EMengineer
Involved In Discussions
Hi everyone. We have device which is powered by external PSU. As most PSU's its output does not have protective earth connection. In our case we have metal enclosure with secondary circuit inside it (voltage up to 24VDC).
Shall the metal enclosure be separated from all circuits by 2 MOP (J.2 insulation example) - no gorunding with enclosure? SIP/SOP is double insulated (for mains voltage). Although SIP/SOP connector may touch the enclosure (in this case it should be plastic or separated by 2 MOP distance?).
Does the separation between secondary and enclosure also would ensure that there is no risk of operator touching the enclosure and the patient at the same time which would deliver too high leakage current?
In this case PSU would be insulation class I and device insulation class II?
Maybe there are points that I missed?
Shall the metal enclosure be separated from all circuits by 2 MOP (J.2 insulation example) - no gorunding with enclosure? SIP/SOP is double insulated (for mains voltage). Although SIP/SOP connector may touch the enclosure (in this case it should be plastic or separated by 2 MOP distance?).
Does the separation between secondary and enclosure also would ensure that there is no risk of operator touching the enclosure and the patient at the same time which would deliver too high leakage current?
In this case PSU would be insulation class I and device insulation class II?
Maybe there are points that I missed?