Interval for Self-Calibrating Measurement Device?
Crusader
We have a gauge that self-calibrates - it even says so in the user manual! The gauge is set for annual calibrations by the manufacturer. It has expired now. The company has decided to not send out for calibration since this thing can self-calibrate. I'm in a pickle. I guess I can extend the calibration to next year based on past results (previous years of external calibrations).
What do you think? :confused: My 3 year recert is next year. I'm nervous that we'll get hit with a finding. I warned the company but I lost the battle.
Michael_M
Do you have a method of 'verification'? For example, we have an optical micrometer that does not require 'calibration', however, we have some XX gage pins (that are calibrated) which we use to verify the device is still measuring correctly.
Can you do something like this?
Crusader
It is self-calibrating, and I do not have a second gauge to check the piece that is used in the self-calibration. There really is no verification.
JoShmo
There is no such thing as "self calibration" and - I've encountered this before - the manufacturer doesn't understand "calibration" per the VIM. The best it is is a self verification. What are the units it's measuring?
Crusader
Are you able to tell us exactly what it is? Might help us trouble shoot for you.
It is a gloss meter - I attached a similar model owner manual. But the owner manual does state it is self-calibrating. I want to do the right thing and keep my employer happy and meet the ISO standard at the same time. And Thanks for replying!
Ron Rompen
A gloss meter 'self calibrates' by measuring the reflection from a known standard. The problem with this is the known standard becomes dull and less glossy over time and with use.
You can purchase a new 'standard' and use it as the annual comparison/verification one, and continue to use the original one for daily self-calibration (at least until it fails to match up with the 'new' standard).
I ran into exactly the same thing a few years ago, and this is how I dealt with it, and this is what my customer (and my auditor) approved.
BoardGuy
Just because a piece of equipment says it “self-calibrates” does not mean that this method assures that it is measuring correctly. ISO 9001, AS9100, and ISO 13485 clearly states that where necessary to ensure valid results, measuring equipment shall be calibrated or verified or both at specified intervals, or prior to use, against measurement standards… (7.6)
Your problem is how you verified that the measurement results from this piece of “self-calibrating” equipment are valid and what records you will retain showing you verified this equipment. Since you do not state what the equipment is its little hard to advice you on a method that could be use to verify the equipment but there is always a way.
BradM
Newer equipment/instrumentation have features/functions that allow for internal adjustment/alignment of the measurement capability. For example, newer analytical balances have internal standard weights that can be used to verify the balance is weighing correctly. However, there are many factors that can affect the accuracy/repeatability of a balance. Too, the internal weights on the balance need to be verified routinely.
Depending on how robust/ accuracy the "self-calibration" routine is on a piece of equipment can extend the calibration frequency/interval. However, for a true calibration of measurement equipment/instrumentation, a known standard with NIST (or equivalent) traceability is needed; and an assessment of uncertainty (or accuracy) ratio.
So if you don't want to pay an outside vendor to perform the verification, you should consider acquiring the standard yourself and determining what interval that standard should be re-verified. Also, you should have a procedure/work instruction written to address the needs.
J Documenting a Calibration Interval Extension Calibration Frequency (Interval) 7
S Absolute Interval measurement of CTS/CSE Waveforms IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S Requirements for Interval Measurement test & Frequency Response test clarification IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
V Using K-Factor(Tolerance Interval) Analysis for Design Verification Statistical Analysis Tools, Techniques and SPC 3
L 21CRF820.70(g)(2) Maintenance Inspection Interval 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
S Is there a difference in the process? Analyze residuals, construct interval estimate Using Minitab Software 2
S Significance of regression, correlation coefficient, hypothesesi, confidence interval Using Minitab Software 8
H Sample Size Calculation (Confidence Interval and Reliability) - Medical Devices Reliability Analysis - Predictions, Testing and Standards 6
A Calibration Interval - AS9100 - Control of Monitoring and Measuring Equipment Calibration Frequency (Interval) 8
R Control Chart Constants and Confidence Interval Statistical Analysis Tools, Techniques and SPC 7
D Torque Wrench Calibration Interval Extension with Non-Use? General Measurement Device and Calibration Topics 1
Marc Calibration Frequency (Interval) Discussions Sub-Forum General Measurement Device and Calibration Topics 0
K Calibration Interval based on Usage Calibration Frequency (Interval) 14
J Digital Scale Re-Calibration Interval Frequency General Measurement Device and Calibration Topics 10
M A Guide or Standard for Calibration Interval (Frequency) Calibration Frequency (Interval) 2
M Which Distribution & Analysis - Half hourly interval of ticket inflow Statistical Analysis Tools, Techniques and SPC 2
P One sided confidence interval for probability plot in Minitab Using Minitab Software 1
G Calculating the Fraction Defective Confidence Interval for a Lot Statistical Analysis Tools, Techniques and SPC 2
K How to Calculate Sample Size at 95% Confidence Interval on Surface Roughness Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
T Calculation of Confidence Interval for the Population Variance Statistical Analysis Tools, Techniques and SPC 2
D "As Found" Condition and Calibration Interval Changes Calibration Frequency (Interval) 5
B How many MSA studies are required and at what interval(s)? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
M Definition Interval - Definition of Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 4
F Rolled Throughput Yield and Confidence Interval Six Sigma 3
0 Constructing Confidence Interval of IQ Scores of Smokers vs. Non Smokers Using Minitab Software 4
J How to Write the Planned Interval to Conduct Internal Audit, Management Review, etc.. ISO 13485:2016 - Medical Device Quality Management Systems 3
B Calibration Interval of Munsell Color Charts General Measurement Device and Calibration Topics 3
C Furnace Classes in relation to AMS 2750 to determine Calibration Interval Calibration Frequency (Interval) 2
A How to Determine Machine Tool Calibration Interval? General Measurement Device and Calibration Topics 4
Z CMM (Coordinate Measuring Machine) Calibration Interval (Frequency) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
MarilynJ6354 Torque Wrench Calibration Frequency (Calibration Interval) Calibration Frequency (Interval) 38
C Formula for Calculating Confidence Interval at 90% Confidence Level Reliability Analysis - Predictions, Testing and Standards 3
M Confidence Interval on Normal Probability Plot Using Minitab Software 2
Q 95% Confidence Interval of a 20% Absentee Rate - Sample Exam Question Six Sigma 10
D Nonadjustable M&TE - 10 year calibration interval? Pin Gages Calibration Frequency (Interval) 11
P Confidence Interval for Production Sampling Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
Y Definition Confidence Interval - Definition of Confidence Interval Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 4
somashekar Calibration interval (Clause 7.6 ISO 13485:2003) - Biomedical test instruments ISO 13485:2016 - Medical Device Quality Management Systems 4
D Thermometer Calibration and Calibration Interval ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Management Review "planned interval" define please ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
T Standard Error vs. Confidence Interval (not confidence level) Statistical Analysis Tools, Techniques and SPC 1
K CI (Confidence Interval) for bias Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
A AS/EN 9100 Required Internal Audit Cycles / Interval Internal Auditing 12
A Measurement Uncertainty (MU) - Confidence Interval & Degrees of Freedom? Measurement Uncertainty (MU) 4
T Software & Consultation Sources needed for Interval Analysis/TAR/TUR. General Measurement Device and Calibration Topics 4
J What's your interpretation of Clause 7.6 - Calibration 'Interval'? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 16
B Gage R&R estimate of confidence interval Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
H What is Predictive Maintenance - MFTB Collection (downtime interval of equipment) IATF 16949 - Automotive Quality Systems Standard 1
L Need a help in finding a research paper - Patel, J. K. (1986) Tolerance interval Capability, Accuracy and Stability - Processes, Machines, etc. 2