Internal Calibration Process Suggestions

AmandaMusser

Quality Rules!
#1
Hello Covers!
I am in need of some suggestions regarding our internal calibration process.
We are an AS9100, ISO 9001, ISO 13485 & FDA certified/registered contract manufacturer of circuit board assemblies. For our calibration process, the majority of our eqpt is sent to an outside vendor for calibration. However, some eqpt is calibrated in-house by our Maintenance Technician.

1. We received the following finding from our Internal Audit "In review of the training record database for Vanguards Calibration Technician, Maint Tech has no record of any basic training associated with ISO 17025 or other calibration standards. Technical training for the oven calibration and pick/place equipment was listed as ?On the Job Training? only. Calibration records and certificates signed off by Mr. Tech cannot be accepted as authentic calibrations conducted by a calibration technician". In order to meet requirements, does our Maintenance Tech need to have training/certification in metrology or does OTJ experience meet this?

2. Calibration/validation/verification... our 1 tech is getting buried with all the eqpt that he calibrates, validates, verifies &/or performs PM on. Is it acceptable for him to train eqpt users to perform the validations &/or verifications and if a calibration is required, it is moved to him & he performs that part?

Thank you in advance!
 
Elsmar Forum Sponsor

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#2
Re: Internal Calibration Process - Suggestions needed

I note you do not mention being accredited. However, the other standard most likely require that personnel be appropriately trained. In such instance, the internal audit finding is valid on its surface. At the very least, I would suggest acquiring a copy of The Metrology Handbook, available from ASQ. Then look to somewhere such as Workplace Training for distance learning in Metrology theory and basics.

Verification between calibrations can be done by users, but calibration truly should be by a trained and experienced professional. Two things to make sure of. The traceability must be documented establishing the link to your measurement standards. Additionally, uncertainty must be estimated. The metrological definition of traceability requires both of these, for traceability. Watch the so-called NIST number as generally it is NOT valid for traceability, as it is a work order number.

Training in ISO/IEC 17025 is desirable but for internal cal may not be needed. However, you should have the "Trilogy" documents there: ISO/IEC 17025, Guide to the Expression of Uncertainty in Measurement (GUM), and the international vocabulary of metrology (VIM) and you want the most recent edition. I think it is now designated JLDG200 but I don't have my copy in front of me.

Hope this helps.
 
Q

QualityEMW

#3
Re: Internal Calibration Process - Suggestions needed

we have had the same issue arise here, the solution for us was a documented procedure for calibrating an item with calibrated (by a lab) referenced in the Calibration record.
 

Jen Kirley

Quality and Auditing Expert
Staff member
Admin
#4
Re: Internal Calibration Process - Suggestions needed

I agree it is important to have these procedures, but I also favor the Handbook that Hershal recommends because it talks to technique, which is critically important to achieve consistent calibration results.
 

somashekar

Staff member
Super Moderator
#5
Re: Internal Calibration Process - Suggestions needed

I am not sure about the AS specific requirement, however following is important.
1. The documented procedure
2. The basis used for the calibration or verification.

Whoever is responsible for the in-house calibration must be competent by way of education / skill / experience which can be evident in the determined competency / training record.
There is no requirement of ISO 17025 or other calibration standards. This for me is just the auditor's own interpretation.
The need here is about the competency in knowing and performing per calibration procedures from the manufacturer of those specific equipment OR Basis used for verification, and availability of necessary infrastructure.
 
Last edited:

AmandaMusser

Quality Rules!
#6
Re: Internal Calibration Process - Suggestions needed

Thanks everyone!
I have now scheduled training for our technician. Better to be safe than sorry :)

I'm now looking at re-vamping our process to allow users to validate/verify equipment based on work instructions created by the technician. If needed, they would submit equipment to the technician for adjustments/calibration.
 

Mikishots

Trusted Information Resource
#7
Re: Internal Calibration Process - Suggestions needed

I am not sure about the AS specific requirement, however following is important.
1. The documented procedure
2. The basis used for the calibration or verification.

Whoever is responsible for the in-house calibration must be competent by way of education / skill / experience which can be evident in the determined competency / training record.
There is no requirement of ISO 17025 or other calibration standards. This for me is just the auditor's own interpretation.
The need here is about the competency in knowing and performing per calibration procedures from the manufacturer of those specific equipment OR Basis used for verification, and availability of necessary infrastructure.

I can help with the AS standard, if it helps others here. In additional to ISO 9001:2008, it requires:

1. A register of M&M equipment and a defined process for calibrating/verifying, including details of equipment type, unique ID, location, frequency of checks, check method and pass/fail criteria.

2. Environmental conditions must be suitable for the calibration, inspection, measurement and testing you do.

3. There must be a method (established, implemented, maintained) of notifying people to send in the devices that are due for calibration or verification (or at a minimum, a method of recall, irrespective of how you do it).

There is also a note added, which states that M&M equipment includes (but is not limited to) test hardware, test software, automated test equipment and plotters used to create inspection data. It also includes personally owned/customer supplied M&M equipment, if present.
 
Thread starter Similar threads Forum Replies Date
M How to Internal Audit the Calibration Process Internal Auditing 12
B IATF 16949 - 7.1.5.3.1 Internal calibration laboratory requirements IATF 16949 - Automotive Quality Systems Standard 9
C MU (Measurement Uncertainty) in an Internal Calibration Laboratory Measurement Uncertainty (MU) 1
I Calibration Void Seals on Internal Equipment General Measurement Device and Calibration Topics 3
S Internal Calibration Laboratory Inspection/Approval General Measurement Device and Calibration Topics 3
S Auditing TS 16949 Cl. 7.6.3.1 - Internal Calibration Laboratory Requirements IATF 16949 - Automotive Quality Systems Standard 8
C Basic Internal Calibration Procedure to check the Accuracy of a Balance Calibration Frequency (Interval) 6
N Internal Calibration vs. Accredited Calibration for Balance - Differences Calibration Frequency (Interval) 3
P Internal Calibration for Laser Temperature Gauges Manufacturing and Related Processes 3
N Scale with Internal Calibration - Do I need to calibrate? How to document this? Calibration Frequency (Interval) 8
L Calibration Procedure for Internal Spline Plug Gage Needed General Measurement Device and Calibration Topics 4
2 TS 16949 Cl. 7.6.3.1 - Internal Calibration Procedure not Traceable to ASTM, EN etc. IATF 16949 - Automotive Quality Systems Standard 20
H "Internal Calibration" using other Calibrated Equipment Calibration Frequency (Interval) 13
K Internal laboratory - applies only to Calibration or also to other laboratories? IATF 16949 - Automotive Quality Systems Standard 6
R Standard Methods - Internal Calibration Laboratory - What if there is no standard? IATF 16949 - Automotive Quality Systems Standard 2
P Internal Calibration or Verification - Personnel Requirements IATF 16949 - Automotive Quality Systems Standard 3
M Internal tool calibration program - Best source to become "knowledgeable"? General Measurement Device and Calibration Topics 5
V Internal Scale and Balance Calibration Method Tolerances and Limits General Measurement Device and Calibration Topics 4
D ISO/TS16949 Internal Calibration Lab need to include Production Equipments IATF 16949 - Automotive Quality Systems Standard 26
C Internal Diameter Measurement and Pin Gage Calibration General Measurement Device and Calibration Topics 6
K Internal Calibration - Testing Laboratory General Measurement Device and Calibration Topics 9
Q Internal Testing and Calibration Laboratories - MSA, Gage R&R and Control Plans IATF 16949 - Automotive Quality Systems Standard 2
M Company Internal Calibration - Is Internal Calibration a LABORATORY? TS16949 General Measurement Device and Calibration Topics 24
C Internal Audit of Internal Calibration Laboratory - Inspection sample size question Internal Auditing 2
D Internal Laboratory - Defining the Laboratory Scope - Inspection, Test, Calibration Inspection, Prints (Drawings), Testing, Sampling and Related Topics 18
L Internal Calibration Laboratory Requirements - Ambient temperature, Humidity, Etc. Statistical Analysis Tools, Techniques and SPC 3
K Should QS/TS certified internal test/calibration labs be required to be ISO17025 ISO 17025 related Discussions 16
V Does an internal calibration laboratory need to be certified on ISO 17025? ISO 17025 related Discussions 3
E MDR internal audit Medical Device and FDA Regulations and Standards News 0
B Internal Auditor Competency - Product Auditors Internal Auditing 9
U Internal Auditor not trained but done Audit for some process Nonconformance and Corrective Action 5
F Internal transfer of work from one line to another? Qualification and Validation (including 21 CFR Part 11) 3
B Looking for 10 Internal Audit Online Training Participants ISO 17025 related Discussions 2
R Monitor production quality - Internal KPIs Manufacturing and Related Processes 5
R IATF 16949 - Outsourcing of internal audits Internal Auditing 10
M Major vs. Minor for Internal Audits? Internal Auditing 10
T Internal Nonconformance procedure thoughts (AS9100) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
M Tips on preparing for IATF 16949 Internal Lead Auditor exam Manufacturing and Related Processes 1
C Internal Audits in a tiny Dx Company Internal Auditing 33
H AS9100 Checklist for Internal Audit needed AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
A What are the pros and cons of using an audit software for internal auditing? General Auditing Discussions 7
F Internal Audit before Pre-Assessment ISO 17025 related Discussions 2
Q Internal audit plan template Internal Auditing 6
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
L Internal audit during COVID-19 restrictions ISO 13485:2016 - Medical Device Quality Management Systems 5
A QMS Overhaul - Reorganizing and Renaming Documents - Internal References Document Control Systems, Procedures, Forms and Templates 18
N Sampling Plan for Internal Audits - ISO 2859 or 3951 - Or Neither? Internal Auditing 6
O ISO13485 implementation - Are internal audits expected before stage 1 audit? Design and Development of Products and Processes 3
B Using Unreleased Documents & Process Maps for Internal Audit purposes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
M Help me improve the definition for internal PPM Manufacturing and Related Processes 3

Similar threads