Hi all

I'm wrestling with the interpretation of the MDR GSPR 23.4 (u) where a requirement in the IFU is that "in the case of implantable devices, the overall qualitative and quantitative information on the materials and substances to which patients can be exposed;"

Obviously manufacturers will not want to release the quantitative formulation of your device into the public domain when you're attempting to complete against other companies.

Additionally I noted that, it's completely plausible that when attempting to be in compliance with the labelling requirements under GSPR 23.4(u) and GSPR 10.4.5 the reach different conclusions and end up contradicting each other.

Whereas when GSPR 10.4 provides a clear understanding on what the context for identifying chemicals on the labelling are... I've not seen any clarification or guidance on how to interpret GSPR 23.4 (u) which is only a single sentence.

Has anybody seen any guidance or anything which would help with the correct interpretation of GSPR 23.4(u)? Or perhaps someone has started discussions with their notified body already on this point and could offer some insight into how the notified bodies are interpreting this?

Any feedback or thoughts would be welcome!
 
Elsmar Forum Sponsor

dgrainger

Trusted Information Resource
#2
Manufacturers many not want to but they have to include the information in the IFU - it's not optional - Also required in the Patient information requirements of Article 18(d).
 

monoj mon

Involved In Discussions
#3
Obviously manufacturers will not want to release the quantitative formulation of your device into the public domain when you're attempting to complete against other companies.
To my understanding, this requirement is asking the manufacturer to provide the final (overall) composition which the patient may be exposed to during implantation. For example, if your product is a copolymer, you are supposed to give the information on the final material(s) but not the individual monomer constituents. If it has a degradation profile, you may qualitatively/quantitatively identify it along with what is degraded and how it will be released from the body.
 
Thread starter Similar threads Forum Replies Date
R MDR, Annex I, 23.1 Interpretation - IFU on the website EU Medical Device Regulations 5
M MDR Article 22 Interpretation EU Medical Device Regulations 3
P MDR Rule 10 interpretation - Active Device EU Medical Device Regulations 3
R MDR Software Rule 11 Formal Interpretation EU Medical Device Regulations 7
F Interpretation of MDR 2017/745 Article 23 - CE Mark Requirements EU Medical Device Regulations 8
D EU MDR Corrigendum Interpretation EU Medical Device Regulations 3
G EU MDR 2017/745 Rule 11 interpretation - Re-classification of a Software as Medical Device Other Medical Device Related Standards 0
L Interpretation of MDR Reportable Event - FDA guidance on Medical Device Reporting 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 16
D IATF16949 - Interpretation of Customer Requirements clauses Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 2
A Interpretation of GMP Requirements for class 1 medical device manufacturer (device GMP exempt, only General controls applicable) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
Q % Study variation low, % tolerance high - GR&R Interpretation help Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
A Interpretation with regards to Ppk > Cpk Capability, Accuracy and Stability - Processes, Machines, etc. 14
A OHSAS 18001 external auditor finding personal interpretation? Occupational Health & Safety Management Standards 5
A OEM branding - My interpretation of the LVD Directive makes us a manufacturer CE Marking (Conformité Européene) / CB Scheme 3
P Average Peel Strength - Interpretation of BS EN 868-5:2018 and ASTM F0088/F0088M Other Medical Device Related Standards 2
B ISO 50001 Interpretation of section 3.3.9 (Outsourcing) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
K Interpretation of significant change - material change Canada Medical Device Regulations 3
A IATF Sanctioned Interpretation No. 7 - Type and Extent of Control (supplemental) IATF 16949 - Automotive Quality Systems Standard 4
U Hand-Held dosing device has no PATIENT - Interpretation of the PATIENT definition IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
G Medical Device "Immediate Container" Interpretation of Definition US Food and Drug Administration (FDA) 2
A Interpretation of Article 16 (2b) - Packaging, samples and Certificate EU Medical Device Regulations 10
M Informational MDCG 2019-3 Interpretation of Article 54(2)b – Pre- market clinical evaluation consultation procedure with the involvement of expert panels Medical Device and FDA Regulations and Standards News 0
JoshuaFroud Interpretation of Clause 5.5.2 in ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 2
qualprod P x I = Value interpretation for residual risk? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
T CSA Z299.3-85 Nuclear interpretation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
D VDA Trigger Matrix in the VDA 6.2 Manual - Interpretation and Use VDA Standards - Germany's Automotive Standards 3
B Interpretation of Customer Specific Requirements of Continental - Records Retention Customer and Company Specific Requirements 6
S Interpretation or Definition of ‘Once Every 5 Days’ ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
N What is the interpretation I-MR-R chart in this question? Statistical Analysis Tools, Techniques and SPC 1
E ME (Medical Equipment) Systems - IEC 60601-1 Clause 16.1 Interpretation IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
P IATF 16949 Cl. 7.1.5.2.1 "Calibration/Verification Records" Interpretation IATF 16949 - Automotive Quality Systems Standard 12
B Clarification on interpretation of some EN ISO 14971:2012 & IEC 62304:2006 req's ISO 14971 - Medical Device Risk Management 46
M IATF 16949 - 7.1.3.1 Plant, Facility, and Equipment Planning - Interpretation IATF 16949 - Automotive Quality Systems Standard 8
R Location Interpretation on Drawing - Hole to Hole? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
T TS 16949 Clause 7.4.1.2 and Sanctioned Interpretation IATF 16949 - Automotive Quality Systems Standard 12
D Interpretation of new IAQG ruling - Audit duration for nonconformance verification AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
K Interpretation of IEC 60601-1-2 Electromagnetic Compatibility – Requirements & tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Heat Treatment - Soaking time interpretation Manufacturing and Related Processes 5
D Interpretation of DOE interaction plot Using Minitab Software 8
A Training material for interpretation & understanding Part 11 requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
L Interpretation of "Any or All"'? IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
G Interpretation of phrase - "Direct Authority" EASA and JAA Aviation Standards and Requirements 5
pittmatj Brazil: Interpretation of RDC 185 Other Medical Device Regulations World-Wide 6
D What is your understanding or interpretation of TS16949 7.4.1.2 IATF 16949 - Automotive Quality Systems Standard 6
M Interpretation of Plus/Minus Draft - Injection Molded Parts Inspection, Prints (Drawings), Testing, Sampling and Related Topics 10
E FDA Significant Change Interpretation - Guidance Document EU Medical Device Regulations 2
0 ISO 9001:2008 interpretation not the same per country? (Netherlands and France) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Multiple Standard Interpretation Queries Occupational Health & Safety Management Standards 4
A Need MSA %GRR Interpretation Advice Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
P EU Conflict Minerals Regulation Interpretation RoHS, REACH, ELV, IMDS and Restricted Substances 3

Similar threads

Top Bottom