Impossible to achieve NDC (Number of Distinct Categories) greater than 2

J

Jirka

#1
Hello everybody,

I have read many of your posts, and no one has solved my problems, so, instead of giving to you my example and make this thread so long such as the others threads which I have read, I had the idea to ask oppositte:

Does anybody can attach me please a R&R study excel sheet using the following data which we usally use in our factory?:

1- Dimension to be measured: Across flats 17.73 - 18.00 mm
2- Tooling used for measuring: Milessimal Micrometer (0.001mm)
3- The studies used are 10 parts x 2 appraisers x 2 trials
4- R&R study has to be performed considering the % of the tolerance (GRR)

The idea that I have is if somebody can fullfill me that sheet with random dimensions in order to get the NDC more greater than 2 and of course with the GRR% lower than 30%.

I can subbmit you my excel file, but I'm afraid that maybe has some mistake, so I prefer if somebody can attach me your excel file and then I would check with my data to find where is the mistake (if exist) because I have tried to create "the ideal" R&R study and it's impossible to achieve a GRR% lower than 30% and at the same time a NDC greater than 2.

We need to give an answer to our auditor next month of why our R&R studies doesn't get the ndc's... :eek:
Thank you for your kind attention! :thanx:
 
Elsmar Forum Sponsor
B

Barbara B

#2
Re: Impossible to achieve NDC greater than 2

Jirka,

could you please give us at least the data so we could see for ourselves why there's a problem with your metrics?

%GRR and ndc could be converted into each other. To get a ndc>=5, %GRR has te be smaller than 23%. For %GRR=30% the ndc is calculated as ndc=3.3 (see AIAG MSA 4th, p.122-123).

Regards,

Barbara
 

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#3
Re: Impossible to achieve NDC greater than 2

Your device should be sufficient for measuring the characteristic you are looking at....but there are some things that crop up the way the calculations work.

ndc is calculated based on PV, or (more importantly) the variation you presented to the device as a part of the study. If you have little variation and yet it represents the variation over time you expect, then ndc expects even more resolution or less gage error to achieve a reasonable value. Even 5 is a weak number if you wish to use SPC. If the data has less variation that what you expect to see over time, then you need to address that. PV is often replaced with the variation from long term data (since it is more real than a handful of specimens).

ndc is not related to tolerance - at least the calculation provided by AIAG. So, you can get a good % tolerance GRR but bad ndc. But, a bad ndc should relate to a high GRR as a % of the total variance.

This is why your data and the rest of the statistics would be handy for us to look at. Generating an academic data set will not be as beneficial. It is far easier to relate to a real situation.
 
J

Jirka

#4
Re: Impossible to achieve NDC greater than 2

Thank you so much for your fast response!

I must apologize cause I did a mistake with my explanations, usually the problem to achieve the proper NDC are with the pitch Ø.

So, I attach you an example of real R&R test. If according the 4th revision of the MSA we must achieve 5 NDC's, how we can reach that? It is real severe for us. I know that one solution could be to buy new measurement toolings which can read 0.0001 because the pitch diameter usually it's in millimess, but I'm afraid that this will not be still enough.

Anyway, what we are worried about is the answer to be given to our auditor...

Thanks again for your support
 

Attachments

Last edited:

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#5
Re: Impossible to achieve NDC greater than 2

Thread measurement is always a painful topic, so don't feel bad - it's not just you.

As far as your data, you have good news and bad news. The good news - if it is true - is that you have little variation between and within your measurements. The average range is .004. The bad news is that ndc looks to divide that range up into statistically significant pieces. The resolution of the device is .001, so if you had no gage error your ndc would be about 4. With error, it is nothing but down.

So, here is the issue. You could increase your ndc by increasing your variation...but it seem your customer may just prefer to have tight data and your current measurement technique. You only need a higher ndc for SPC to sense process variation. But .004 out of a .105 spec would not support more expensive gaging for typical measurements - especially if the variation is truly what you can expect over time.
 
J

Jirka

#6
Re: Impossible to achieve NDC greater than 2

Thanks for your support and most of all thank you to tell me that I'm not alone in that kind of thread issues :D

Yes, we really have a very accurate data by measuring the pitch diameter after many years of training of measuring in the same way, but although we have that accuracy, you can see that the GRR result it's more near to 30% than to 10%. In coated pitch diameter R&R's I would never trust a result of less than 10% because it's really really difficult to achieve, at least measured in 0.001 devices.

About the NDC, if we should have more variety for get a better NDC, then the GRR would increase, so it is a really neverending story.

In fact, any of our customers has ever complaint about NDC, it's just our auditor which wants to know why we can't achieve the 5 NDC's requested... so, what would you suggest me to explain him? That in this case NDC's would have more sense in SPC and that our customers are good with our GRR results and accuracy?

I hope to find a good "defence" and then be able to handle it properly...

Thanks again for your support.
 

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#7
Re: Impossible to achieve NDC greater than 2

I don't have the MSA book handy today - but I will in a day or so. But, I recall there are some tidbits buried in there that will help. The biggest defense is that the MSA book is a reference, not a requirement - so the auditor should not see the recommendation of ndc=5 as a mandate. I need to verify it does not crop up in the PPAP book under the MSA section - because that would make it a mandate (although misguided one).

Assuming it is not in the PPAP book, the justification is that the variation of your process is extremely small compared to the specification (very capable.) If you are not required to do SPC, then there is no point gaining a higher degree of statistical resolution (which is what ndc represents). If it is not specified verbally, it is very clear from the calculations used to develop ndc if the auditor was familiar with them.

Other misguided requirements come from customers that mandate this ndc requirement across the board in the CSRs - without having any idea what they are doing. After training the customer on this issue, they may have to provide a written waiver of their own requirement - which would be all the auditor would need.
 

Jim Wynne

Staff member
Admin
#8
Re: Impossible to achieve NDC greater than 2

I don't have the MSA book handy today - but I will in a day or so. But, I recall there are some tidbits buried in there that will help. The biggest defense is that the MSA book is a reference, not a requirement - so the auditor should not see the recommendation of ndc=5 as a mandate. I need to verify it does not crop up in the PPAP book under the MSA section - because that would make it a mandate (although misguided one).

Assuming it is not in the PPAP book, the justification is that the variation of your process is extremely small compared to the specification (very capable.) If you are not required to do SPC, then there is no point gaining a higher degree of statistical resolution (which is what ndc represents). If it is not specified verbally, it is very clear from the calculations used to develop ndc if the auditor was familiar with them.

Other misguided requirements come from customers that mandate this ndc requirement across the board in the CSRs - without having any idea what they are doing. After training the customer on this issue, they may have to provide a written waiver of their own requirement - which would be all the auditor would need.
All good advice. :bigwave:

As far as the PPAP manual is concerned, there's a note in MSA section (page 6 of the 4th edition) that says "Gage R&R acceptability criteria are defined in the Measurment Systems Analysis reference manual."

What that means is anyone's guess, given the fact that the MSA manual emphasizes the fact (as you point out) that it offers guidelines, not requirements. There will always be exceptions to the "rules," and we can only hope that customers and auditors will listen to reason. Otherwise, things like this become instances of customers unwittingly asking suppliers to make stuff up so the right numbers appear in the little boxes.
 

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#9
Re: Impossible to achieve NDC greater than 2

As far as the PPAP manual is concerned, there's a note in MSA section (page 6 of the 4th edition) that says "Gage R&R acceptability criteria are defined in the Measurement Systems Analysis reference manual."

What that means is anyone's guess, given the fact that the MSA manual emphasizes the fact (as you point out) that it offers guidelines, not requirements.
I was suspicious that was the case, and, as you mention. MSA book is emphatic that there is no one way or criteria to approve a gage - especially in the section with the 10/30 criteria. :rolleyes:

I admire their intentions, but they on occasion trip over their own feet.
 

Jim Wynne

Staff member
Admin
#10
Re: Impossible to achieve NDC greater than 2

I was suspicious that was the case, and, as you mention. MSA book is emphatic that there is no one way or criteria to approve a gage - especially in the section with the 10/30 criteria. :rolleyes:

I admire their intentions, but they on occasion trip over their own feet.
That's because they don't have actual feet, just guidelines.:tg:
 
Thread starter Similar threads Forum Replies Date
D Occurrence virtually impossible (thus low occurrence number) - FMEA Question Quality Tools, Improvement and Analysis 7
Jim Wynne Be careful about saying "impossible" Coffee Break and Water Cooler Discussions 1
A "Impossible" Is Often The "Untried" Coffee Break and Water Cooler Discussions 3
A That's impossible - George Washington supposedly threw a dollar across the Potomac Funny Stuff - Jokes and Humour 11
V What plans have been put in place to achieve quality objectives? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
D Is Failure to achieve a Goal a Major Non-Conformance? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
S Determining sample size for inspection to achieve x% confidence re defects Misc. Quality Assurance and Business Systems Related Topics 10
C How to achieve Page Ownership with a Wiki Document Control Systems, Procedures, Forms and Templates 2
C CRMs (Customer Relationship Management) to help achieve ISO 9001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
P ISO 15408 - How to achieve Certification to this Standard? Other ISO and International Standards and European Regulations 2
S Have you ever "used" an auditor to achieve a goal? Quality Manager and Management Related Issues 18
R How to Achieve Multi Objective (Responses) Optimization in Minitab Using Minitab Software 17
K Can we achieve certification to cover both the FSMS and QMS in one Audit? Food Safety - ISO 22000, HACCP (21 CFR 120) 2
U Defining Essential Performance to Achieve Freedom from Unacceptable Risk IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
T In-House Calibration of Thread Plugs/Rings to achieve a 4:1 Ratio Measurement Uncertainty (MU) 4
N New Seven Management Tools - Corporate Objective - Achieve 20% sales increase Misc. Quality Assurance and Business Systems Related Topics 3
S Suitable Sterilization Cycle to achieve Sterility Assurance Level of <10-6 ISO 13485:2016 - Medical Device Quality Management Systems 1
L How to achieve Inspector Consistency? Specialty vehicle manufacturer Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
M Means to Achieve Objectives and Targets - ISO14001 4.3.4 Miscellaneous Environmental Standards and EMS Related Discussions 1
R New to Quality - How to achieve goals company has given me Misc. Quality Assurance and Business Systems Related Topics 5
J Hindu Temple In Selangor First To Achieve ISO 9001:2000 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
Icy Mountain Personnel safety to achieve product quality - TS16949 Section 6.4.1 IATF 16949 - Automotive Quality Systems Standard 2
E Requirements to achieve accreditation for Automotive EMC Testing General Measurement Device and Calibration Topics 4
T Can a Fabless Semiconductor Company achieve TS16949 Certification (Registration)? IATF 16949 - Automotive Quality Systems Standard 48
D Can Minitab Return specification limits required to achieve a given Cpk? Using Minitab Software 11
S ISO 17025 Clause 5.9 - Calibration results - How to achieve what ISO wants ISO 17025 related Discussions 7
D Survey Question - Requires we have a process in place to achieve conformance IATF 16949 - Automotive Quality Systems Standard 3
S Personnel safety to achieve product quality as required by TS 16949 Clause 6.4.1 IATF 16949 - Automotive Quality Systems Standard 4
A ISO/TS16949, Clause 6.2.2.4 - Motivating employees to achieve quality objectives IATF 16949 - Automotive Quality Systems Standard 6
M Continuous Improvement - How to achieve - What tools/methods ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
B Gage R&R with NDC=1 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
J Question on very low NDC number with tolerable GRR ratio's Capability, Accuracy and Stability - Processes, Machines, etc. 7
C Acceptable NDC for %GR&R part inspection to Tolerance (%Tolerance) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
bobdoering "nds" or Number of Discriminate Samples - the Necessary Tool to Work With "ndc"! Imported Legacy Blogs 0
L Number of Distinct Categories" or NDC Calculation in MSA Studies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 8
J Is a ndc less than five is also acceptable? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
R Count NDC from Graphik Sample Range vs R chart Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
B Process Control - Selecting Parts - NDC & Study Variation Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
B Capability Study => Check %Study Variance and NDC? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
B NDC - Gauge R&R Studies with results where NDC = 1 or 0 Manufacturing and Related Processes 1
Y Poor Gage R&R Results - High R&R result and poor ndc value Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
B Help with NDC (Number of Distinct Categories) Please Six Sigma 2
R ndc=0 and GRR=14% ??? Interpretation - basic question Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
S Using Range/Tolerance to Guesstimate NDC or % SV? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
C Conclusions from weird GR&R (Gage R&R) Results - 9 of 12 points have NDC of 1 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
M MSA Manual asks for NDC higher than 5 - NDC (number of distinct categories) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
T Gage R&R Sheet Formula error?! NDC less than 4 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
V Can Anyone Give a Simple Explanation on ndc (number of distinct categories)? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
J %Study %Tolerance NDC - R&R Criteria Acceptability when the Tolerance is used Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
D Gage R&R %Tolerance NDC Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12