Implications of failing an Anvisa Audit for ISO 13485 Certification

kys123

Starting to get Involved
Hello everyone,

I'd like to know more about the potential implications of nonconformances found during a third-party audit by Anvisa (the Brazilian regulatory agency as you may or not know), for a European manufacturer that is ISO 13485 certified and was looking to export their products to Brazil.

While I do not have the specific details about the nonconformances found, I am curious about how failing an international audit might impact the manufacturer's ISO 13485 certification.

Are there any potential consequences that could arise from the nonconformances (depending on how serious they were), and if so, what would they be?
 

yodon

Leader
Super Moderator
I was hoping someone would jump on this but since nobody has yet and it's aging a bit, let me toss out what I *think* (I don't know for sure). Check your agreement with your CB.

I was thinking that agreements with certification bodies require notification of findings from other audits. I may well be wrong.

I would not look at it as an audit "failure" (I presume they didn't require you to pull your device from the market?) but, as you later not, identification of nonconformities. I presume you're going to take action to resolve them and that will only strengthen your system / product. I would be forthcoming with the CB and show them the steps taken. It just shows you're properly operating your QMS.

I wouldn't volunteer it in an audit but, as noted above, I think it may be required to be reported to them.
 

Sidney Vianna

Post Responsibly
Leader
Admin
While I do not have the specific details about the nonconformances found,
That is a critical piece of information. Maybe the nonconformances were specific against Brazilian regulations only.

Regulatory noncompliance is no reason for yanking anyone’s management system certification, as long as the organization reacts appropriately to correct the situation, what could even be departure from a market with an unreasonable regulatory exposure and/or corrupt inspectors from a regulatory body.
 

kys123

Starting to get Involved
I was hoping someone would jump on this but since nobody has yet and it's aging a bit, let me toss out what I *think* (I don't know for sure). Check your agreement with your CB.

I was thinking that agreements with certification bodies require notification of findings from other audits. I may well be wrong.

I would not look at it as an audit "failure" (I presume they didn't require you to pull your device from the market?) but, as you later not, identification of nonconformities. I presume you're going to take action to resolve them and that will only strengthen your system / product. I would be forthcoming with the CB and show them the steps taken. It just shows you're properly operating your QMS.

I wouldn't volunteer it in an audit but, as noted above, I think it may be required to be reported to them.

Hello yodon, thanks for jumping on in the discussion.

Sorry for not making it very clear, I am currently employed at a consulting company in Brazil who was interested in the manufacturer's products.
The stakeholders are looking at it as a "failure" because they failed to meet the Anvisa GMP standard, which to be fair is very similar to ISO 13485 (it does have some specific points as Sidney pointed out), and will probably not even have a chance of receiving the brazilian GMP certificate, having wasted quite some time and a hefty fee.

The word around is that they had quite a few major nonconformances, which is why I was curious about their certification due to IMDRF rules/CBs/etc., however, I do not know what exactly happened but I might find out soon and share if possible.
 
Last edited:

kys123

Starting to get Involved
I was hoping someone would jump on this but since nobody has yet and it's aging a bit, let me toss out what I *think* (I don't know for sure). Check your agreement with your CB.

I was thinking that agreements with certification bodies require notification of findings from other audits. I may well be wrong.

I would not look at it as an audit "failure" (I presume they didn't require you to pull your device from the market?) but, as you later not, identification of nonconformities. I presume you're going to take action to resolve them and that will only strengthen your system / product. I would be forthcoming with the CB and show them the steps taken. It just shows you're properly operating your QMS.

I wouldn't volunteer it in an audit but, as noted above, I think it may be required to be reported to them.

That is a critical piece of information. Maybe the nonconformances were specific against Brazilian regulations only.

Regulatory noncompliance is no reason for yanking anyone’s management system certification, as long as the organization reacts appropriately to correct the situation, what could even be departure from a market with an unreasonable regulatory exposure and/or corrupt inspectors from a regulatory body.

Hey guys, hope you are doing well. I finally got some more information regarding what happened:

The audit uncovered almost 30 nonconformities, with around 5 of them being considered major.

While I did not get to read the entire report yet, the two most concerning nonconformities were that the company was (is?) manufacturing injectable products without effective measures for water quality control (their water system was not validated and only being monitored at a single tank) and microbiological control (only tested in the sterilization tank, no control whatsoever over the equipment used for filling syringes).
What can you do... :magic:
 
Thread starter Similar threads Forum Replies Date
K Change in address on ERLM US FDA: Implications to import US Medical Device Regulations 0
C Labelling and UDI Requirements - Implications of serial numbers US Medical Device Regulations 1
P Transferring medical data from a device (Sec 201(h)): regulatory implications US Medical Device Regulations 3
S Device with different options for a sub-component - CE marking implications EU Medical Device Regulations 2
M Changing Investigational Device from Non-Sterile to Sterile - what are the implications? Other Medical Device and Orthopedic Related Topics 3
H Expanding our business and the implications to our QMS Manufacturing and Related Processes 2
JoCam False information provided for Medical Device Registration - What are the implications? Other Medical Device Related Standards 3
B Implications of Dropping ISO 9001 Certification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
H Question about implications of performing Firmware upgrade via MDDS - Medical Device Data Systems 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
M Informational TGA – European implementation of Medical Device and IVD Regulations – Implications for Australia Medical Device and FDA Regulations and Standards News 0
M Automatic Data Gathering Requirements and Privacy Implications Medical Information Technology, Medical Software and Health Informatics 0
B Manufacturing Company Address Change - implications for FDA Other US Medical Device Regulations 1
M Informational TGA – Brexit – Implications for therapeutic goods in Australia Medical Device and FDA Regulations and Standards News 0
M "Minor" Address Change Implications on Product Labeling Requirements Misc. Quality Assurance and Business Systems Related Topics 2
M New Canadian Safety Standard? SPE 3000 - What are the implications? Canada Medical Device Regulations 1
M Implications of an import detention Other US Medical Device Regulations 3
S Change of Medical Device Company Registered Address - Implications Quality Manager and Management Related Issues 3
M CE Mark outside of Europe - Implications EU Medical Device Regulations 4
M What the FDA considers a Prescription Device - Decisions and Implications Other US Medical Device Regulations 18
S What are the implications of a major NC during a TS 16949 surveillance audit? IATF 16949 - Automotive Quality Systems Standard 3
K Business Impact Analysis Questionnaire - Financial Implications Business Continuity & Resiliency Planning (BCRP) 1
K Implications of References in Documentation Document Control Systems, Procedures, Forms and Templates 4
Hershal ILAC P14 and its implications General Measurement Device and Calibration Topics 0
B Implications for CE Marking due to Device Modifications by the Purchaser CE Marking (Conformité Européene) / CB Scheme 1
bobdoering Myth or Mythunderstanding: Implications of the Economic Design of Control Charts Statistical Analysis Tools, Techniques and SPC 1
T Implications on the Legal Status of the "Declaration of Conformity" EU Medical Device Regulations 8
M TS 16949 and its HR Implications - Job Descriptions? Competencies? IATF 16949 - Automotive Quality Systems Standard 2
M QS9000 element 4.1.6.1 implications QS-9000 - American Automotive Manufacturers Standard 4
K ISO 9001 2000 - Implications for a small company ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
J The Implications of ISO9001 - Quality Improvements vs. Customer Pressure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
I Failing concentricity Manufacturing and Related Processes 3
Watchcat Are De novos failing? Other US Medical Device Regulations 4
Sidney Vianna Interesting Discussion How major corporations in the West keep failing to uphold their pledges of abolishing child labor in their cocoa supply chains Sustainability, Green Initiatives and Ecology 3
M Medical Device News Olympus Medical pleads guilty to failing to file events notice: Justice Department Medical Device and FDA Regulations and Standards News 0
S ISO13485 2016 - What the upshot is of failing to upgrade on a medical diagnostic device ISO 13485:2016 - Medical Device Quality Management Systems 1
M Arguing with the test lab about failing in an EMC test IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
G Options for failing a VDA POTENTIAL audit VDA Standards - Germany's Automotive Standards 2
N How would you approach failing quality objective measurables? Quality Tools, Improvement and Analysis 12
AnaMariaVR2 FDA Global Clampdown: 8 Warning Letters - Failing to Register Plants US Food and Drug Administration (FDA) 0
B Gage R&R Newbie - Everything is Failing! Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 14
F What's the next step after failing an AQL? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
H Hydrophilic Coating of Catheter Shafts Failing Durability Testing ISO 13485:2016 - Medical Device Quality Management Systems 2
QMMike Why is this gage R&R failing? Quality Tools, Improvement and Analysis 10
Jim Wynne First Year Teacher - Why US public schools are failing Coffee Break and Water Cooler Discussions 87
B Analysis of failing parts - Mining equipment distributor Reliability Analysis - Predictions, Testing and Standards 3
P A Quick look at a failing business Career and Occupation Discussions 6
Marc Are ISO 9000 Registrations failing to grow? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
Y Gap assessment comparing RDC ANVISA 665/2022 and RDC ANVISA 16/2013? Other Medical Device Related Standards 0
P Brazil - ANVISA - distance selling Other Medical Device Regulations World-Wide 0
L ANVISA’s RDC 665 replaces previous regulations including RDC 16/2013 and IN 08/2013, and comes into effect on May 2, 2022 Other Medical Device Regulations World-Wide 2

Similar threads

Top Bottom