Interpretation of GMP Requirements for class 1 medical device manufacturer (device GMP exempt, only General controls applicable)

Aphel

Involved In Discussions
#1
Dear all,
Would Need your Support for interpratation 21 CFR 820 --- Section 820.180 (Records).
As a class 1 device manufacturer our devices are exempt from GMP - apart from §820.198 and §820.180. This is clear for me.

My question:
820.180 - 1st sentence - requires: "all records by this part shall be maintained at the manufacturing Establishment..."
What does "all records" mean exactly?

Does it mean ALL records required within GMP sections e.g. design Control, purchasing, production, CAPA,... have to be created and filed?
OR
Does it mean only those records of applicable sections - in my case 820.198 - have to be created and filed?

Thank you very much in advance!
Regards, Aphel
 
Elsmar Forum Sponsor

shimonv

Trusted Information Resource
#2
When you read "All records required by this part", I take it to mean part 820.
In other words, you need to apply record controls to all QSR records.

Shimon
 

William55401

Quite Involved in Discussions
#3
Aphel. You only need to apply record controls to the QSR sections that you must comply with. In this case, you are only responsible for the complaint handling element (and MDR since it is mentioned).
 

Tidge

Trusted Information Resource
#4
My question:
820.180 - 1st sentence - requires: "all records by this part shall be maintained at the manufacturing Establishment..."
What does "all records" mean exactly?

Does it mean ALL records required within GMP sections e.g. design Control, purchasing, production, CAPA,... have to be created and filed?
OR
Does it mean only those records of applicable sections - in my case 820.198 - have to be created and filed?
When you read "All records required by this part", I take it to mean part 820.
In other words, you need to apply record controls to all QSR records.
@shimonv is correct. You may be exempt from section 820.30 Design Controls (consult the class I device table in that section). I am not aware of any other element of 21 CFR 820 which exempts manufacturer of class I devices.

You should also review the Preamble, including the response 160.
 

Aphel

Involved In Discussions
#5
Dear all,
Thank you for your response.

To be clear now:

As a medical device class 1 GMP exempt manufacturer - sections §820.198 and §820.180 apply (as part of "General controls")

So I Need not to establish required procedures (apart from §820.198) BUT I Need to establish required records... correct?

Example:
§820.50 requires a procedure for supplier Evaluation, and results of supplier Evaluation have to be recorded.

In my case - I Need not to establish the required procedure (because I am GMP exempt) , but I Need to document activities out of supplier Evaluation (due to the requirements of §820.180) - correct?

Thanks once again for your help to clarify my issue...
BR Aphel
 
Thread starter Similar threads Forum Replies Date
A General Controls - GMP interpretation? General Requirements Concerning Records 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
H IEC 61010-1:2010/A1:2016 - Clause 10.5.3 - (b) - Interpretation CE Marking (Conformité Européene) / CB Scheme 6
M Rebuild interpretation Federal Aviation Administration (FAA) Standards and Requirements 3
O Article 120 - significant change interpretation EU Medical Device Regulations 7
Ashland78 9.3.2.1 Sanctioned Interpretation IATF 16949 - Automotive Quality Systems Standard 2
W 7.1.5.3.2. External laboratory Sanctioned Interpretation IATF 16949 - Automotive Quality Systems Standard 4
F Interpretation of 60601-1 Figure J.5 (Isolation, example 5) for a rechargeable, battery operated device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Sanctioned interpretation #10 - ISO 17025 IATF 16949 - Automotive Quality Systems Standard 2
somashekar Interpretation of 6.3 Infrastructure in the context of new machines ISO 13485:2016 - Medical Device Quality Management Systems 7
Q Contingency plan - IATF Sanctioned Interpretation 17 IATF 16949 - Automotive Quality Systems Standard 4
U MDD Article 12/ MDr Article 22 interpretation EU Medical Device Regulations 2
R MDR, Annex I, 23.1 Interpretation - IFU on the website EU Medical Device Regulations 5
M MDR Article 22 Interpretation EU Medical Device Regulations 3
I Interpretation of the MDR GSPR 23.4 (u) EU Medical Device Regulations 5
D IATF16949 - Interpretation of Customer Requirements clauses IATF 16949 - Automotive Quality Systems Standard 3
P MDR Rule 10 interpretation - Active Device EU Medical Device Regulations 6
Q % Study variation low, % tolerance high - GR&R Interpretation help Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
A Interpretation with regards to Ppk > Cpk Capability, Accuracy and Stability - Processes, Machines, etc. 14
A OHSAS 18001 external auditor finding personal interpretation? Occupational Health & Safety Management Standards 5
A OEM branding - My interpretation of the LVD Directive makes us a manufacturer CE Marking (Conformité Européene) / CB Scheme 3
P Average Peel Strength - Interpretation of BS EN 868-5:2018 and ASTM F0088/F0088M Other Medical Device Related Standards 2
B ISO 50001 Interpretation of section 3.3.9 (Outsourcing) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
K Interpretation of significant change - material change Canada Medical Device Regulations 3
R MDR Software Rule 11 Formal Interpretation EU Medical Device Regulations 7
A IATF Sanctioned Interpretation No. 7 - Type and Extent of Control (supplemental) IATF 16949 - Automotive Quality Systems Standard 4
F Interpretation of MDR 2017/745 Article 23 - CE Mark Requirements EU Medical Device Regulations 8
U Hand-Held dosing device has no PATIENT - Interpretation of the PATIENT definition IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
G Medical Device "Immediate Container" Interpretation of Definition US Food and Drug Administration (FDA) 2
A Interpretation of Article 16 (2b) - Packaging, samples and Certificate EU Medical Device Regulations 10
D EU MDR Corrigendum Interpretation EU Medical Device Regulations 3
M Informational MDCG 2019-3 Interpretation of Article 54(2)b – Pre- market clinical evaluation consultation procedure with the involvement of expert panels Medical Device and FDA Regulations and Standards News 0
G EU MDR 2017/745 Rule 11 interpretation - Re-classification of a Software as Medical Device Other Medical Device Related Standards 0
JoshuaFroud Interpretation of Clause 5.5.2 in ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 2
qualprod P x I = Value interpretation for residual risk? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
T CSA Z299.3-85 Nuclear interpretation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
D VDA Trigger Matrix in the VDA 6.2 Manual - Interpretation and Use VDA Standards - Germany's Automotive Standards 3
B Interpretation of Customer Specific Requirements of Continental - Records Retention Customer and Company Specific Requirements 6
S Interpretation or Definition of ‘Once Every 5 Days’ ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
N What is the interpretation I-MR-R chart in this question? Statistical Analysis Tools, Techniques and SPC 1
E ME (Medical Equipment) Systems - IEC 60601-1 Clause 16.1 Interpretation IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
P IATF 16949 Cl. 7.1.5.2.1 "Calibration/Verification Records" Interpretation IATF 16949 - Automotive Quality Systems Standard 13
B Clarification on interpretation of some EN ISO 14971:2012 & IEC 62304:2006 req's ISO 14971 - Medical Device Risk Management 46
M IATF 16949 - 7.1.3.1 Plant, Facility, and Equipment Planning - Interpretation IATF 16949 - Automotive Quality Systems Standard 8
R Location Interpretation on Drawing - Hole to Hole? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
T TS 16949 Clause 7.4.1.2 and Sanctioned Interpretation IATF 16949 - Automotive Quality Systems Standard 12
D Interpretation of new IAQG ruling - Audit duration for nonconformance verification AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
K Interpretation of IEC 60601-1-2 Electromagnetic Compatibility – Requirements & tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Heat Treatment - Soaking time interpretation Manufacturing and Related Processes 5
D Interpretation of DOE interaction plot Using Minitab Software 8
A Training material for interpretation & understanding Part 11 requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2

Similar threads

Top Bottom