Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive)

#1
Hi all.
Looking for some enlightening.
I'm working for a automotive giant in a testing lab as a quality engineer.
This lab provides testing for the plant. To be more precise, we are simulating real life conditions for different car parts. The lab is 17025 accredited with a declared scope of Testing, by the national accreditation organization. As you all may know, as an accredited lab you have to use traceable measuring instruments.
In a continuous search for optimization and cost efficiency I was thing that we might be able to perform some calibrations internally. Now comes the interesting part. We are already performing those type of calibrations, with traceable reference standards, working procedure and uncertainty calculation, all of this not as a official calibration, more as intermediate check in-between third-party calibrations.
Our national accreditation institution states that we have to acreditate separtly in order to able perform those calibrations. But I have different opinion on that.
Since we have already all of the procedures and experience and we already have those audits each year, why we can not incorporate the calibrations in our base qms? Of course with all special requirements:
Working procedure
Qulified personnel
Measurent uncertainty budget
Use of traceable working standards.
Audits on techinal part
Interlabs
Etc.
Dose anybody here have any kind of experience on this subject? Or even a hint.

Thank you all!
 
Elsmar Forum Sponsor

dwperron

Trusted Information Resource
#2
This used to be an easier question to answer.

In 17025:2005 it is crystal clear:
"5.4.6.1 A calibration laboratory, or a testing laboratory performing its own calibrations, shall have and
shall apply a procedure to estimate the uncertainty of measurement for all calibrations and types of
calibrations."
So it was obvious that a testing lab was allowed to calibrate their own equipment. Of course, it would be necessary to meet all of the requirements mandated for a 17025 calibration laboratory. It's just that calibration would not be part of your accreditation, and you thus could not calibrate for an outside party.

In 17025:2017 it gets a bit murky:

"7.6.2 A laboratory performing calibrations, including of its own equipment, shall evaluate the
measurement uncertainty for all calibrations."
Unfortunately, this verbiage is not as clear regarding a testing lab performing its own calibrations, but the intent is still there. After all, why would they mention "including of it own equipment" if they we only referring to calibration laboratories?

So I would agree with your position.
Good luck!
 
#3
This used to be an easier question to answer.

In 17025:2005 it is crystal clear:
"5.4.6.1 A calibration laboratory, or a testing laboratory performing its own calibrations, shall have and
shall apply a procedure to estimate the uncertainty of measurement for all calibrations and types of
calibrations."
So it was obvious that a testing lab was allowed to calibrate their own equipment. Of course, it would be necessary to meet all of the requirements mandated for a 17025 calibration laboratory. It's just that calibration would not be part of your accreditation, and you thus could not calibrate for an outside party.

In 17025:2017 it gets a bit murky:

"7.6.2 A laboratory performing calibrations, including of its own equipment, shall evaluate the
measurement uncertainty for all calibrations."
Unfortunately, this verbiage is not as clear regarding a testing lab performing its own calibrations, but the intent is still there. After all, why would they mention "including of it own equipment" if they we only referring to calibration laboratories?

So I would agree with your position.
Good luck!
Thank you so much. That is very encouraging. I mean, we acknowledge that we have to be evaluated for this part to. But why the need for a separate accreditation? From my perspective this is a birocratic thing.
Do you have a any references from other accreditation institutions that are addressing this issue?
 

dwperron

Trusted Information Resource
#4
The folks at Eurolab say "This clause has remained almost unchanged "
https://www.aphl.org/programs/food_...ments/EUROLAB Handbook ISO IEC 17025 2017.pdf

Canadian Association For Laboratory Accreditation Inc. (CALA) , an organization that accredits testing, not calibration, labs says:
"7.6.2 Internal Calibrations If a laboratory performs internal calibrations on its balances, thermometers and pipettes, it must have a documented procedure that includes the estimation of uncertainty and it must use staff that is appropriately trained. CALA reserves the right to include a calibration expert on assessment teams to laboratories doing their own calibrations. "
https://cala.ca/wp-content/uploads/P07-CALA_Application-2017.pdf
So they speak about testing labs doing their own calibrations.

UKAS states:
"•7.6.2 requires evaluation of measurement uncertainty for all calibrations, including those a laboratory performs on its own equipment (i.e. “in-house” calibrations)"
https://www.ukas.com/download/publi...ditation/ISO-IEC-17025-2017-Presentation.pptx
 

AllTheThings

Involved In Discussions
#5
Hey all, I thought I would resurrect this discussion. I appreciate Sam starting this, and the cites from CALA and UKAS.

I am currently starting to get my lab in shape for ISO 17025 accreditation as a test-only lab. We intend to calibrate our own environmental chambers (temperature and humidity) using traceable accredited calibration humidity and temperature probes.

Usually, we send stuff out, but chambers are special. They don't move easily, if at all. They end up on different schedules, given trying to calibrate around multi-month tests. We are far more familiar with the controllers, sensors, and uncertainties than the cal houses tend to be. We also have enough of them that beyond the convenience, the costs start to break even. When we acquire new to us (but used) chambers or conduct heavy maintenance/sensor replacements, we already run a calibration as a check, before having an external provider in for an accredited calibration.

This all leads in the direction of us doing the calibration work in-house.

We have started putting together uncertainty budgets and more formal procedures for the activity.

My questions (edit, plural :) ): What will an auditor want to see to accept this path? Do we need to note anything special in our documents regarding the scope of the calibrations we do, and the uncertainty approach? Do we need to cite 7.6.2 as justification for not outsourcing the cal? Any impartiality concerns we need to address?
 

dwperron

Trusted Information Resource
#6
My questions (edit, plural :) ): What will an auditor want to see to accept this path? Do we need to note anything special in our documents regarding the scope of the calibrations we do, and the uncertainty approach? Do we need to cite 7.6.2 as justification for not outsourcing the cal? Any impartiality concerns we need to address?

You will need to notify your accreditation body that you perform calibrations on some of your equipment. That heads up will allow them to send an assessor familiar with calibration, and with the proper documentation to audit your calibration program.
Yes, you will need to set up a 17025 compliant calibration program. That will include everything that a calibration lab is required to do under 17025, such as is covered in 17025 Sections 6.4 through 6.6. You will need to produce and validate (with records) your calibration procedures as per Section 7.2. Section 7.7 would normally require you to participate in a proficiency testing program, I would check with your accreditation body and see if they would require that or if you can set up your own internal method to comply with Section 7.7. You will need to keep calibration records as per Section 7.8.4.

That's a start
Good luck!
 

AllTheThings

Involved In Discussions
#7
dwperron,

Thanks for the insights. That is really helpful for me to consider.

Luckily, we have the structure in place for much of those specifics under our ISO 9001 certification...But still have some gaps. With respect to 7.7...Well, 7.7.2, I have been giving that some thought. I may see if the AB lets us satisfy it under 7.7.2b by comparing the calibrated references we send out to in-house working standards. Otherwise, PT for just temperature and humidity isn't too expensive. I will share the experience towards the end of the year, when we are past the audit.
 

Jim Wynne

Staff member
Admin
#8
You will need to notify your accreditation body that you perform calibrations on some of your equipment. That heads up will allow them to send an assessor familiar with calibration, and with the proper documentation to audit your calibration program.
I think there might be some confusion. The OP is preparing for 17025 accreditation.
 

dwperron

Trusted Information Resource
#9
I think there might be some confusion. The OP is preparing for 17025 accreditation.
He is with a 17025 accredited test laboratory that is looking into calibrating their own equipment. That means he will be assessed for his ability to calibrate his equipment to 17025 as well. He was asking what was involved in that process.
 

Jim Wynne

Staff member
Admin
#10
He is with a 17025 accredited test laboratory that is looking into calibrating their own equipment. That means he will be assessed for his ability to calibrate his equipment to 17025 as well. He was asking what was involved in that process.
Please read the 2nd paragraph of the post you quoted.
 
Thread starter Similar threads Forum Replies Date
AllTheThings ISO-17025 accredited TEST lab performing internal calibrations (Yes: Here is how) ISO 17025 related Discussions 6
M Must a company be 17025 accredited to perform internal calibrations? ISO 17025 related Discussions 7
J Performing Internal Calibrations - Dial gauges, micrometers, height gauges, calipers General Measurement Device and Calibration Topics 8
F Internal Laboratory Manual IATF 16949 - Automotive Quality Systems Standard 4
G Internal Laboratory Scope IATF 16949 - Automotive Quality Systems Standard 1
B Internal audit checklist Internal Auditing 5
V Internal Audit Software IATF 16949 - Automotive Quality Systems Standard 5
J Internal Audit Schedule IATF Internal Auditing 3
J API Q1 4.3.2.2 Personnel Competence for Internal Auditors Oil and Gas Industry Standards and Regulations 4
D Internal auditor competence assessment Internal Auditing 12
C ISO 14001 Internal Audit - Opportunity for Improvement ISO 14001:2015 Specific Discussions 2
T Ford CSR for Internal Auditors 2021 Internal Auditing 4
P Does FDA require certification for quality system internal audit for auditor? Qualification and Validation (including 21 CFR Part 11) 1
P Looking to outsource Internal Audit - MDSAP competent auditor needed Other Medical Device Regulations World-Wide 9
D INTERNAL LABORATORY IATF 16949 - Automotive Quality Systems Standard 4
S Transition training for Internal Auditor from ISO 9001: 2008 to the ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
J IATF CB Auditor questions Internal Auditor Competence IATF 16949 - Automotive Quality Systems Standard 19
J Outsourced Internal Audit requirements for Aerospace Suppliers AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 21
D Number of people to be interviewed during an internal audit? Internal Auditing 10
cgaro62 Using internal audits as part of document reviews Document Control Systems, Procedures, Forms and Templates 21
S Is it required to complete Internal Audits within one year? ISO 13485:2016 - Medical Device Quality Management Systems 29
C First Round of QMS Internal Audits - Ethical Dilemma Internal Auditing 10
T How to conduct combined internal audits (9001 and 13485) Internal Auditing 1
Q Easy CARs for Internal Audit ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14
A Internal audit plan and processes for ISO 14001:2015 ISO 14001:2015 Specific Discussions 3
C API Q1 internal audit report Internal Auditing 3
P Filled in F48/F49 for internal audit ISO 17025:2017 Internal Auditing 2
A ISO 9001 Internal Audits - No production right now due to furloughs Internal Auditing 5
J Internal audit random sampling methodology Internal Auditing 2
G Organizing internal audit program for an Integrated QHSE Management System Internal Auditing 13
W How do you phrase your internal audit questions? Internal Auditing 3
M IATF - Internal Audit 3 year span Internal Auditing 4
T Internal NCR - API Q1 5.4.2b - API 16C Clause 4.1.3 Internal Auditing 9
Q ISO 9001-2015 Internal audit finding Internal Auditing 14
lanley liao How to understand this words that the planning of internal audit shall take into consideration the results of previous audits? Oil and Gas Industry Standards and Regulations 10
A Add MDSAP to Internal Audit Schedule Medical Device Related Regulations 0
J IATF 16949 Internal Audit question - Auditor's responsibility Internal Auditing 6
M ISO13485:2016, MDSAP and Internal Audits ISO 13485:2016 - Medical Device Quality Management Systems 8
S IATF 16949 Internal Audit Example IATF 16949 - Automotive Quality Systems Standard 18
R AS9100D internal audit checklist or ISO 9001 2015 to AS9100 D AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S Would this be a second site for the purposes of internal and third party audits? General Auditing Discussions 4
V Internal Auditor Competency KPI IATF 16949 - Automotive Quality Systems Standard 15
J ISMS - Internal Audits Internal Auditing 5
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 8
S ISO 9001:2015 Internal Auditing Internal Auditing 8
A Internal Audit Questions ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
L Documented Information in Internal Audits Process (9.2) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
salaheddine96 Internal audit planning Internal Auditing 2
D CB and customer audits considered as internal audits? General Auditing Discussions 9
M ISO 9001 Major Nonconformance Internal Audit Schedule/COVID-19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 18

Similar threads

Top Bottom