Let's say I use the symbol for irradiation sterilization on the label of a sterile implant. Do I have to print the symbol in the instrucitons for use?

My current thinking is that 13.6(a) stipulates that all information (with a few exceptions) that is printed on the label also needs to be included in the IFU. This would include the symbols, right?

Do you have any input on this?

/K
 
Elsmar Forum Sponsor

Michael Malis

Quite Involved in Discussions
#2
Re: Label symbols also in IFU?

Hi all,

Let's say I use the symbol for irradiation sterilization on the label of a sterile implant. Do I have to print the symbol in the instrucitons for use?

My current thinking is that 13.6(a) stipulates that all information (with a few exceptions) that is printed on the label also needs to be included in the IFU. This would include the symbols, right?

Do you have any input on this?

/K
The short answer is yes!

You need to explain your symbols in the IFU. Also some areas of the world (such as Canada) do not recognize or accept the use of symbols. In order to make use of symbols, a legend defining these symbols must accompany the product.

Hope this helps,
Mike
 
K

karha096

#3
Thanks, but my question is not whether you need to explain the use of symbols you invent yourself, but rather if the symbols standardized by 980 aslo need to be printed in the IFU.
 

Michael Malis

Quite Involved in Discussions
#4
Thanks, but my question is not whether you need to explain the use of symbols you invent yourself, but rather if the symbols standardized by 980 aslo need to be printed in the IFU.
My message above was referring to allsymbols. Regardless of ISO 980, ISO 15523 or any others, - all symbols must be explained in the IFU.
This action is allowing regions and countries that are not familiar with EU regulations to understand it.
 
M

MIREGMGR

#5
The labeling for a device that is marketed/used only within the EC only needs to explain those symbols that are not standardized in EN980 or ISO15223.

The labeling for a non-OTC device that is marketed/used in the US, if it includes symbols other than the limited set that are listed in the FDA guidance "Use of Symbols on Labels and in Labeling of In Vitro Diagnostic Devices Intended for Professional Use" (2004 edition), must include English text that explains the symbols. Some FDA examiners expect the use of English text with all symbols. The FDA does not recognize EN980 and ISO15223.

The labeling for a non-OTC device that is marketed/used in Canada must state and identify certain information in either English or French text. I'm under the impression that the use of symbols for other information, or symbols in addition to the required text in the case of specifically required information, is allowed.

Requirements in other countries are dependent on their medical device regulatory systems. Many countries allow importation of devices that are labeled to conform with FDA or EC requirements, for instance, and those regulatory systems do not require the inclusion of a separate IFU in certain circumstances.
 
Last edited by a moderator:

Michael Malis

Quite Involved in Discussions
#6
...The labeling for a non-OTC device that is marketed/used in Canada must state and identify certain information in either English or French text. I'm under the impression that the use of symbols for other information, or symbols in addition to the required text in the case of specifically required information, is allowed....
:nope:CANADA does not recognize or accept the use of symbols without package insert or leaflet defining those symbols! Also only Quebec province, require French language.
 
Last edited:
M

MIREGMGR

#7
:nope:CANADA does not recognize or accept the use of symbols without package insert or leaflet defining those symbols! Also only Quebec province, require French language.
Could you provide a link or reference to that restriction in HC's "Guidance for the Labelling of Medical Devices (...)" or the "Medical Device Regulations"?

As to languages...Section 23 of the "Medical Device Regulations" and Section 1.3 of "Guidance for the Labelling of Medical Devices (...)" both state that information required to be provided on a device label (except for manufacturer name/address and the product's identifying and batch-code numbers) shall be in English and French if the device is to be sold to the general public, and in either English or French or both if the device is for professional use only, with the device's IFU available to be provided in the other official language upon request. There is no mention in either document of Quebec specificity.
 
C

Cinti

#8
Hi,

I need one information. EN 980:2008 contains 29 symbols. can anybody list out the symbols that should be print out in labels, carton box and IFU.
 
M

MIREGMGR

#9
EN 980:2008 contains 29 symbols. can anybody list out the symbols that should be print out in labels, carton box and IFU.
Each of the standardized symbols communicates the proper action or understanding in regard to some circumstance. You determine which of those circumstances apply to your product, then use the corresponding symbols.

In most instances, it's appropriate for the same information to be communicated on both the box/case/carton label and the individual-product label. If the individual-product label carries the full range of regulatory and safety/effectiveness information to be communicated, the IFU may be focused only on its specific task of instructions for use. There is however no harm in repeating on the IFU the information from the labels, as long as it is correct and consistent.
 

sreenu927

Quite Involved in Discussions
#10
Hi Cinti,

All it depends on your type of medical device. What is your device?

As a minimum, every medical device requires, SN or LOT, REF, Manufacturing date, Use by date, Consult Instructions for Use, Manufacturer, if requires sterilization, then type of sterilization symobl, if for single use, then single use only symbol, if IVD, IVD symbol, etc...

Regards
Sreenu
 
Thread starter Similar threads Forum Replies Date
Q Instructions for Use (IFU) - On the Manufacturer's Website vs. On the Product Label US Food and Drug Administration (FDA) 15
U Intended Users in IFU (Instructions for Use) EU Medical Device Regulations 5
S Providing IFU (Instructions for Use) in multiple languages EU Medical Device Regulations 0
Z IFU (Instructions for Use) Symbol - 7010:M002 - can it be in black? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
N Updating class I device IFU with instructions to use class II cleared device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Ajit Basrur Incoming Inspection for Instructions for Use (IFU) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
T Product Instructions (IFU) - Placement Requirements - ISO 13485:2003/2016 Other ISO and International Standards and European Regulations 2
D 207/2012 Update for MDR (?) - Instructions for Use (IFU) EU Medical Device Regulations 9
O Regulatory Notification for IVD IFU Instructions Revision ISO 13485:2016 - Medical Device Quality Management Systems 3
J Medical Device Procedure Pack IFU's (Instructions for Use) ISO 13485:2016 - Medical Device Quality Management Systems 5
R How and Where can I find IFU (Instructions For Use) examples? ISO 13485:2016 - Medical Device Quality Management Systems 6
J Domestic and International Medical Device IFU (Instructions for Use) Other US Medical Device Regulations 3
G Medical Device IFU (Instructions for Use) Translation Requirements EU Medical Device Regulations 4
E Changes to IFU (Instructions for Use) on a CE Marked Medical Device ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
sagai Electronic Instructions for Use (IFU) of Medical Devices EU Medical Device Regulations 10
T Providing Instructions for Use (IFU) vs. 'Help' pages 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
P EU Electronic IFU (Instructions for Use) Requirements EU Medical Device Regulations 2
L Device Description in CSDT Format - Medical Device Instructions for Use (IFU) Other Medical Device Regulations World-Wide 6
C Authorized Representative Contact Information in IFU (Instructions for Use) EU Medical Device Regulations 2
S Can we put two Manufacturer's address in IFU (Instructions For Use)? CE Marking (Conformité Européene) / CB Scheme 2
O Harmonized Standards on Medical Device IFU (Instructions for Use) ISO 13485:2016 - Medical Device Quality Management Systems 7
G IFU (Instructions for Use) - Labeling Requirements EU Medical Device Regulations 1
K Different Indications for Use in Different Countries -- how to handle IFU? Other Medical Device Regulations World-Wide 4
S Shelf Life Test for IFU and Implant Card Other Medical Device Related Standards 1
M MDR - eIFU - When you can supply an electronic IFU instead of a paper IFU EU Medical Device Regulations 5
J Warnings/Cautions in Medical Device IFU Medical Device and FDA Regulations and Standards News 4
Ed Panek IFU Contact Requirements - FDA and MDD/MDR US Food and Drug Administration (FDA) 1
C Does an accessory need an IFU if it use is discussed in the Parent device IFU? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
R MDR, Annex I, 23.1 Interpretation - IFU on the website EU Medical Device Regulations 5
J EU 207/2012 E-IFU Service Video EU Medical Device Regulations 1
A Which guidance do you use for guiding the requirements of IFU for 510(k) submission? Medical Device and FDA Regulations and Standards News 5
B Do IFU designs have to be document controlled under ISO 13485? Document Control Systems, Procedures, Forms and Templates 2
K No IFU, where the requirements would go? EU Medical Device Regulations 5
P IFU Translation Requirements for EU EU Medical Device Regulations 2
V Preparing the IFU in compliance with MDR 745, Chapter III EU Medical Device Regulations 2
E Commission Regulation (EU) No. 207/2012 E-IFU Applicability EU Medical Device Regulations 5
K REF symbol with product code in the IFU EU Medical Device Regulations 1
F Control of medical device IFU content Other Medical Device and Orthopedic Related Topics 3
B Does EN ISO 15223-1:2016 include the graphic symbols to be added to software and IFU? Other Medical Device Related Standards 9
K REF code and Consult IFU in Shipping carton boxes. EU Medical Device Regulations 1
S IFU on web page - Consequences if the most recent one is not uploaded EU Medical Device Regulations 3
I 2017/745 (MDR) - Instruction for use (IFU) - manufacturer's website EU Medical Device Regulations 15
H IFU in different language other than English EU Medical Device Regulations 5
QIE Electronic IFU - new project, need input Other Medical Device and Orthopedic Related Topics 0
M Include IFU in Class II Medical Device Marketing Brochure? US Food and Drug Administration (FDA) 2
M EU Regulations or Standards for instruction for use (IFU ) EU Medical Device Regulations 3
P IVDD - IFU related requirements - Type of information/statement expected EU Medical Device Regulations 1
K Label - IFU requirement, maybe? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
W Electrosurgical devices/surgery IFU contraindications EU Medical Device Regulations 2
Q Product Instruction for Use Manual (IFU) - Obligation to list the standard used 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1

Similar threads

Top Bottom