Hi Everyone, would like to get clarified few clauses – interpretation of IEC 60601-1-2 - Electromagnetic compatibility –Requirements and tests
1. Is sub-system’s EMC/I test reports are acceptable, what is the condition of acceptability & technical consideration? For example we procure X-ray generator, X-ray tube, and x-ray detector that are IEC tested, if we change these models from one vendor to another for the same terminal specifications, should we repeat the entire SYSTEM testing? Please clarify.
2. What are the requirements applicable for EQUIPMENT and SYSTEMS specified for use only in a shielded location?
3. For example, if the attenuation by shielding material is less than 20db, will the attenuation by the room shielding material on Radiation and conducted Emission taken into account and that value is relaxed from the Radiated/conducted Emission levels that are measured?
If not, what is the rationale behind this, can anyone please explain?
4. Radiated Emission – to apply the 20db relaxation – for the equipment when installed in a shielded location. Reference to the following “note” in IEC 60601-1-2.
“It is essential that the actual RF shielding effectiveness and filter attenuation of the shielded location be verified to ensure that they meet or exceed the specified minimum values”
What is the verification methods to use the relaxation, testing is conducted in-situ with shield present otherwise how it will be done?
5. We find the following point from 4th edition of IEC,
“Specific medical environments are now clearly defined as; Emergency Medical Services Environment, Home Healthcare Environment, Professional Healthcare Facility Environment and Special Environments.
Professional healthcare environment is considered a controlled EM environment. MRI, x-ray and other medical equipment requiring shielded rooms have special label and user information detailing the use in special environments. X-ray generators operating in intermittent mode may have the quasi-peak limits relaxed by 20dB for discontinuous conducted and radiated emissions.”
Can anyone clarify on following points:
- Does it mean that the testing has to be done during X-ray firing or is it ok to test in stand by condition?
- What does it mean – “controlled EM environment”?
Please help me out
1. Is sub-system’s EMC/I test reports are acceptable, what is the condition of acceptability & technical consideration? For example we procure X-ray generator, X-ray tube, and x-ray detector that are IEC tested, if we change these models from one vendor to another for the same terminal specifications, should we repeat the entire SYSTEM testing? Please clarify.
2. What are the requirements applicable for EQUIPMENT and SYSTEMS specified for use only in a shielded location?
3. For example, if the attenuation by shielding material is less than 20db, will the attenuation by the room shielding material on Radiation and conducted Emission taken into account and that value is relaxed from the Radiated/conducted Emission levels that are measured?
If not, what is the rationale behind this, can anyone please explain?
4. Radiated Emission – to apply the 20db relaxation – for the equipment when installed in a shielded location. Reference to the following “note” in IEC 60601-1-2.
“It is essential that the actual RF shielding effectiveness and filter attenuation of the shielded location be verified to ensure that they meet or exceed the specified minimum values”
What is the verification methods to use the relaxation, testing is conducted in-situ with shield present otherwise how it will be done?
5. We find the following point from 4th edition of IEC,
“Specific medical environments are now clearly defined as; Emergency Medical Services Environment, Home Healthcare Environment, Professional Healthcare Facility Environment and Special Environments.
Professional healthcare environment is considered a controlled EM environment. MRI, x-ray and other medical equipment requiring shielded rooms have special label and user information detailing the use in special environments. X-ray generators operating in intermittent mode may have the quasi-peak limits relaxed by 20dB for discontinuous conducted and radiated emissions.”
Can anyone clarify on following points:
- Does it mean that the testing has to be done during X-ray firing or is it ok to test in stand by condition?
- What does it mean – “controlled EM environment”?
Please help me out