risk Management 1
F EMC (Electromagnetic Compatibility) Conformity Charts - IFU Requirements Misc. Quality Assurance and Business Systems Related Topics 3
P IFU for Sterling Silver Surgical devices ISO 13485:2016 - Medical Device Quality Management Systems 3
O Regulatory Notification for IVD IFU Instructions Revision ISO 13485:2016 - Medical Device Quality Management Systems 3
E Does posting your IFU on your website constitute elabelling? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
M Laminating abbreviated Operators Manual to IFU's and DFU's ISO 13485:2016 - Medical Device Quality Management Systems 1
B Brazil ANVISA Medical Device Labeling/IFU Requirements Other Medical Device Regulations World-Wide 1
J Canada Bulk Packaging and IFU requirements for Class IIa and IIb Medical Devices Canada Medical Device Regulations 2
J Medical Device Procedure Pack IFU's (Instructions for Use) ISO 13485:2016 - Medical Device Quality Management Systems 5
R How and Where can I find IFU (Instructions For Use) examples? ISO 13485:2016 - Medical Device Quality Management Systems 6
M Orthopedic IFU ( Instruction For Use ) EU Medical Device Regulations 3
J Domestic and International Medical Device IFU (Instructions for Use) Other US Medical Device Regulations 3
G Medical Device IFU (Instructions for Use) Translation Requirements EU Medical Device Regulations 4
sagai Electronic Instructions for Use (IFU) of Medical Devices EU Medical Device Regulations 10
R Class II Medical Device - Marketing vs. Indication For Use (IFU) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
T Providing Instructions for Use (IFU) vs. 'Help' pages 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
B Brazil requirements to have translations of IFU to Portuguese certified Other Medical Device Regulations World-Wide 16
C IFU Language Requirements for FDA Clinical Study conducted in EU EU Medical Device Regulations 3
P EU Electronic IFU (Instructions for Use) Requirements EU Medical Device Regulations 2
B Language Requirements - Labels/IFU (Professional Use Medical Devices) in Bangladesh Other Medical Device Regulations World-Wide 2
S IFU for combined OTC Medical Device in light of IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
D Electronic IFU's on Class III and IV Medical Devices World Wide Other Medical Device Regulations World-Wide 1
L Device Description in CSDT Format - Medical Device Instructions for Use (IFU) Other Medical Device Regulations World-Wide 6
M Manufacturing Address - Section 5.3 of EN 591 - In Vitro Diagnostic IFU EU Medical Device Regulations 4
Similar threads


















































Top Bottom