I've got several points of confusion regarding requirements of 21 CFR 820.70(a) (General production & process controls).

Any guidance to any of the points below would be much appreciated:

1. "Where deviations from device specifications could occur as a result of the manufacturing process..."
Are there any examples of devices NOT subject to this criteria? It seems to me that ALL devices would meet this criteria, and hence be subject to all the requirements that follow...

2. "Monitoring and control of process parameters and component and device characteristics during production;"
Assuming there are no quantifiable/measurable "parameters", would it be sufficient just to document in-process acceptance activities at different stages of production to fulfill this?

3. "Compliance with specified reference standards or codes;"
I'm assuming this is mainly applicable to processes like sterilization or clean-room practices for which standards or codes exist? ...but the way the regulation is worded, it appears that this is a requirement regardless (i.e. "Where process controls are needed they shall include:...").

4. "Criteria for workmanship which shall be expressed in documented standards or by means of identified and approved representative samples."
By "documented standards", does acceptance criteria (for in-process and finished product), or receiving inspection criteria (for outsourced components) qualify? ...or does "standards" mean an established national or international standard?

-----
It seems to me that 820.70(a) is already covered by subsequent parts: 820.72 (Equipment), 820.75 (Process validation), and Subpart H (Acceptance activities).

But then, perhaps I'm missing the intent of the regulation here.
(if so, please set me straight)....
 
Elsmar Forum Sponsor

Mark Meer

Trusted Information Resource
#2
Re: Help interpreting process controls 820.70(a)

Sorry... perhaps the lengthy post is deterring replies. I'll be more concise:

With respect to 21 CFR 820.70(a) (General production and process controls):

- Do all the subpoints apply to all devices? (the way it's worded it appears so...)
- When it mentions "standards", does this necessarily mean an established national or international standards?

I find this section very difficult to interpret... any help would be greatly appreciated.
 

Mark Meer

Trusted Information Resource
#4
Re: Help interpreting process controls 820.70(a)

Geez, if this is the case, the FDA sure is unnecessarily wordy about it!

"Each manufacturer shall develop, conduct, control, and monitor production processes to ensure that a device conforms to its specifications. Where deviations from device specifications could occur as a result of the manufacturing process, the manufacturer shall establish and maintain process control procedures that describe any process controls necessary to ensure conformance to specifications. Where process controls are needed they shall include:"

could be simply abbreviated to:

"Each manufacturer shall establish and maintain process control procedures to develop, conduct, control, and monitor production processes to ensure that a device conforms to its specifications. These controls shall include..."

And: "Compliance with specified reference standards or codes;" could much better worded as "Compliance with established criteria".

The present wording is almost begging for misinterpretation... :bonk:
 

kreid

Involved In Discussions
#5
For your point 1) Software as a Medical Device (SaMD). The manufacturing process should not/would not impact the design specification for software.
 

Mark Meer

Trusted Information Resource
#6
Good point! I hadn't considered software stand-alones. Although, I suppose writing software to installation disks could be considered a manufacturing process.

I still am not completely clear what this requirement is trying to accomplish that is not already covered by subsequent requirements (e.g. equipment control, acceptance activities).

The fact is that any manufacturing process has the potential that "deviations from device specifications could occur". So it's unclear to me what the burden of documentation is expected here...

Presently, we maintain:
- Manufacturing work-instructions
- Documented in-process acceptance at delineated stages
- Equipment control procedures (SOPs)
- Documented final acceptance criteria

To me - assuming the wording "standards" can be replaced with "criteria" - this would meet 820.70(a) requirements.

Am I missing something?
 
Thread starter Similar threads Forum Replies Date
J Interpreting Process Capability results and ratios Capability, Accuracy and Stability - Processes, Machines, etc. 1
P Interpreting the Process Gap - Open Non-Conformances & ISO Clause ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
Q Interpreting China Medical Device regulations/standards China Medical Device Regulations 1
P ISO 80369-7 standard - Interpreting which Parts should be in scope Other Medical Device Related Standards 7
K PACS - interpreting MDD and Borderline Guidance CE Marking (Conformité Européene) / CB Scheme 5
M List of Packaging Contents - Medical Devices - Interpreting EU Directive 93/42/EEC CE Marking (Conformité Européene) / CB Scheme 3
D Interpreting Normal vs Weibull Capabilities Capability, Accuracy and Stability - Processes, Machines, etc. 4
P Interpreting Span Measurement - P95-P5 Six Sigma 1
B Interpreting Deviations 5 & 6 in Annex ZA in ISO 14971:2012 ISO 14971 - Medical Device Risk Management 1
B Interpreting "misuse" when assessing Hazardous Situations ISO 14971 - Medical Device Risk Management 2
SteveK Interpreting clause 4.1.5 in ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 9
M Help interpreting 21 CFR Part 806 (corrections and removals) Other US Medical Device Regulations 1
G Interpreting Type 1 MSA (Measurement Systems Analysis) Results Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
M Help interpreting MIL-STD-105E Single Sampling Plans Tables Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
M Biocompatibility in Respiratory Products - Help interpreting whitepaper IEC 60601 - Medical Electrical Equipment Safety Standards Series 22
I Interpreting Product Realization (Clause 7) in ISO 9001:2008 for Service Industry ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
V GD&T Interpreting Datums in Two Single Segmented Position Tolerances Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
M Interpreting X bar and R Control Charts Statistical Analysis Tools, Techniques and SPC 2
B Help with interpreting stock market terminologies Coffee Break and Water Cooler Discussions 4
V Interpreting Zinc Plating Specification GMW 3044 - 6K96/48 APQP and PPAP 2
K Interpreting and Applying 7.3.2 Design and Development Inputs ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
S Interpreting the Calibration Report for my Dial Indicator General Measurement Device and Calibration Topics 3
B Length Measure - Interpreting Calibration Results and Measurement Uncertainty (MU) Measurement Uncertainty (MU) 3
S Interpreting Level of Confidence - Round Robin for Tensile Testing - Help needed Statistical Analysis Tools, Techniques and SPC 4
M Interpreting AS9100 Clause 8.2.2 Internal Audit Requirements Internal Auditing 28
T Interpreting my t test in Minitab Using Minitab Software 2
S Interpreting Linear Regression Results from Minitab Using Minitab Software 15
optomist1 Interpreting LMC for Pattern of Slots (GD&T) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
optomist1 Interpreting Minitab Test For Equal Variances Using Minitab Software 3
V Interpreting Minitab Gauge R&R Results Using Minitab Software 6
A ISO 2081 - Interpreting Coating Thickness Other ISO and International Standards and European Regulations 3
L Interpreting Injection Molding Tooling Documents - Cavities Manufacturing and Related Processes 3
C Interpreting Gage R&R Results - 3 operators, 3 iterations and 10 parts Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
M Interpreting Measurement Uncertainty for Temperature PRT probes Measurement Uncertainty (MU) 2
N Interpreting Hygrometer Calibration Uncertainty - Temperature Coefficient Measurement Uncertainty (MU) 1
mr500 Interpreting Decimals in the mm state - Off Wall Question Inspection, Prints (Drawings), Testing, Sampling and Related Topics 14
J Interpreting clause 7.5.2.1 (validation of software used in production & service) ISO 13485:2016 - Medical Device Quality Management Systems 2
E Interpreting Partial Least Square Results Using Minitab Software 1
C Interpreting Outside Laboratory Calibration Certificate Measurement Uncertainty (MU) 3
M Interpreting MIL-PRF-19500 (Performance Spec, Semicon Dev) - What is LTPD of 20 - ??? Manufacturing and Related Processes 4
T Rules for interpreting control charts Statistical Analysis Tools, Techniques and SPC 2
Q Interpreting Responsibility and Authority Clause 5.5.1 ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 18
H Interpreting 'Evaluation of compliance' in ISO 14001 Internal Auditing 4
P Micrometer Gage R&R study - Interpreting the data and suggestions Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
W Interpreting a Normal Probability Plot Statistical Analysis Tools, Techniques and SPC 4
Wes Bucey "Downsizing" ramifications - Evaluating and Interpreting the News World News 0
Q Interpreting calibration result of Air Particle Counter on Counting efficiency General Measurement Device and Calibration Topics 1
A Particle Counter Results - Interpreting Data for a Class 100,000 Cleanroom Other Medical Device Related Standards 12
M GD&T Q&A session - Interpreting FCF (Feature Control Frame) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 43
N TL 9000 - Help Interpreting Normalization Unit for Measurements TL 9000 Telecommunications Standard and QuEST 2
Similar threads


















































Top Bottom