Incomplete Process Sheets Closure - DHR - Doc Control

Q

QA-RA Jones

#1
All:

I recently started a position with a manufacturing company of medical devices. They have process sheets/routers/work instructions in place for every job that runs through the shop. When I arrived in this position there were several of these routers sitting in a box that were missing signatures for applicable steps (even though they say that the step was definitely completed), dates for the work completed, changes in quantities throughout that were not explained or associated to NCMRs, etc. At this point, some of these routers are 1-4 yrs old and have never been closed out due to these documentation gaps.

How can I move to close these out? Can I create a Letter to File to close the routers without the applicable information?

Any thoughts on this would be appreciated! Thank you!
 
Elsmar Forum Sponsor

yodon

Staff member
Super Moderator
#2
Sounds like a clear case for initiating a CAPA. You have a systemic process issue and you need to both make the corrections and eliminate the root cause(s). Let the CAPA drive how you close out the routers. There should be some defensible rationale behind whatever you do and without the CAPA, you'd just be making (what might appear to be) arbitrary decisions.
 
Q

QA-RA Jones

#3
I completely agree with you on the issuance of a CAPA. This portion is complete and we have already identified solutions to making sure this does not happen in the future (root cause eliminated). My only issue at this point is specifically related to how we can close out the past documents that have gaps associated with them.

Things we know about these process sheets:
We know the final quantity delivered. (inspections are complete for the final quantity)
We know that no complaints have been received on the related products produced through the incomplete routers.

Correctively for the routers with gaps:
Could we close these documents out through just identifying a statement in the CAPA resolution that we know what was delivered and when and that we also know no complaints have been received on the quality of the product, therefor the routers will be closed?
 

yodon

Staff member
Super Moderator
#4
I think you're on the right track. I would probably suggest that the rationale for closure (corrections) include some indication that there's no impact to safety or efficacy of the product. The 'no complaints' may not be the strongest argument but can be supporting.

These are tough and companies get in tough positions trying to dig out. What you're doing sounds like it will put you in a defensible position.
 
Thread starter Similar threads Forum Replies Date
H Incomplete IVD kit under IVD Regulation EU Medical Device Regulations 1
I Audit Nonconformances - Can reported corrective actions be incomplete? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
M Incomplete Unannounced Audit - No Active Production at the time ISO 13485:2016 - Medical Device Quality Management Systems 11
H Incomplete forms and records missing date or a name Records and Data - Quality, Legal and Other Evidence 6
D Incomplete DHF: Re-do or plug gaps? Design and Development of Products and Processes 6
C Incomplete Internal Audit at time of Re-Certification Audit Internal Auditing 5
J How to deal with incomplete forms Nonconformance and Corrective Action 14
B Filling Out FDA Form 3654 when Testing is Incomplete 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Q Rectification and/or Corrections to existing 'incomplete' records Records and Data - Quality, Legal and Other Evidence 7
D Monitoring Number of Incomplete Items & p-chart - Medical Records Statistical Analysis Tools, Techniques and SPC 2
M Design Input - 820.30(c) - Addressing incomplete, ambiguous... requirements Design and Development of Products and Processes 1
F Rules of process outsourcing in China China Medical Device Regulations 0
B Process / Procedure - Radiographic (X-Ray) Non-Film Document Control Systems, Procedures, Forms and Templates 0
P Certification process for registered device China Medical Device Regulations 11
J NCR- Failure of contract review process - NADCAP audit AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
Melissa Process Validation of Rotary Heat Sealer Speeds Design and Development of Products and Processes 4
C Validation of process for releasing the UDI EU Medical Device Regulations 4
Melissa Risk Management Process, How far do I need to go? ISO 14971 - Medical Device Risk Management 10
C Validation of process for production and servicing 5.7.1.5 API Spec Q1, 9th Edition Oil and Gas Industry Standards and Regulations 3
Z IA Construction Procurement Process Internal Auditing 3
H Risk Management Plan in agile process ISO 14971 - Medical Device Risk Management 11
DuncanGibbons Process flow & PFMEA for production planning and simulation activities? Process Maps, Process Mapping and Turtle Diagrams 7
N Help with understanding Process Controls Manufacturing and Related Processes 7
M VDA 6.3 – Workshop for Certified Process Auditor VDA Standards - Germany's Automotive Standards 1
C Process Consistency Manufacturing and Related Processes 5
I PFD (Process Flow Diagram) approach. Setup details as part of PFD? APQP and PPAP 5
K Business Process Flowchart Process Maps, Process Mapping and Turtle Diagrams 2
L Looking for Control Plan and FMEA Stamping process FMEA and Control Plans 2
J License renewal process in Iran Other Medical Device Regulations World-Wide 0
J Definition Outsourced process - Clear definition - 13485 Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 5
S IS0 13485 process flows ISO 13485:2016 - Medical Device Quality Management Systems 2
H Verification Process for the existing MDS Product IEC 62304 - Medical Device Software Life Cycle Processes 5
C Revalidation of Process Equipment for Equipment Transfered to New Facility. ISO 13485:2016 - Medical Device Quality Management Systems 5
Q Process map Evaluation and Analysis Method Process Maps, Process Mapping and Turtle Diagrams 5
S In process inspection Manufacturing and Related Processes 2
S Brexit 100% inspection during in process inspection Manufacturing and Related Processes 11
A Complaint review as part of the complaint handling process? ISO 13485:2016 - Medical Device Quality Management Systems 3
mustomutlu Process Validation Final Report Other Medical Device and Orthopedic Related Topics 2
Q Process Matrix_Audit Matrix Quality Management System (QMS) Manuals 4
Q Process description for outsourced processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
Q Process matrix examples of ISO 9001 & 14001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
J Need Help with FPY Data in Assembly Process Manufacturing and Related Processes 7
A When someone refuses to follow a process.... Misc. Quality Assurance and Business Systems Related Topics 27
E Software maintenance Process Software maintenance Process to IEC 6204? IEC 62304 - Medical Device Software Life Cycle Processes 3
R AS5553 Clause 3.1.7 f - "Implement a returns process....." AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
normhowe "The Problem with Quality Management: Process orientation, controllability and zero-defect processes as modern myths" Book, Video, Blog and Web Site Reviews and Recommendations 2
Judy Abbott General temperature used in the blasting process and laser process Manufacturing and Related Processes 2
B SOP for CNC turret punching machine for sheet metal process Manufacturing and Related Processes 0
A API Monogram audit review process Oil and Gas Industry Standards and Regulations 4
R AS9102 FAI Change in Material / Process Supplier AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4

Similar threads

Top Bottom