All:

I recently started a position with a manufacturing company of medical devices. They have process sheets/routers/work instructions in place for every job that runs through the shop. When I arrived in this position there were several of these routers sitting in a box that were missing signatures for applicable steps (even though they say that the step was definitely completed), dates for the work completed, changes in quantities throughout that were not explained or associated to NCMRs, etc. At this point, some of these routers are 1-4 yrs old and have never been closed out due to these documentation gaps.

How can I move to close these out? Can I create a Letter to File to close the routers without the applicable information?

Any thoughts on this would be appreciated! Thank you!
 
Elsmar Forum Sponsor

yodon

Staff member
Super Moderator
#2
Sounds like a clear case for initiating a CAPA. You have a systemic process issue and you need to both make the corrections and eliminate the root cause(s). Let the CAPA drive how you close out the routers. There should be some defensible rationale behind whatever you do and without the CAPA, you'd just be making (what might appear to be) arbitrary decisions.
 
Q

QA-RA Jones

#3
I completely agree with you on the issuance of a CAPA. This portion is complete and we have already identified solutions to making sure this does not happen in the future (root cause eliminated). My only issue at this point is specifically related to how we can close out the past documents that have gaps associated with them.

Things we know about these process sheets:
We know the final quantity delivered. (inspections are complete for the final quantity)
We know that no complaints have been received on the related products produced through the incomplete routers.

Correctively for the routers with gaps:
Could we close these documents out through just identifying a statement in the CAPA resolution that we know what was delivered and when and that we also know no complaints have been received on the quality of the product, therefor the routers will be closed?
 

yodon

Staff member
Super Moderator
#4
I think you're on the right track. I would probably suggest that the rationale for closure (corrections) include some indication that there's no impact to safety or efficacy of the product. The 'no complaints' may not be the strongest argument but can be supporting.

These are tough and companies get in tough positions trying to dig out. What you're doing sounds like it will put you in a defensible position.
 
Thread starter Similar threads Forum Replies Date
H Incomplete IVD kit under IVD Regulation EU Medical Device Regulations 1
I Audit Nonconformances - Can reported corrective actions be incomplete? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
M Incomplete Unannounced Audit - No Active Production at the time ISO 13485:2016 - Medical Device Quality Management Systems 11
H Incomplete forms and records missing date or a name Records and Data - Quality, Legal and Other Evidence 6
D Incomplete DHF: Re-do or plug gaps? Design and Development of Products and Processes 6
C Incomplete Internal Audit at time of Re-Certification Audit Internal Auditing 5
J How to deal with incomplete forms Nonconformance and Corrective Action 14
B Filling Out FDA Form 3654 when Testing is Incomplete 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Q Rectification and/or Corrections to existing 'incomplete' records Records and Data - Quality, Legal and Other Evidence 7
D Monitoring Number of Incomplete Items & p-chart - Medical Records Statistical Analysis Tools, Techniques and SPC 2
M Design Input - 820.30(c) - Addressing incomplete, ambiguous... requirements Design and Development of Products and Processes 1
A When someone refuses to follow a process.... Misc. Quality Assurance and Business Systems Related Topics 15
E Software maintenance Process Software maintenance Process to IEC 6204? IEC 62304 - Medical Device Software Life Cycle Processes 3
R AS5553 Clause 3.1.7 f - "Implement a returns process....." AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
normhowe "The Problem with Quality Management: Process orientation, controllability and zero-defect processes as modern myths" Book, Video, Blog and Web Site Reviews and Recommendations 2
Judy Abbott General temperature used in the blasting process and laser process Manufacturing and Related Processes 2
B SOP for CNC turret punching machine for sheet metal process Manufacturing and Related Processes 0
A API Monogram audit review process Oil and Gas Industry Standards and Regulations 4
R AS9102 FAI Change in Material / Process Supplier AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
A Process mapping Process Maps, Process Mapping and Turtle Diagrams 1
R MDEL Process Canada Medical Device Regulations 4
optomist1 Rates Daily or Hourly Process Improvement Training Consultants and Consulting 2
S Manufacturing Process FDA FOIA Medical Device and FDA Regulations and Standards News 3
S Manufacturing Process FDA FOIA US Food and Drug Administration (FDA) 4
B Toyota PPAP Process - Three Questions APQP and PPAP 3
R Changes vs CMO - How can we simplify this process? Supplier Quality Assurance and other Supplier Issues 3
A Ethics Committee Review Process for IVD Products EU Medical Device Regulations 2
V Laser Welding Process - Impact on Electrical Properties Reliability Analysis - Predictions, Testing and Standards 4
Q Process: Knowledge Section 7.1.6 of ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
L Documented Information in Internal Audits Process (9.2) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
A Sampling plan for in-process QC (medical devices) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
R MRB (Material Review Board) Process using MS Sharepoint or MS Teams Manufacturing and Related Processes 2
M Clinical Benefit of device that only aids in a process for managing or treating disease EU Medical Device Regulations 2
C In-process inspection - Tooling and assembly lines for automotive companies AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
M Efficacy of an IT process after a cyber attack ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
N Riveting - special process Manufacturing and Related Processes 11
M Material incoming to the production process reflected in PFMEA FMEA and Control Plans 9
A API Spec Q1 Purchasing Process - Supplier Reevaluation based on Supplier Risks 5.6.1.4 Oil and Gas Industry Standards and Regulations 17
B Handling lower detection limits for SPC and process performance Statistical Analysis Tools, Techniques and SPC 1
D Measurables for Plastic Injection molding process Manufacturing and Related Processes 1
S Cleaning process center change ISO 13485:2016 - Medical Device Quality Management Systems 4
Z Rapid audit template for plastic parts manufacturing process Manufacturing and Related Processes 12
R Inspection and Work order process Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3
A How to reduce the process SPC monitoring Capability, Accuracy and Stability - Processes, Machines, etc. 3
John Predmore Configuration Management as a process instead of a procedure AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
R PCBA process validation Qualification and Validation (including 21 CFR Part 11) 2
U Internal Auditor not trained but done Audit for some process Nonconformance and Corrective Action 5
B Two excellent examples of process capability analysis from Quality Magazine Capability, Accuracy and Stability - Processes, Machines, etc. 5

Similar threads

Top Bottom