In-tolerance Probability - Level of Confidence - Uncertainty and Proficiency

Marc

Hunkered Down for the Duration
Staff member
Admin
#1
Date: Thu, 14 Oct 1999 11:36:54 -0400
From: "James D. Jenkins"
To: Greg Gogates
Subject: RE: Uncertainty and Proficiency RE8

Dick,

Pertaining to the level of confidence being the most important thing, you are absolutely correct and while the GUM annex G doesn't use the term "In-tolerance Probability", it uses "Level of Confidence". And when talking about the "level of confidence" of a device in which the specifications are utilized as error containment of the measuring parameter bias, we are talking about the confidence in the error of the readings being contained within those specifications, that is "In-Tolerance Probability". As you pointed out, the confidence level is really the most important thing because of the large effect it has on the uncertainty value. That is why the concerns expressed in the GUM annex G must be dealt with relative to the category B estimates. And of course to deal with them we should use the methods of statistics given for such a task to maintain conformance to the GUM and common sense. You should also read my response in "RE: Uncertainty and Proficiency RE4".

I have found two techniques for determining the "In-tolerance Probability" for a particular devices parameter, with method number one being the most accurate. The mathematical methodology for doing this can be found at http://www.isgmax.com under the section for the Freeware software program "Category B Degrees of Freedom Estimator". You can also download this program and methodology for you own evaluation and use. Dr. Howard Castrup developed this method for estimating the "delta u" variable in Equation G.3 of the GUM so that the degrees of freedom can be computed for Category B uncertainties. I am sure he realized as you do that this "Level of Confidence" is extremely important and deserves some attention.

The "Category B Degrees of Freedom Estimator" gives us 3 options for formulating our findings in a statistical statement with the resultant being that "x lies within +/- L with C% confidence".

Options
1. Containment Limits and Probability.
2. Containment Limits and History of Observations.
3. Containment Limits and Number of Cases.

The two data collection techniques for determining the "In-tolerance Probability" that I have used:

1. Using historical information from previous calibration performance tests we can formulate the "In-Tolerance Probability" pertaining to the parameter in question with actual values. Now you may find that you lack enough historical information to be of statistical value either for the particular unit even when combined with others of the same make/model being serviced under the same conditions. In that case you could contact either the service which provides the calibration for you or the manufacturer and request information pertaining to the in-tolerance probability of the devices parameter in question for the given assigned calibration interval. I have found that technicians with years of experience with a particular make/model already know this probability answer. But hard data is always preferred and with a little digging in the archives they should be able to give you enough data to support your analysis. Enter this information into the "Category B Degrees of Freedom Estimator" or for those math lovers use the formula and a sharp pencil.

2. Not preferred, but in a crunch, an estimate from someone with experience with the device in question is better than not addressing the issue at all. Even if all we have is the data from the current calibration and hopefully the uncertainty of that data, we can still estimate a reasonable close value of "In-Tolerance Probability"(Level of Confidence). If you use a software database for your analyses like the one included with UncertaintyAnalyzer, you can tag and revisit your analysis for improvement once you acquire a reasonable amount of data to support a statistical estimate. Use this estimate as the "Confidence Level" in your equation.

There are numerous advantages to using software in computing and storing all the details pertaining to the analysis, although it is extremely important for those using software to still understand the concepts as well as have an expectation of the resultants value. I find that using software is much more efficient with a higher probability of getting the correct mathematical answer. As a Metrologist, I want to invest my time identifying and dealing with the physical components of the measurement and let my computer perform the redundant math.

I hope this answers your questions. I find that demonstrating this process with the software (as I do in the measurement uncertainty classes I teach) is somewhat easier than trying to explain this in writing where I am unable to sense the comprehension or deal with live questions. Further information pertaining to the software I use and the classes I teach can be found at http://measurementuncertainty.com .

Sincerely,
James D. Jenkins
Quametec Corporation
 
Elsmar Forum Sponsor
Thread starter Similar threads Forum Replies Date
Ronen E Tolerance intervals (?) question - Flow Rate Probability Range Statistical Analysis Tools, Techniques and SPC 6
G 0.00005" tolerance on micrometer with 0.0005" resolution? General Measurement Device and Calibration Topics 4
N Drawing tolerance vs. Measurement device General Measurement Device and Calibration Topics 4
I Control Plan (Product/Process specification/ Tolerance) acceptance FMEA and Control Plans 27
J Mechanical inspection techniques of close tolerance parts Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
Proud Liberal Small radius with tight tolerance on a torus General Measurement Device and Calibration Topics 4
L GRR for a tolerance that has changed Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
T How to justify this High %Tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
D Tolerance definition based on expected Cp/cpk Reliability Analysis - Predictions, Testing and Standards 14
N % Tolerance - Type 1 study on the gages, then a gage R&R (ANOVA) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
S Tolerance limits for micro balance General Measurement Device and Calibration Topics 1
J Determining SPC tolerance Statistical Analysis Tools, Techniques and SPC 21
Q % Study variation low, % tolerance high - GR&R Interpretation help Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
R Uncertainty in measurement larger than tolerance Measurement Uncertainty (MU) 2
WEAVER Electronic Weighing Scale Calibration Tolerance Manufacturing and Related Processes 1
I Determining Calibration Tolerance of a Measurement Device General Measurement Device and Calibration Topics 2
D Tolerance when verifying an analytical balance General Measurement Device and Calibration Topics 3
E Discussion between co-worker on tolerance and uncertainty and how to apply it. Thoughts? 17025 ISO 17025 related Discussions 1
C Thread Tolerance - Can I use a 3A Ring on a part that the drawing calls out for a 2A? Manufacturing and Related Processes 5
T GRR based on part tolerance or process variation. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
A Excel OOT (Out of Tolerance) Formula Excel .xls Spreadsheet Templates and Tools 3
N Tolerance Interval plots in Minitab Statistical Analysis Tools, Techniques and SPC 0
S Question about a basic additive variance/tolerance example Quality Tools, Improvement and Analysis 2
G Assigning a calibration tolerance - An x-y coordinate machine - Uncertainty as my verification tolerance General Measurement Device and Calibration Topics 4
R Feeler Gauge Calibration - What tolerance is used for calibrating feeler gauges? General Measurement Device and Calibration Topics 4
S SPC (Statistical Process Control) for Unilateral Tolerance - Questions Statistical Analysis Tools, Techniques and SPC 6
B Gage R&R Acceptable (10-30%), deduct Total Variation from Tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
Y How to fix weight tolerance for plastic injection molded part? Does it vary with material groups? Manufacturing and Related Processes 1
F Position tolerance check gage manufacturing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
M GD&T tolerance or band? Using a symbol like parallelism // Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
J What is the difference between Process Variation and Tolerance? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
J Gage R&R without tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
A Calibration of Pressure Gauge, How to Determine Calibration Tolerance General Measurement Device and Calibration Topics 1
V Using K-Factor(Tolerance Interval) Analysis for Design Verification Statistical Analysis Tools, Techniques and SPC 3
C Acceptable NDC for %GR&R part inspection to Tolerance (%Tolerance) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
O Cpk for Unilateral Tolerance Capability, Accuracy and Stability - Processes, Machines, etc. 9
S Doubts about SPC taken in Machining - Part has +-0.01 Tolerance Statistical Analysis Tools, Techniques and SPC 5
A Calibration tolerance for data logger General Measurement Device and Calibration Topics 3
N Straightness Gage Tolerance General Measurement Device and Calibration Topics 0
G Calculating Tolerance of a Panel Meter with a 2 degrees of an Arc General Measurement Device and Calibration Topics 0
Hami812 Cpk Setting Tolerance - Cart before the Horse? (Wifi Routers) Capability, Accuracy and Stability - Processes, Machines, etc. 3
kedarg6500 Excel spreadsheet for Unilateral Tolerance Study wanted Excel .xls Spreadsheet Templates and Tools 2
U Control Charting a Feature with Bonus Tolerance Quality Tools, Improvement and Analysis 10
J Process Capability - How much percentage of tolerance? VDA Standards - Germany's Automotive Standards 3
B Gage R&R MSA - Variation, Tolerance, Variance by different methods IATF 16949 - Automotive Quality Systems Standard 23
R Fixing of tolerance limit - Electronic weighing balance with the capacity of 150 Kgs General Measurement Device and Calibration Topics 5
bobdoering Unilateral Tolerance Capability Calculations Capability, Accuracy and Stability - Processes, Machines, etc. 1
bobdoering Unilateral Tolerance Capability Calculation Imported Legacy Blogs 0
H Tolerance, decimal places, and measuring equipment capability Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
G Out of Dimensional Tolerance - MRB Audit Issue Nonconformance and Corrective Action 11

Similar threads

Top Bottom