Hello everyone,
We have an out-of-control calibration on one of our Thermocouple Thermometer. This requires outside calibration based on specification/documents that what we have.We have the manual of the said instrument and it states that it can only be calibrated by an authorized personnel from that manufacturer. There is also a provisions there, that it can be calibrated. It must be send to the OEM where we purchase the said instrument. Here are my concerns:
1. Would it be possible for us to calibrate the said equipment using the calibration procedure on the manual?Please advice.
2. Can we provide an ECN to change the calibration of the said equipment from outside to in-house calibration? Please comment.
3. Aren't we violating the 4.11 element regarding calibration? Please comment.
Any feedback would be appreciated.
Thank you very much.
Vash Stampede
CJacobsen
You certainly can calibrate the instrument in-house - providing you have:
1 - An approved calibration procedure. The one in the manufacturers manual is ok to use if your system allows you to. In otherwords, your SOP that covers calibration should specify the sources of calibration procedures. Just make sure that 'manufacturer' is one of them.
2 - You have the proper, calibrated standards and equipment needed to calibrate the thermocouple.
I don't know why you would need an ECN to change where an instrument is calibrated. As long as in your documentation states what sources you can use for calibration procedures, there is no reason why you cannot calibrate the instrument in-house. The consideration here is the use of the thermocouple. If it is never used to calibrate or verify another instrument and only used to take readings, no reason to have it calibrated outside.
You are only violating 4.11 (or 7.6 in the FDIS) if in your own procedure you do not allow yourself the flexibility to decide where an instrument is best calibrated.
Jerry Eldred
I totally agree with the first reply. As thermocouple thermometers is within one of my areas of specialty, please feel free to email me directly and let me know the specifics (manufacturer and model). I may very well have calibrated that type.
They are not too difficult to calibrate. Although some manufacturers in recent years have been attempting to convince users not to calibrate their own equipment by such comments.
If you will give me some details on the meter, I ma be able to give some advice on what is needed to calibrate it correctly. Even if you do use it to measure other items, if you calibrate it correctly, and have the proper documentation coverage, it should not be a problem.
G OEM (Original Equipment Manufacturer) Calibration of In-House Measurement Equipment General Measurement Device and Calibration Topics 19
A AS9100 - 7.1.5.2 In-House Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 25
dhuot1990 In-House Calibration Program - Biotech Industry General Measurement Device and Calibration Topics 3
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
L In house calibration - Our CMM's are outside certified can I use them to certify our standards General Measurement Device and Calibration Topics 4
Ed Panek ISOTECH TTI-10 found out of calibration - Calibration house was wrong? General Measurement Device and Calibration Topics 7
H In-house device calibration record - Actual values obtained required? General Measurement Device and Calibration Topics 9
T In house Calibration? Measurement equipment traceability requirements - ISO9001 General Measurement Device and Calibration Topics 12
G In-House Calibration of Gas Flow Meters General Measurement Device and Calibration Topics 6
4 In-House Thread Wire Calibration General Measurement Device and Calibration Topics 5
C ISO17025 for In-House Calibration Labs: Is it required? ISO 17025 related Discussions 30
A Is NIST traceability required if a Refractometer calibration is done in-house? Calibration Frequency (Interval) 6
D Measurable objectives for our in-house calibration department General Measurement Device and Calibration Topics 6
M In house Volt Meter Calibration General Measurement Device and Calibration Topics 5
N ISO 17025 Requirements for Environmental Conditions for In-House Calibration ISO 17025 related Discussions 5
W Force Calibration - In-House Load Cell Calibration General Measurement Device and Calibration Topics 7
M Looking for a copy or template of a In House Calibration Report for a Dial Indicator General Measurement Device and Calibration Topics 1
C Flow Meter Calibration - Outside Calibration Service vs. In-House Calibration General Measurement Device and Calibration Topics 3
S Using Prepared In-House Reference Materials (RM) for Equipment Calibration & Verifica General Measurement Device and Calibration Topics 3
S In-House Equipment Calibration Report Format & Template Example Needed Document Control Systems, Procedures, Forms and Templates 1
1 Pharma Industry stance on ISO 17025 Accreditation - In-House Calibration Laboratories ISO 17025 related Discussions 3
T Reference Spectrum Generator for In-House Calibration General Measurement Device and Calibration Topics 2
F Temperature Requirements for an In-House Calibration Laboratory General Measurement Device and Calibration Topics 8
T In-House Calibration of Thread Plugs/Rings to achieve a 4:1 Ratio Measurement Uncertainty (MU) 4
R In-House Calibration Facility Requirements - Setting up a new in-house laboratory General Measurement Device and Calibration Topics 4
B In-House pH Meter Calibration General Measurement Device and Calibration Topics 13
B Developing In-House Calibration Capacities General Measurement Device and Calibration Topics 5
E Gauge Block Grade recommended use for in house calibration General Measurement Device and Calibration Topics 1
C Setting up Pressure Calibration In-House Lab General Measurement Device and Calibration Topics 7
J In-House Calibration Requirements - Use of Grade 2 set of gage blocks General Measurement Device and Calibration Topics 2
D TS 16949 In-House Calibration Requirements - Certification Required? IATF 16949 - Automotive Quality Systems Standard 26
T In-House Calibration - Digital Micrometer found Out of Calibration General Measurement Device and Calibration Topics 10
R Self (In-House) Calibration of Measuring Tools (vernier, micrometers etc.) General Measurement Device and Calibration Topics 17
D Calibration vs. Verification - Can we calibrate dial indicators in-house? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
ScottK Soliciting recommendations for a fast turnaround calibration house - Mass Flow Meter Manufacturing and Related Processes 1
M Gage Block In-House Calibration - Blocks lost and users keep exchanging blocks General Measurement Device and Calibration Topics 2
V Is permission from Registrar needed for deciding on in-house calibration? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
H In-house Calibration - Verniers using gauge blocks General Measurement Device and Calibration Topics 5
J Thread Gage Calibration: In-House or Outsourced? General Measurement Device and Calibration Topics 7
E Can anyone recommend a good/inexpensive calibration house? General Measurement Device and Calibration Topics 9
S Calibration System was In-House now Outside (External Service Provider) General Measurement Device and Calibration Topics 14
U Calibration House - Where I can get a PG 13.5 Gut thread ring gage calibrated? General Measurement Device and Calibration Topics 1
T In-house Calibration Procedures in our FAA Approved Repair Station Federal Aviation Administration (FAA) Standards and Requirements 11
BradM Certificate for In-House Calibration General Measurement Device and Calibration Topics 25
K What is the difference between in house calibration and verification? General Measurement Device and Calibration Topics 21
A Auditor has rejected our in-house gage verification - Calibration in 17025:2005 ISO 17025 related Discussions 21
J In-House Calibration Certificate wording - Confidence Level or Level of Confidence? Document Control Systems, Procedures, Forms and Templates 3
B In-House Calibration form needed Document Control Systems, Procedures, Forms and Templates 1
J Using Outside Accredited Calibration Laboratories vs. In-house Calibrations General Measurement Device and Calibration Topics 10
C In-House Particle Counter Calibration General Measurement Device and Calibration Topics 12