Inhouse Calibration Plus TUS in accordance with AMS 2750 rev E?

B

BizPro

#21
Hello!

Nadcap heat treating, AMS2750, and pyrometry in general is not only my expertise, it is my favorite special process.

1. YES! You can perform TUS (Temperature Uniformity Survey), equipment calibrations such as temp controllers, chart recorders, and timers in house...given you have the appropriate required calibrations for your standards used. Sensitivity is very easy to check once someone shows you how to do it (Fluke 724 or other multi-calibrator).
2. You will need ATP (approved test procedures) for how you perform TUS.
3. You will need to ensure that your standards are calibrated every three months (data logger and multi-calibrator).
4. You will need a data logger (data recording unit), calibrated thermocouples, and a multi-calibrator (Fluke 724). You may need a timer depending the Make and Model of your recorder (chart paper w/pens require a calibrated timer).
5. You will need diligent record keeping. In the top 10 findings for Nadcap Heat Treating (AC7102 Rev.F now), 6/7 of the findings are directly related to pyrometry and in my experience it is usually a test certification/record issue (not always).
6. You need to know the use of different types of thermocouples (ie: K, E, J, N, S, R, B and more) and permissible usage as required by AMS2750E.
7. Be aware of additional customer pyrometry requirements. Boeing and Lockheed are examples of primes that have more stringent requirements than AMS2750E. So often calibration companies will certify a SAT or a TUS to BAC5617 and unfortunately their paperwork (and their service) does not meet the Boeing requirements (thermocouples is one of those areas).
8. Hire an experienced consultant to perform an internal audit for you to AC7102 or have them do a gap assessment to see where you are at.
9. Don't forget you need calibration stickers!
10. Ensure you have a good understanding of AMS2750 and the Nadcap checklist along with any customer requirements. My professional experience is in Quality and Operations Management in the following areas: ISO 9001, AS9100, ISO 17025, Nadcap Heat Treat, Chemical Processing, Materials Testing, Welding, Conventional and Non-Conventional Machining. With that being said, I find the most challenging area with a lot of companies is AMS2750E and Nadcap Heat Treating.

This doesn't cover everything you need but should start you on the right track. You will need a good QA person to review all internally generated test certifications (hint - correction factors and offsets!).
 
Last edited by a moderator:
Elsmar Forum Sponsor
B

BizPro

#22
Shane and Others -

Type N is a good base metal T/C even though it cost more. Per AMS2750 -Type N may be calibrated as well in certain circumstances (expandable vs. non-expandable). Type K and E are never allowed to be re-calibrated per AMS 2750. They may be re-used IF and ONLY IF it is under 500F and you have OE (objective evidence) that you have implemented the U-Formula and follow it. This is per AMS 2750, BAC5617 has more stringent T/C requirements.
 

TrivialSublime

Involved In Discussions
#23
Just a helpful correction here. BAC 5621 is Boeings specification for Temperature Control of Processing of Materials (i.e. their Pyrometry spec). BAC 5617 looks to be a heat treat spec. I am sure it mentions BAC 5621 in applicable documents but that?s kind of a round about way of getting there.

Also if Boeing legacy product (Douglas) is going to be heat treated then that calls in DPS 1.700, Douglas' Pyrometry spec.

My recommendation on recorders would be digital to eliminate the whole paper chart thing (pens, paper, etc.) and have them linked into your internal internet server for storage and easy access to records. Disc space is cheap. In the long run the expense and time spent will save you money.

Honeywell makes a good product called EZtrend that comes with both controller and recorder built in with multiple channels.

Omega makes good handhelds that can be used for calibration and SAT. We use an SSI datalogger still going strong after 6 years.

My own opinion about the K and N thermocouple changes from AMS 2750 Rev C and Rev D. The rumor I heard was Boeing has issues with the K thermocouples (T/C) having reliability problems. When the spec rev?d to D we went to N and use Ks as control and record T/C and discard those after a year without recalibrating which is not allowed as you state.

The Ks hold up quite well and it?s a shame to have to scrap them but it makes life easier.

Just my two cents on that.
 
H

hackyue

#24
Re: Calibration Plus TUS in accordance with AMS 2750 rev E

Hi,there Yes we are still Nadcap approved ,yes it is customer requiments,
we need to buy new equipment To do Temperature Uniformity Survey(TUS)as I this moment company from outside do this for us,we are looking for savings like I said,My question is Can WE DO IT INHOUSE
I don't think there is any problem about doing all the test in house.

All of our customers accept it.

And it saves much money, and the efficiency is better
 
M

Mtvated

#25
My past and current employers have perform SAT and TUS in-house. It would be more flexible to schedule calibrations around production and very cost effective.
 
S

sol49720

#26
You could rent your needed Fluke meter instead of purchasing one, analyse the use cost over time against purchase cost to determine what is best for your individual circumstance. Don't forget to pull your data from the furnace sensors from the furnace recorder, not from the data logger or you will not meet AMS 2750E 3.5.13.3.2. You are using the data logger to verify your furnace recording is correct and within calibration.
 

dgriffith

Quite Involved in Discussions
#27
... Don't forget to pull your data from the furnace sensors from the furnace recorder, not from the data logger or you will not meet AMS 2750E 3.5.13.3.2. You are using the data logger to verify your furnace recording is correct and within calibration.
...not from the data logger... This must be a misprint. Why have a reference data logger to check the furnace data recorder if your not going to use the reference data. :D Likely you meant pull the data from both?
 
M

MTProcessing

#28
I think he's referring to the requirement to capture data from the control thermocouple during a survey in addition to the data from the survey thermocouples. The survey thermocouples have to go through a separate datalogger. Some setups allow the survey logger to also capture the data from the control thermocouple, but this does not meet the requirement. The control thermocouple data must be captured through the normal control instrument.

Just one of the 2750 items that can make a person's head spin.
 
R

robanbieber

#30
Hi bwdreiko,
Actually the CL25 sure could be used as a secondary standard. You would have to have it calibrated to a primary standard if I remember correctly.
We had a NADCAP auditor ask us why we weren't doing that for our data logger and we strictly don't because I like having that unit being calibrated by the manufacturer.
That being said, you can use (double check AMS2750E, don't think that?s changed in regards to this) , we also use the CL25 for furnace Instrument calibrations.
I do not think the Omega CL25 could be used as a secondary standard instrument for calibrating your TUS dataloggers or sensors.. as ams2750E states that the required calibration accuracy is +/- 0.3F (+/-0.2C). The omega CL25 only has accuracy of +/- 0.5F... see attached three sheets for omega CL25 specs plus ams requirement table... am I missing something?
The CL25 could be used to calibrate the intruments on the oven or be used in the TUS or SAT but it could not be used to calibrate the TUS loggers or the sensors.... I am looking for an instrument that can meet those AMS specs that does not cost a fortune.. anybody have any suggestions?
 

Attachments

Last edited by a moderator:
Thread starter Similar threads Forum Replies Date
M Inhouse Calibration of Regulated Power Supplies (RPS) using calibrated Multimeter General Measurement Device and Calibration Topics 7
T Inhouse CMM Verification using Artifact (metronom/tetronom) vs. Outsourcing General Measurement Device and Calibration Topics 11
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 0
N IATF Calibration Lean in Manufacturing and Service Industries 5
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 26
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W Next Calibration Due Date Calibration Frequency (Interval) 5
S Where do l get calibration standards to run a calibration lab? Other ISO and International Standards and European Regulations 2
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 1
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
Crimpshrine13 Calibration of pH Meter Probe Calibration and Metrology Software and Hardware 3
F ESD workbench "calibration" Manufacturing and Related Processes 2
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
F Nist traceable calibration certificates General Measurement Device and Calibration Topics 1
G Tool tracebility and First calibration requirements for aerospace (AS9100) organisation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Calibration of Digital thermometer with surface probe General Measurement Device and Calibration Topics 1
T Calibration or Verification -> Cm and Cmk, etc. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
G Is repeatability required for equipment calibration? General Measurement Device and Calibration Topics 10
was named killer AS 9100D - Calibration Instructions - Controlled documents? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
B Photovoltaic Tester Calibration General Measurement Device and Calibration Topics 0
B Temperature loop: partial calibration General Measurement Device and Calibration Topics 10
G Glass scale calibration at 74 degrees F? General Measurement Device and Calibration Topics 4
A Digital Timer Calibration Requirements General Measurement Device and Calibration Topics 3
Ron Rompen Barcode Verifier Calibration and Use Calibration and Metrology Software and Hardware 5
S Is Optical Parallel Set required calibration? ISO 17025 related Discussions 1
A Calibration of Power Supplies AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
D Calibration Process Flow Map Example Wanted General Measurement Device and Calibration Topics 3
M Calibration or Verification? What terminology to use ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1

Similar threads

Top Bottom