Insulation requirements of BF type infant skin temperature probe
william guo
Starting to get Involved
#1
Hello experts,
The skin temperature probe of infant incubator was classfied to Type BF applied parts, requirements of 1MOPP based on mains voltage shall be met. Question is where the isolation I can put for this 1MOPP?
1. the insulation of temperature probe. We tested the electric strength btween input conectors pins and probe head with 1500Vrms/1min, no breakdown. but the problem is how to locate the inuslation on the probe, where is it?
2. if we considerred no insulation exists on the probe, can we put isolation on the signal boad by using isolation compoments such as optocoupler, this can complicate the signal processing and circuits design.
Does anyone has experience about this?
Elsmar Forum Sponsor
Thread starter Similar threads Forum Replies Date
W Insulation requirements between F-type applied part and enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W HF surgery device accessory insulation requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W Is anybody familiar with IEC61010-1 Insulation Requirements? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Insulation diagram for ECG monitoring and diagnosis system. IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K Doubt on insulation class of a device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M 2xMOPP insulation for Applied Part B. IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
Q Isolation/Insulation Diagrams - IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
V PCB internal insulation treated as insulation formed by insulating compound IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Creepage distance accross insulation of step up transformer EU Medical Device Regulations 2
S Who compiles the Critical Component List, Insulation Diagram etc.? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P Question on Table 6 "Test voltages for solid insulation forming a MEANS OF PROTECTION IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P MOOP and insulation coordination IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K IEC60601 3rd Edition Insulation Thickness for electrosurgical devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N Cable Connectors - Insulation does not go all the way into the back of the connector Manufacturing and Related Processes 4
A Insulation between Secondary Accessible Parts and Earth IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P IEC 60601-1-11 - Conversion to Double Insulation IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Is insulation between F type applied part and other parts a must? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
N Insulation Diagram for Water Tub IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
V External power supply with reinforced insulation as a simple secondary circuit? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P Understanding IEC 60601-1:2005, Clause 8.1 b) - Short Circuiting Double Insulation IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
R Determining Working Voltage - Drawing a Medical Device Insulation Diagram IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
W Insulation Grade for In Vitro Diagnostic (IVD) Device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Do insulation layers in double insulation have sequence requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S Can enclosure of semiconductor or electrolytic capacitor provides basic insulation? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
N Insulation Barrier Testing - According IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 22
T Insulation diagram for a medical electrical equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
H Insulation of FRP (fiberglass reinforced plastic) Tanks Manufacturing and Related Processes 3
B Control Chart - Minimum point & average insulation thickness for Automotive Wires Statistical Analysis Tools, Techniques and SPC 9
A Medical device with NON medical PC - USB galvanic insulation ISO 13485:2016 - Medical Device Quality Management Systems 2
P UDI-PI requirements on reusable surgical device, do we need serialisation? ISO 13485:2016 - Medical Device Quality Management Systems 0
Sidney Vianna Interesting Discussion Should ISO 9004 be changed from a guidance standard to a requirements standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
J WAIVED ON Q1 - We Don't have to comply with FORDS customer specific requirements IATF 16949 - Automotive Quality Systems Standard 2
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
Techilady Ford Special Characteristics Communication and Agreement Form (SCCAF) provides production requirements? FMEA and Control Plans 4
P Requirements for being an European Representative EU Medical Device Regulations 3
C ISO 14001:2015 6.1.3 Compliance Obligations - Legal requirements monitoring ISO 14001:2015 Specific Discussions 0
lanley liao How to understand correctly the requirements of API monogram products Oil and Gas Industry Standards and Regulations 0
A GPS Regulatory requirements? Manufacturing and Related Processes 6
J SSQR-01 Vision Requirements- Color Blindness AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
P ISO 14644 Class 8 Cleanroom Air Filter Requirements Other Medical Device Related Standards 4
J Requirements as a Distributor for Incoming Inspection of Purchased Finished Medical Device Medical Device Related Regulations 0
R Weld Penetration Standards and Requirements Manufacturing and Related Processes 2
S Registration of Medical Device in Hong Kong - labeling requirements Other Medical Device Regulations World-Wide 0
L Requirements for Cables used for Low-Voltage DC Mains (12-36VDC) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M AS9100D Registrar pre-audit requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 15
S Determining the requirements for the products and services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
D LiPo battery minimum requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
V MDSAP Requirements for Device Designer Canada Medical Device Regulations 1
Similar threads
Top Bottom