Installation Qualification (IQ) and Operational Qualification (OQ) for Laser Marking
J.Enger
We are being asked by one of our customers do do an IQ/OQ (Installation Qualification/Operational Qualification) for laser marking for medical implants/tools. I have done pretty much everything, except the Functional Tests for the OQ part. The only thing I can come up with is the Focus operation, but that is something that is done at every setup. I have sent an email to the company for help on that part as well,but any help from you guys on here would be appreciated.
ScottK
Not out of the crisis
Re: Installation Qualification (IQ) and Operational Qualification (OQ) for Laser Mark
I was with a company that did this for the same reason, same type of customer. I'm pulling this from memory as I am no longer with that company...
It kind of depends on your machine.
Definitely consider the focus. Can you adjust the power? If so that would be a consideration.
We also looked at the print range and did a worst case by printing something in the corners of the largest size piece the machine can handle.
And varying font sizes as well, we went to the smallest legible text size.
And material - we did several materials so we ran each test for each material.
Are you printing bar codes for these customers? If so that's a whole other level and you'll need some sort of calibrated bar code verifier/grader to check the bar codes you print... I would make that part of set-up QC check in normal operation too.
J.Enger
Thanks for the info. I was able to generate the IQ/OQ to their acceptance.
If they do have us do the marking, I have several quotes for different Bar Code Verifier/Qualifiers to purchase, which we will split the costs over the different parts they want to have us mark.
S Barcode Scanner IOPQ (Installation/Operational/Performance Qualification) Requirement Qualification and Validation (including 21 CFR Part 11) 9
W Packaging / Heat Sealer Validation - IOQ (Installation/Operational Qualification) Qualification and Validation (including 21 CFR Part 11) 15
W IQ, OQ, PQ (installation, operational & performance qualification) compared to PPAP ISO 13485:2016 - Medical Device Quality Management Systems 5
H Basic questions about Installation, Operational & Performance Qualification - PCBA Qualification and Validation (including 21 CFR Part 11) 11
C ISO 13485 and IQ (Installation Qualification), OQ (Operational Qualification) ISO 13485:2016 - Medical Device Quality Management Systems 30
C Generate IQ, OQ & PQ (installation/operational/performance qualification) Report Document Control Systems, Procedures, Forms and Templates 11
J FDA Microscope IQ (Installation Qualification) Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
JoCam Rental, service & installation - regulatory requirements EU Medical Device Regulations 2
S Quality management system (Well head installation & maintenance) Quality Management System (QMS) Manuals 5
I Veeva installation and the Loader Tool Document Control Systems, Procedures, Forms and Templates 3
GURU SDI Leak Testing - O-ring installation in the groove of Bore ID Problem Solving, Root Cause Fault and Failure Analysis 4
A Best Method of Auditing an Installation Process at a client job-site ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
S Do you test the download and installation of a mobile medical device app? Software Quality Assurance 6
Q UDI and Standalone Software - UDI per installation? Other Medical Device Related Standards 6
A What is the definition of Installation that requires Record Keeping vs Set-up 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
L Medical Device Installation Records from Third Party Installer 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S ISO 13485 Installation Activities 7.5.1.2.2 ISO 13485:2016 - Medical Device Quality Management Systems 4
G Nonconformity Reporting at Installation Sites Service Industry Specific Topics 8
K Supplier or outsourced? Medical Device installation, sales, training Other ISO and International Standards and European Regulations 7
N Definition of Large-Scale Fixed Installation per the RoHS2 Directive EU Medical Device Regulations 3
E IEC 62304 5.2.2 f - Installation and acceptance of device software IEC 62304 - Medical Device Software Life Cycle Processes 1
S Time & Motion Study - Dish TV Installation Quality Tools, Improvement and Analysis 4
J ISO 9001 Audit Questions relating to Installation of Lifts ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
T Office Relocation Installation Inspection Sheet Document Control Systems, Procedures, Forms and Templates 5
S ETSO, TSO and EASA Form 1 - Do I need the ETSO approval for the installation? EASA and JAA Aviation Standards and Requirements 3
S SW (Software) - Customer vs. Company Installation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
T Site Installation Inspection and Testing for a HVAC Contractor Service Industry Specific Topics 2
A Installation and Validation Procedure for IT Laboratory ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
K Starrett RGC4040-24 CMM User and Installation Manual General Measurement Device and Calibration Topics 2
R Is there a Specification for safety wire installation - Aerospace, SAE or otherwise AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
D Tips/Documentation samples required for X-ray Film Processor Installation needed. ISO 13485:2016 - Medical Device Quality Management Systems 1
T Installation Requirements for Bench Top Device ISO 13485:2016 - Medical Device Quality Management Systems 1
MarilynJ6354 Service & Installation Parts Identification - Install equipment at a customer's site Service Industry Specific Topics 6
C Installation Document that meet NFPA 70 & 72 Various Other Specifications, Standards, and related Requirements 3
C NFPA 72 and NFPA 72 Standards - Electrical Installation and Fire Alarm Systems Various Other Specifications, Standards, and related Requirements 3
Q IQ, OQ & PQ (Installation, Operation, and Performance Qualifications) on a CMM ISO 13485:2016 - Medical Device Quality Management Systems 2
Q Software installation issues After Work and Weekend Discussion Topics 3
I Medical Gases Installation - Can be classified as MD and ISO 13485 certificated? ISO 13485:2016 - Medical Device Quality Management Systems 5
S Quality Standards on Refrigeration (Installation) field IATF 16949 - Automotive Quality Systems Standard 3
C Auditing of Installation of Product at Customers Sites General Auditing Discussions 6
M Seeking: Good process definition (map or whatever) for an installation business ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
Q How to Calibrate Software - Microcontroller software installation General Measurement Device and Calibration Topics 1
F 4.9 Installation & Servicing Processes QS-9000 - American Automotive Manufacturers Standard 1
F Component maintenance qualification - Regulation (EU) No 1321/2014 Article 5 EASA and JAA Aviation Standards and Requirements 6
I IATF Lab Scope Testing Qualification and Competency Documentation IATF 16949 - Automotive Quality Systems Standard 3
R Qualification list of businesses approved to Off-Shore Aerospace work. AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
S Qualification question - ISO 13485 - Setting up a small lab Reliability Analysis - Predictions, Testing and Standards 2
E Qualification for non gmp service providers Supplier Quality Assurance and other Supplier Issues 1
V Process, component or full product re-qualification: leaded to unleaded solder Qualification and Validation (including 21 CFR Part 11) 8
DuncanGibbons Understanding the applicability of Design of Experiments to the IQ OQ PQ qualification approach Qualification and Validation (including 21 CFR Part 11) 5