In a risk analysis, how can we tie mobile app security breach to ISO 14971?

S

snoopy2017

Hi everyone,

In a risk analysis, how can we tie security breach (e.g. losing confidentiality of patient information) to ISO 14971? What is the severity level of harm for loss of confidentiality of information in a mobile app? I would think we should do that exploitability analysis first as per FDA's 2016 guidance on cybersecurity. Has anyone had first hand experience doing this analysis tied to 14971? If so, could you provide some guidance or a sample template of this type of a security risk analysis? Thank you. I would appreciate any reply.
 

QAengineer13

Quite Involved in Discussions
I would highly recommend you purchasing and reading AAMI TIR 57 Principles for medical device security which addresses security risk management in the context of ISO 14971, it creates a clear linkages between consideration of safety and security, this TIR is recognized by the FDA and referenced in their post market guidance .

Example: Images from the TIR 57

1540853190526.png

1540853158328.png


Anothe recommendation is EN 82304-1 Health Software General requirmetns for product safety to look into which applies to the SAFETY and SECURITY of Health Software Products designed to operate on general computing platforms and intended to be placed on the market without dedicated hardware.

-Rk
 

Mark Meer

Trusted Information Resource
Haven't had to go through the process myself (yet), but at the highest level of assessing potential harm, perhaps start by assessing:

- Can a breach be used to interfere with the device function?
- Can a breach be used to corrupt/overwrite existing data? If so, what would be the worst-case result?
- Can data potentially stolen/read in event of breach be used to personally identify a patient?
- Can this data be used to infer patient diagnoses and treatments?

Perhaps a bit simplistic depending on your application, but a starting point for what it's worth...
MM.
 
T

tomshoup

The short answer to snoopy2017's question is to use the steps of 14971 to assess the cyber risk: RISK ANALYSIS: identify the hazard (loss of confidentiality), identify the sequence of events that can lead to such a breach, identify the resuting hazardous situation (loss of confidentiality = HIPPA violation, public exposure, etc.), identify the severity and probability. Then RISK EVALUATION: is this risk acceptable. If the risk is acceptable, stop. If the risk is not acceptable, identify and implement control measures to reduce the risk and verify their implementation and effectiveness.

FDA's guidance on cybersecurity as it relates to submissions parallels 14971:
1550878828166.png

Also, the Open Web Application Security Project has a good discussion on threat modeling that will help with this.
www.owasp.org/index.php/Application_Threat_Modeling.

Regards,
Tom 1550878828166.png
 
Thread starter Similar threads Forum Replies Date
T Article 22 procedure pack risk analysis EU Medical Device Regulations 1
S Risk analysis on patient monitor ISO 14971 - Medical Device Risk Management 5
J Determination of software safety class (62304) prior to software risk analysis ISO 14971 - Medical Device Risk Management 3
G FDA requirements for risk analysis US Food and Drug Administration (FDA) 2
L HA vs risk analysis ISO 14971 - Medical Device Risk Management 2
K Help with ISO 14971: Benefit-Risk Analysis ISO 14971 - Medical Device Risk Management 3
L Risk analysis Manufacturing and Related Processes 4
J Risk Analysis for Proficiency Testing Reliability Analysis - Predictions, Testing and Standards 1
S What is your favorite Usability Risk Analysis tool? IEC 62366 - Medical Device Usability Engineering 5
H Risk Analysis and Probability of Occurrence ISO 14971 - Medical Device Risk Management 3
Booker Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
P Benefit risk analysis on pFMEA ISO 14971 - Medical Device Risk Management 10
silentmonkey Overall Benefit/Risk Analysis - Risk Management VS Clinical Evaluation ISO 14971 - Medical Device Risk Management 3
R The term "Benefit Risk Ratio" in EU MDR, do I need to present benefit risk analysis as a RATIO Risk Management Principles and Generic Guidelines 4
D Risk Analysis using Monte Carlo Simulation instead of Scoring and Heat Map Risk Management Principles and Generic Guidelines 2
E Normal Condition Hazards in Risk Analysis ISO 14971 - Medical Device Risk Management 3
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
R ECG Risk Analysis Standards ISO 14971 - Medical Device Risk Management 2
adir88 Documenting Risk Control Option Analysis ISO 14971 - Medical Device Risk Management 8
MrTetris Should potential bugs be considered in software risk analysis? ISO 14971 - Medical Device Risk Management 5
M IATF 16949 (6.1.1 - Planning and Risk Analysis for a remote site) Process Maps, Process Mapping and Turtle Diagrams 5
Dobby1979 Risk Analysis & Technical File - What detail goes in the Risk Management Report ISO 14971 - Medical Device Risk Management 5
M An example of risk analysis of class I MD ISO 14971 - Medical Device Risk Management 36
T Risk analysis of QMS software - Validating software we use for QMS ISO 13485:2016 - Medical Device Quality Management Systems 7
B Grouping of Products for Risk Analysis ISO 14971 - Medical Device Risk Management 9
A Risk-benefit Analysis - Hazard Analysis (HA) and FMEAs ISO 14971 - Medical Device Risk Management 18
R The difference b/w FMEA & Risk analysis as per iso 14971 ISO 14971 - Medical Device Risk Management 8
K Risk Analysis Updates due to complaints ISO 14971 - Medical Device Risk Management 10
S The Severity of a Medical Device Hazard - Risk Analysis Clarification ISO 14971 - Medical Device Risk Management 6
Ed Panek Transition to IEC 60601 4th Edition - Risk Analysis and test submissions CE Marking (Conformité Européene) / CB Scheme 2
Q Risk / benefit Analysis in Risk Management Report CE Marking (Conformité Européene) / CB Scheme 12
R IATF 16949 Clause 6.1.2.1 - Lessons Learned and Risk Analysis IATF 16949 - Automotive Quality Systems Standard 6
S Risk analysis 6.1 and contingency plans 6.1.2.3, are they related? IATF 16949 - Automotive Quality Systems Standard 26
B Software Class A - Lengthy further risk analysis IEC 62304 - Medical Device Software Life Cycle Processes 9
W Biocompatibility Risk Analysis for Clinical Practitioner 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
FoGia Risk Analysis of a Medical Device Accessory ISO 14971 - Medical Device Risk Management 4
S How we can use risk analysis for suppliers IATF 16949 - Automotive Quality Systems Standard 6
I Medical Device Software Risk Analysis ISO 14971 - Medical Device Risk Management 4
Q Risk Analysis - Same Risk Treatment for Context and Interested Parties ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
C Risk Analysis for COTS/OTS Risk Management Principles and Generic Guidelines 4
M IATF 16949 Cl. 8.7.1.4 - Risk analysis for decision making about rework IATF 16949 - Automotive Quality Systems Standard 2
E Risk Analysis - Events which may cause to Data Loss ISO 14971 - Medical Device Risk Management 12
W Risk Benefit Analysis - ISO 14971:2012 Requirements ISO 14971 - Medical Device Risk Management 27
F Medical Device HACCP (Hazard Analysis and Critical Control Point) Risk Management ISO 14971 - Medical Device Risk Management 2
Q Risk Tools in ISO 31010 - Root Cause Analysis vs. Cause-and-effect Analysis ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
S Organizing Risk Analysis and Controls for a New Medical Device (ISO 14971) ISO 14971 - Medical Device Risk Management 4
S Please review my Risk Analysis Table ISO 14971 - Medical Device Risk Management 13
K Risk Analysis and "Information for Safety" / Labeling ISO 14971 - Medical Device Risk Management 10
M Risk analysis - ISO/TS 16949 clause 7.2.2.2 IATF 16949 - Automotive Quality Systems Standard 2
C Help with Risk/Benefit Analysis Self-help Device for Diabetics ISO 14971 - Medical Device Risk Management 3

Similar threads

Top Bottom