Interpreting Measurement Uncertainty for Temperature PRT probes
Mike Boothe
Calibration Uncertainty statements for Temperature PRT probes state a value for a cretain range, example:
±6.3 mK from -40 to 0 °C
±11.5 mK from 0 to 420 °C
±26 mK from 420 to 600 °C
±43 mK from 600 to 661 °C
(0 °C is set at ice point to ±3 mK)
This is stating that the worst case error for the range is the ±xx mK value. Can it be assumed that the worst case error is at the extreeme of the range?
If so then can the range be plotted out vs uncertainty so at a point along the range the uncertainty from the line slope can be used?
Is this an acceptable practice?
Getting back to my example, if I wanted to find the uncertainty at 155 °C (from 0 to 420 °C) the line slope would be at about 6 mK (from 3 to 11.5 mK).
jfgunn
I do not think that you can draw the line and interpolate as you suggest. If this were the case, the uncertainty would have been spec'd as a y=mx+b equation.
Based on your example, the uncertarinty at 155C would be +/-11.5mK.
I look forward to other responses.
blueicecube2
i agree, interpolation doesn't work here. MU aren't exactly linearly calculated.
as a general rule, when it comes to choosing an MU, choose the worse, max.
safer that way too..
although the best way is to understand the calculation behind it & align you assumptions accordingly to justify your choice
B Length Measure - Interpreting Calibration Results and Measurement Uncertainty (MU) Measurement Uncertainty (MU) 3
P Interpreting Span Measurement - P95-P5 Six Sigma 1
G Interpreting Type 1 MSA (Measurement Systems Analysis) Results Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Q Interpreting Tensile Charts General Measurement Device and Calibration Topics 0
Q Interpreting China Medical Device regulations/standards China Medical Device Regulations 1
P ISO 80369-7 standard - Interpreting which Parts should be in scope Other Medical Device Related Standards 7
K PACS - interpreting MDD and Borderline Guidance CE Marking (Conformité Européene) / CB Scheme 5
M List of Packaging Contents - Medical Devices - Interpreting EU Directive 93/42/EEC CE Marking (Conformité Européene) / CB Scheme 3
D Interpreting Normal vs Weibull Capabilities Capability, Accuracy and Stability - Processes, Machines, etc. 4
B Interpreting Deviations 5 & 6 in Annex ZA in ISO 14971:2012 ISO 14971 - Medical Device Risk Management 1
B Interpreting "misuse" when assessing Hazardous Situations ISO 14971 - Medical Device Risk Management 2
S Interpreting clause 4.1.5 in ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 9
M Interpreting Process Controls - 21 CFR Part 820.70(a) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
M Help interpreting 21 CFR Part 806 (corrections and removals) Other US Medical Device Regulations 1
M Help interpreting MIL-STD-105E Single Sampling Plans Tables Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
M Biocompatibility in Respiratory Products - Help interpreting whitepaper IEC 60601 - Medical Electrical Equipment Safety Standards Series 22
I Interpreting Product Realization (Clause 7) in ISO 9001:2008 for Service Industry ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
V GD&T Interpreting Datums in Two Single Segmented Position Tolerances Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
M Interpreting X bar and R Control Charts Statistical Analysis Tools, Techniques and SPC 2
B Help with interpreting stock market terminologies Coffee Break and Water Cooler Discussions 4
V Interpreting Zinc Plating Specification GMW 3044 - 6K96/48 APQP and PPAP 2
K Interpreting and Applying 7.3.2 Design and Development Inputs ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
S Interpreting the Calibration Report for my Dial Indicator General Measurement Device and Calibration Topics 3
J Interpreting Process Capability results and ratios Capability, Accuracy and Stability - Processes, Machines, etc. 1
S Interpreting Level of Confidence - Round Robin for Tensile Testing - Help needed Statistical Analysis Tools, Techniques and SPC 4
M Interpreting AS9100 Clause 8.2.2 Internal Audit Requirements Internal Auditing 28
T Interpreting my t test in Minitab Using Minitab Software 2
S Interpreting Linear Regression Results from Minitab Using Minitab Software 15
optomist1 Interpreting LMC for Pattern of Slots (GD&T) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
optomist1 Interpreting Minitab Test For Equal Variances Using Minitab Software 3
V Interpreting Minitab Gauge R&R Results Using Minitab Software 6
A ISO 2081 - Interpreting Coating Thickness Other ISO and International Standards and European Regulations 3
L Interpreting Injection Molding Tooling Documents - Cavities Manufacturing and Related Processes 3
C Interpreting Gage R&R Results - 3 operators, 3 iterations and 10 parts Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
N Interpreting Hygrometer Calibration Uncertainty - Temperature Coefficient Measurement Uncertainty (MU) 1
mr500 Interpreting Decimals in the mm state - Off Wall Question Inspection, Prints (Drawings), Testing, Sampling and Related Topics 14
J Interpreting clause 7.5.2.1 (validation of software used in production & service) ISO 13485:2016 - Medical Device Quality Management Systems 2
E Interpreting Partial Least Square Results Using Minitab Software 1
C Interpreting Outside Laboratory Calibration Certificate Measurement Uncertainty (MU) 3
M Interpreting MIL-PRF-19500 (Performance Spec, Semicon Dev) - What is LTPD of 20 - ??? Manufacturing and Related Processes 4
T Rules for interpreting control charts Statistical Analysis Tools, Techniques and SPC 2
Q Interpreting Responsibility and Authority Clause 5.5.1 ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 18
H Interpreting 'Evaluation of compliance' in ISO 14001 Internal Auditing 4
P Micrometer Gage R&R study - Interpreting the data and suggestions Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
W Interpreting a Normal Probability Plot Statistical Analysis Tools, Techniques and SPC 4
Wes Bucey "Downsizing" ramifications - Evaluating and Interpreting the News World News 0
Q Interpreting calibration result of Air Particle Counter on Counting efficiency General Measurement Device and Calibration Topics 1
A Particle Counter Results - Interpreting Data for a Class 100,000 Cleanroom Other Medical Device Related Standards 12
M GD&T Q&A session - Interpreting FCF (Feature Control Frame) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 43
N TL 9000 - Help Interpreting Normalization Unit for Measurements TL 9000 Telecommunications Standard and QuEST 2