Insulation class II device with metal enclosure
EMengineer
Hi everyone. We have device which is powered by external PSU. As most PSU's its output does not have protective earth connection. In our case we have metal enclosure with secondary circuit inside it (voltage up to 24VDC).
Shall the metal enclosure be separated from all circuits by 2 MOP (J.2 insulation example) - no gorunding with enclosure? SIP/SOP is double insulated (for mains voltage). Although SIP/SOP connector may touch the enclosure (in this case it should be plastic or separated by 2 MOP distance?).
Does the separation between secondary and enclosure also would ensure that there is no risk of operator touching the enclosure and the patient at the same time which would deliver too high leakage current?
In this case PSU would be insulation class I and device insulation class II?
Maybe there are points that I missed?
Peter Selvey
Insulation Class (i.e. Class I, Class II) is purely mains related and specifically the connection to the mains. It lets the user know if they need a two pin plug or three pinned plug. No matter what the insulation strategy is for various parts of the equipment, the equipment remains Class I or Class II (one or the other) in principle based on the mains connection (there are some weird exceptions for functional earths ... but that is another story).
After the mains connection, different strategies can be used as long as they are effective. For example you could bond (deliberately connect) all the metal parts of the enclosure together such that it is not possible for the 24V to get out like a faraday cage. Or at least, bond any metal parts that are near secondary circuits. This is kind of like earthing and probably the most common way. The bonding needs to be good enough that it can carry the maximum current from the 24V supply without breaking. For the sake of avoiding confusion, you could also connect the 0V line of the 24V circuit to the metal parts as well, that makes it clear especially if you are dealing with test labs.
If the metal enclosure can contact the patient, there could be a lot more to the story. It will depend on the classification (B, BF CF etc), parts of the equipment that contact and so on. It is not simple to explain.
The SIP/SOP can be excluded under 8.4.2c or can be designed in such a way that there is insulation around the part.
EMengineer
Thank you Peter for the response. So insulation class is a PSU matter. If device is separate from the PSU it does not have any insulation class.
To add to your answer I found other thread with your response about Annex J insulation examples:
Isolation of secondary grounded circuits
To sum up - no need to worry about low voltage secondary circuit within unearthed metal enclosure (if mains part is in external PSU). The most important thing is to pay attention to the correct separation of the applied part (2 MOPP for working voltage and 1 MOPP for mains voltage).
Although I have a little fear how testing house will evaluate this part (I saw in other threads that manufacturer had to put Y caps to comply with annex J example, even when it is only informative annex).
EMengineer
I couldn't find any specific information in the standard regarding the scenario where an unearthed metal enclosure becomes accidentally earthed through contact with other equipment, it's possible that I may have missed something. Could you please clarify if there are any specific considerations in the IEC 60601-1 standard for this situation? The earthing probability seems to be high, but the applied part remains unaffected due to proper separation. In this particular scenario where mains voltage is present on unearthed metal parts (Fig 18 of 601-1), even though unintentional earthing can occur at times, how should the evaluation of leakage current be conducted?
I have another question regarding leakage currents, specifically in a device with a CF applied part. If the operator can simultaneously touch both the metal controller and the patient, should the controller be considered as a part to which the requirements for the applied part (in this case, CF) are applicable (as per 601-1 cl. 4.6)? Then the controller shall be insulated from the other circuits by 2 MOPP, be of insulated material and pass 10 uA leakage current requirement?
K Doubt on insulation class of a device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Insulation from MAINS of the PE of a CLASS I PS used as FE IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
O Insulation Diagram IEC60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
F Grounded (PE) Cable foil as Means of Patient Protection and solid insulation levels for secondary circuits IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
H Regarding Double Insulation & Caution Symbol on Product according to IEC 61010-1 CE Marking (Conformité Européene) / CB Scheme 6
X Difference between the Figure J.6 and J.7 insulation examples IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
X Insulation between opposite terminals of battery IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
W Insulation requirements of HV pulse circuits IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
G 60601-1 8.8.3. short circuit of both side during insulation test IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
V Insulation Diagram for PB560 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Aluminum anodization as MOP insulation? Clause 8.5.1.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Insulation diagram for ECG monitoring and diagnosis system. IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
W Insulation requirements of BF type infant skin temperature probe IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M 2xMOPP insulation for Applied Part B. IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
Q Isolation/Insulation Diagrams - IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
V PCB internal insulation treated as insulation formed by insulating compound IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W HF surgery device accessory insulation requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
E Creepage distance accross insulation of step up transformer EU Medical Device Regulations 2
S Who compiles the Critical Component List, Insulation Diagram etc.? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P Question on Table 6 "Test voltages for solid insulation forming a MEANS OF PROTECTION IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P MOOP and insulation coordination IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K IEC60601 3rd Edition Insulation Thickness for electrosurgical devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N Cable Connectors - Insulation does not go all the way into the back of the connector Manufacturing and Related Processes 4
A Insulation between Secondary Accessible Parts and Earth IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P IEC 60601-1-11 - Conversion to Double Insulation IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Is insulation between F type applied part and other parts a must? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
N Insulation Diagram for Water Tub IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W Is anybody familiar with IEC61010-1 Insulation Requirements? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
V External power supply with reinforced insulation as a simple secondary circuit? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W Insulation requirements between F-type applied part and enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
P Understanding IEC 60601-1:2005, Clause 8.1 b) - Short Circuiting Double Insulation IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
R Determining Working Voltage - Drawing a Medical Device Insulation Diagram IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
W Insulation Grade for In Vitro Diagnostic (IVD) Device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Do insulation layers in double insulation have sequence requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S Can enclosure of semiconductor or electrolytic capacitor provides basic insulation? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
N Insulation Barrier Testing - According IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 22
T Insulation diagram for a medical electrical equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
H Insulation of FRP (fiberglass reinforced plastic) Tanks Manufacturing and Related Processes 3
B Control Chart - Minimum point & average insulation thickness for Automotive Wires Statistical Analysis Tools, Techniques and SPC 9
A Medical device with NON medical PC - USB galvanic insulation ISO 13485:2016 - Medical Device Quality Management Systems 2
P EN 62304 Safety Class of an ECG device designed to be used in clinical situations when patient is in immediate danger. IEC 62304 - Medical Device Software Life Cycle Processes 1
S Clean Room Class ISO 8 EU Medical Device Regulations 3
adztesla Design change and Process change Class 3 medical device Medical Device and FDA Regulations and Standards News 2
C CMO for class 2, 510k exempt device Manufacturing and Related Processes 4
E Class II symbol - where to apply it? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Class Ir device have to be certified under MDR ? EU Medical Device Regulations 2
B Under MDR, are VATS and MICS still considered just class Ir? ISO 13485:2016 - Medical Device Quality Management Systems 1
Leigh76 CE Marking of Class II Bulk NonSterile Products for Kit Packers EU Medical Device Regulations 1
E Class II ME EQUIPMENT and subclause 8.9.1.12 IEC 60601 - Medical Electrical Equipment Safety Standards Series 10