"Internal Calibration" using other Calibrated Equipment

H

Haim Moshe

#1
Hello Everyone!
I am new to this forum but had a the chance to sniff around lately.

We have a model shop in company which handles only R&D requests.
it is a wonderfull workshop with top edge equipment and a lot of measuring equipment

Until lately they never calibrated the TE saying it was for R&D purposes and as such it is not required.

Eventualy they understood the need for calibration and so we are now calibrating the majority of the stuff.

However; there are some TEs for which we are doing "Internal Calibrating" using other calibrated equipment.

we are keeping the reference equipment in a closet and the room is being air conditioned almost 24/7 (with measurments indicating RT and 50%RH)

My question is - it is allowed?

Haim
 
Last edited by a moderator:
Elsmar Forum Sponsor
S

SteveK

#2
Re: "Internal Calibrating"

Hi Haim,

Welcome on board!

In answer to your question - I certainly hope so, this is what we do. We use calibrated 'Master' gauges to calibrate gauges on the shop floor - these are locked away. Not had any problems with this during audits. It is traceability that is the ultimate requirement IMO.

Steve
 
H

Haim Moshe

#3
Re: "Internal Calibrating"

Thanks Steve

I felt it is OK but wanted something to confirm this
I reviewed the ISO9000 req. but it is worthless in this matter
ISO17025 is an killer

How often you are calibrating your masters? annually?

Haim
 

BradM

Staff member
Admin
#4
Hello Haim!:bigwave:

I do have some questions here in regards to your question.

As Steve said, "yes" it's OK to perform calibrations in in-house. Apart from that, you might want to answer some questions to assure you have an adequate process to do them.

  1. Do you have a valid procedure to do the work?
  2. Are the operators trained on the procedure and have the knowledge to do the work?
  3. You state you have environmental controls. Are they sufficient to assure there is no impact/effect from the environment on the calibrations and/or standards?
  4. Have you established some plan of remedial action if the standard is found out of tolerance? This may also involve risk assessment and such.
  5. Are all the results being adequately documented and reviewed?
  6. Are your standards of sufficient accuracy?
I know that's a lot of questions.:D I was just pulling some out of the air. There may be more.

You might consider getting ISO10012 and assuring your process meets those requirements:
http://www.iso.org/iso/catalogue_detail?csnumber=26033

17025 is a fairly stringent requirement, but is a great one to follow. You might consider reading it and verifying that your process satisfies its requirements.

Hope this helps.:)
 
S

SteveK

#5
Hi again Haim,

All our master gauges are calibrated annually by a calibration/test house. Obviously we retain the certificates for the masters. The other things Brad asks for is spot on. After an audit I had a minor NC because I did not state what we would do in our procedure if a gauge came back and it was found out of calibration.

Steve
 
D

Duke Okes

#6
Of course you need to ensure that your internal calibration devices have the appropriate accuracy/precision in order to accomplish the required accuracy/precision of the final device.
 
G

George Weiss

#7
The original question of calibration did not reference an ISO STD, but 9001 is assumed.
There is not much help there, because you are in the ISO realm of "write what you do", and then "do as you write". There are some steps in the completeness and correctness of the process. Your comment about ISO-17025:2005 being an unwritten OMG is understandable. There is another group of guidance documents, which are old, but put you on a good course. MIL STD 45662, costs $0, and is easily reviewed, and puts out a good collection of methodologies. Referring to this standard might be a good first step in a circling of what to do. One of the draw-backs to doing a better job, and referencing more documents, is a mushrooming of technicalities. An evaluation of your over-all calibration plan level being: 1) a few calipers, 2) some small meters and power supplies, or 3) a developing variety of equipment and tools. Making a simple procedure and not involving all the MIL STD 45662, Z540.1, Z540.3, ISO-17025 requirements, might make the most sense. You might find that you have a measurement need well below the ability of the caliper’s abilities. Testing to your limited needs is another means of not over killing the "performance verification" of the caliper, or other item. Looking at, and reviewing the mentioned standards will give you a likely down-stream point depending how your business develops.
 

apestate

Quite Involved in Discussions
#8
The metrology industry and regulations define traceability as an unbroken chain of measurements going back to national standards, each accompanied with a statement of uncertainty.

Measurement uncertainty is the key to traceability. If I calibrate my own gage blocks, I might create a calculation of measurement uncertainty. I might even be right... but what guarantee do you have that I'm not just some hack?

Therefore, ISO 17025 and laboratory accreditation gives you some universal assurance that the calibration and measurement uncertainty calculation has been done right.

Now, let's say you send your gage blocks to an accredited lab who have adequate capabilities. They are traceable to the standard inch as defined by NIST.

Now, we only really need to calibrate our tools so we can measure our products. We don't need to calibrate our products, so that THEY are traceable to national standards. We simply need to qualify them as good or bad. Therefore, it's OK to skip the measurement uncertainty when you calibrate your calipers, micrometers, bore gages, comparator, indicator, etc.

According to the definition of traceability, you can calibrate your hand tools as long as the masters you are using have been calibrated and accompanied with a statement of inaccuracy estimation.

You can't calibrate a micrometer "in-house" and then use that micrometer to calibrate a caliper's ID jaws. However, if you send that micrometer to an accredited laboratory first, you can use it to calibrate the ID jaws.

That's the rule as I understand it. :>
 
G

George Weiss

#9
The previous post is a very good ISO-17025:2005 posture. All of the descriptions would be of full concern if you are attempting to comply with ISO-17025:2005. If you do go done this course, you might be over killing your needs unless you are willing to go the fully route and become accredited. Claiming 17025 compliance is a flag with no value, and lots of effort. You are claiming ISO 9001, and want a course to follow. Maybe just posting a thread asking for a procedure to verify you gages with your cabinet stored standard.
Have you thought of just marking the gages "reference only", and then just periodically checking them as a functional confidence check, with a date and test value checklist form?.
 
H

Haim Moshe

#10
Thank you all :thanx: for the contributing information. I will summerize and prepare some doc on the matter. I will post it for review and comments. again thx
 
Thread starter Similar threads Forum Replies Date
B IATF 16949 - 7.1.5.3.1 Internal calibration laboratory requirements IATF 16949 - Automotive Quality Systems Standard 9
C MU (Measurement Uncertainty) in an Internal Calibration Laboratory Measurement Uncertainty (MU) 1
I Calibration Void Seals on Internal Equipment General Measurement Device and Calibration Topics 3
A Internal Calibration Process Suggestions General Measurement Device and Calibration Topics 6
S Internal Calibration Laboratory Inspection/Approval General Measurement Device and Calibration Topics 3
S Auditing TS 16949 Cl. 7.6.3.1 - Internal Calibration Laboratory Requirements IATF 16949 - Automotive Quality Systems Standard 8
C Basic Internal Calibration Procedure to check the Accuracy of a Balance Calibration Frequency (Interval) 6
N Internal Calibration vs. Accredited Calibration for Balance - Differences Calibration Frequency (Interval) 3
P Internal Calibration for Laser Temperature Gauges Manufacturing and Related Processes 3
N Scale with Internal Calibration - Do I need to calibrate? How to document this? Calibration Frequency (Interval) 8
L Calibration Procedure for Internal Spline Plug Gage Needed General Measurement Device and Calibration Topics 4
2 TS 16949 Cl. 7.6.3.1 - Internal Calibration Procedure not Traceable to ASTM, EN etc. IATF 16949 - Automotive Quality Systems Standard 20
K Internal laboratory - applies only to Calibration or also to other laboratories? IATF 16949 - Automotive Quality Systems Standard 6
R Standard Methods - Internal Calibration Laboratory - What if there is no standard? IATF 16949 - Automotive Quality Systems Standard 2
P Internal Calibration or Verification - Personnel Requirements IATF 16949 - Automotive Quality Systems Standard 3
M Internal tool calibration program - Best source to become "knowledgeable"? General Measurement Device and Calibration Topics 5
V Internal Scale and Balance Calibration Method Tolerances and Limits General Measurement Device and Calibration Topics 4
D ISO/TS16949 Internal Calibration Lab need to include Production Equipments IATF 16949 - Automotive Quality Systems Standard 26
M How to Internal Audit the Calibration Process Internal Auditing 12
C Internal Diameter Measurement and Pin Gage Calibration General Measurement Device and Calibration Topics 6
K Internal Calibration - Testing Laboratory General Measurement Device and Calibration Topics 9
Q Internal Testing and Calibration Laboratories - MSA, Gage R&R and Control Plans IATF 16949 - Automotive Quality Systems Standard 2
M Company Internal Calibration - Is Internal Calibration a LABORATORY? TS16949 General Measurement Device and Calibration Topics 24
C Internal Audit of Internal Calibration Laboratory - Inspection sample size question Internal Auditing 2
D Internal Laboratory - Defining the Laboratory Scope - Inspection, Test, Calibration Inspection, Prints (Drawings), Testing, Sampling and Related Topics 18
L Internal Calibration Laboratory Requirements - Ambient temperature, Humidity, Etc. Statistical Analysis Tools, Techniques and SPC 3
K Should QS/TS certified internal test/calibration labs be required to be ISO17025 ISO 17025 related Discussions 16
V Does an internal calibration laboratory need to be certified on ISO 17025? ISO 17025 related Discussions 3
Q ISO 9001-2015 Internal audit finding Internal Auditing 12
lanley liao How to understand this words that the planning of internal audit shall take into consideration the results of previous audits? Oil and Gas Industry Standards and Regulations 10
A Add MDSAP to Internal Audit Schedule Medical Device Related Regulations 0
J IATF 16949 Internal Audit question - Auditor's responsibility Internal Auditing 6
M ISO13485:2016, MDSAP and Internal Audits ISO 13485:2016 - Medical Device Quality Management Systems 8
S IATF 16949 Internal Audit Example IATF 16949 - Automotive Quality Systems Standard 7
R AS9100D internal audit checklist or ISO 9001 2015 to AS9100 D AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S Would this be a second site for the purposes of internal and third party audits? General Auditing Discussions 4
V Internal Auditor Competency KPI IATF 16949 - Automotive Quality Systems Standard 14
J ISMS - Internal Audits Internal Auditing 3
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 5
S ISO 9001:2015 Internal Auditing Internal Auditing 8
A Internal Audit Questions ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
L Documented Information in Internal Audits Process (9.2) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
salaheddine96 Internal audit planning Internal Auditing 2
D CB and customer audits considered as internal audits? General Auditing Discussions 9
M ISO 9001 Major Nonconformance Internal Audit Schedule/COVID-19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 18
E MDR internal audit Internal Auditing 1
B Internal Auditor Competency - Product Auditors Internal Auditing 9