Internal Results vs. Independent Testing Laboratory - Inkometer

Southern Cross

Involved In Discussions
#1
Our product is tested with an instrument called an inkometer. The product is a resin solution used to manufacture printing ink. At the moment we test on our own machine, then send a sample to the independent lab. We are seeing multiple rejections of our product, requiring re-adjustment prior to sending the product to our customer. We have known about our problems for some time, finally I have time to try to address them. I've done a Gage R&R test last year and the total R&R was 23%. Not too bad for a process industry instrument. Nevertheless, we are still encountering rejections.

The specification at the independent lab is 23.5 to 25. Roughly 40% of the samples we send fall outside that range according the the independent lab. See the enclosed files for the graphical information.

At this stage I'm more interested in the instrument itself. It is used by placing 1.32 mls. of solution on the rollers. The machine runs at 800rpm for one minute (90oF) and then takes a reading of the force exerted on one of the rollers. I've been checking the instrument each day recently using the calibration weights and I'm seeing significant variations without even using a sample. Cp without a sample is 1.3.

I need to run an experiment to test for variations in result based on (1) temperature and (2) sample volume. The problem is that I can't easily control how much volume is placed on the instrument, so a standard full factorial DOE is not possible. What is the best way to approach this?

1.JPG

2.JPG
 
Elsmar Forum Sponsor
J

justncredible

#3
Ignore the sample volume on the rollers, you might instead use viscosity or some other type to monitor the ink sample.

From what you have said I would be looking at the test rollers for bearing wear and tension spring wear.

It has been some years since I worked in printing, and it was web presses. The bearings on the rollers gave us alot of problems. As did the glue. We used a hand roller to calculate the spread of inks, all very rough ways to try and control a process.

You might increase the sample number to smooth out the noise of the process. Do extra runs to get a decent picture of what is affecting the process as well. Make sure you track the humidity as well as temps, also the basis weights of the stock. Ink laydown and saturation also can play into a study.

Also I would send a extra sample to a diffrent lab and make sure the 2 samples from the labs match each other.

Printing vexxed me to no end......600 bags a min and I have no idea why they all came out looking fine..
 

Hershal

Metrologist-Auditor
Trusted Information Resource
#4
The previous reply is right on track. Also, to reduce the potential variation from your monitoring items, have the weights and any temperature monitoring device calibrated by an accredited cal lab and make sure you get the uncertainties.
 

Southern Cross

Involved In Discussions
#5
We already use a rheometer to control viscosity and psuedoplasticity. What we have found over the last few years is that tack is independent of viscosity in our product.

The instrument is only a year or two old and it's the latest model. It is calibrated, with temperature, on a regular basis. Our balances are calibrated on a regular basis.

What I'm seeing at this stage, even just using the weights, there is a good deal of variation in results from day to day within a range of about 0.4 units. Not so good, but I'm going to have to live with it.

My interest at the moment is in trying to determine how much the temperature and sample volume change the result. I've been working in the industry for ~20 years and I've never seen anyone actually try to systematically work this out before.

Fortunately we're not the ink manufacturer or user. I don't have to concern myself with paper variation due to humidity etc. That's our customer's problem and they keep that information to themselves.
 

Stijloor

Leader
Super Moderator
#6
<snip>My interest at the moment is in trying to determine how much the temperature and sample volume change the result. I've been working in the industry for ~20 years and I've never seen anyone actually try to systematically work this out before.
Could a Design of Experiment (DOE) be of help here?

Stijloor.
 

Southern Cross

Involved In Discussions
#7
Could a Design of Experiment (DOE) be of help here?

Stijloor.
That's what I'd like to do, but I won't be able to control the sample volume accurately. Best I can do is add some or take some away, then weigh how much is applied by difference. I'm thinking that multiple regression is going to be the best approach?
 
J

justncredible

#8
How much temperature variation do you have in a normal day? Is the unit in a controlled lab?

I do think your on the right track, also track the humidity. Make sure it is well grounded, a ungrounded paper conversion press can produce so much static it will actually warp light. Make sure plenty of brushes contact the paper. You might also see if the pumps are shearing the ink. A viscosity test before and after might show shearing of the ink.

To find the interactions first you must figure out the gage issue. How precise and accurate is it? Will it be good enough to study the tackiness of the ink? How do they calibrate the gage? Do you have a GR&R for it?
 
J

justncredible

#9
How much temperature variation do you have in a normal day? Is the unit in a controlled lab?

I do think your on the right track, also track the humidity. You might also see if the pumps are shearing the ink. A viscosity test before and after might show shearing of the ink.

To find the interactions first you must figure out the gage issue. How precise and accurate is it? Will it be good enough to study the tackiness of the ink? How do they calibrate the gage? Do you have a GR&R for it?

Get a thermometer and measure the weights on the unit at every degree change thruout the day. Springs are pretty wild and do things unforseen.
 
D

DsqrdDGD909

#10
That's what I'd like to do, but I won't be able to control the sample volume accurately. Best I can do is add some or take some away, then weigh how much is applied by difference. I'm thinking that multiple regression is going to be the best approach?
There are some highly accurate adjustable micropipettes.
Capacity Increments Accuracy (CV%)
0.2 - 2.0mL 0.01mL ±0.7 ±0.5 @ 0.2mL
±0.5 ±0.2 @ 2mL
1 - 10mL 0.1mL ±0.5 ±0.3 @ 1mL
±0.5 ±0.15 @ 10mL
 
Thread starter Similar threads Forum Replies Date
S Minimum Retention Time for Records of internal audit results as per AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
lanley liao How to understand this words that the planning of internal audit shall take into consideration the results of previous audits? Oil and Gas Industry Standards and Regulations 10
C How should I present Internal Audit results to Management in Management Review Internal Auditing 18
Casana The right way to Report Internal Audit Results Internal Auditing 20
C Reporting Profile of a Surface Results on an Internal Cylindrical Feature Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
O Should Staff have Access to see Internal and/or CB Audit Results? Internal Auditing 5
L Internal Audit Recommendations - CAPA from audit results Nonconformance and Corrective Action 7
Q Evaluating Internal Audit Results Internal Auditing 24
F Software Program for tracking internal audit results needed Quality Assurance and Compliance Software Tools and Solutions 2
J Process Audit Checklist to document Internal Process Audit results Process Audits and Layered Process Audits 19
K Are internal audit results a true reflection of the organisations quality system Internal Auditing 2
Q IATF 16949 9.2 Internal Auditor IATF 16949 - Automotive Quality Systems Standard 13
T Internal Auditor Technologies AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
Q Internal lab IATF 16949 - Automotive Quality Systems Standard 1
D What are the rules regarding coverage in internal audits? IATF 16949 - Automotive Quality Systems Standard 5
D What are the IATF rules regarding doing internal audits remotely? IATF 16949 - Automotive Quality Systems Standard 5
S 7.1.5.3.1 Internal Lab Scope IATF 16949 - Automotive Quality Systems Standard 15
K Best Internal Auditor Training Internal Auditing 18
S 9001 + 14001 Internal Auditing Process checklist, flowchart or questionnaire ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
B Single fault analysis and negligible probability of IC internal fault IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T AS9100D Risk-Based Internal Audit Schedule AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 10
S AS9120 - New Internal Auditor Training requirement? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
A Internal Audits - skipped clauses? Internal Auditing 11
B Does anyone copyright their internal QMS documents? Document Control Systems, Procedures, Forms and Templates 24
B Establishing topics for IATF internal audit processes Internal Auditing 9
L Internal Audits during the MDR Transition Period EU Medical Device Regulations 5
T Robust internal audit program AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
J Average number of Nonconformances during internal quality audit for Medical Device Manufacturers Internal Auditing 3
Q AS9100:D Counterfeit internal audit questions AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
D ISO 9001:2015 Internal Audit Check Sheet ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
K Qualified MDR internal auditor? EU Medical Device Regulations 8
L Internal audits for IATF IATF 16949 - Automotive Quality Systems Standard 7
E FDA & Internal Audits US Medical Device Regulations 3
LabCat Internal SOP for a validated (ISO) method? ISO 17025 related Discussions 2
L Documenting internal audit of customer specific requirements IATF 16949 - Automotive Quality Systems Standard 10
T Internal and external communication procedure for Food Safety Food Safety - ISO 22000, HACCP (21 CFR 120) 2
R Looking for ISO 13485 Internal Audit Checklist ISO 13485:2016 - Medical Device Quality Management Systems 8
E Cybersecurity for Internal Tool Validation Medical Device and FDA Regulations and Standards News 1
G During internal audit - finding poor action plans ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 18
G Opening meeting - internal audit ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
G NC criteria during internal audits ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
A Information on Process-based Internal Auditing Needed Internal Auditing 6
N Internal NCR Meetings Benchmarking 15
H Ever given an Operations Manager an internal Corrective Action Request? Nonconformance and Corrective Action 5
TheAxiomaticHamster Internal ISO17025:2017 Compliant Reporting ISO 17025 related Discussions 2
Mr Roo Discovered dishonesty after performing an internal audit General Auditing Discussions 5
Q Internal calibration (stopwatch) IATF 16949 - Automotive Quality Systems Standard 3
C Internal Laboratory (7.1.5.3.1) IATF 16949 - Automotive Quality Systems Standard 3
W IATF 9.2.2.1 Internal Audit how to determine risk IATF 16949 - Automotive Quality Systems Standard 12
B Consistently Late Internal Audits- Any Suggestions? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 116

Similar threads

Top Bottom