Intrinsic Uncertainty and Accuracy of a Medical Electrical Safety Analyzer
A
akmal122g
#1
Hello. Can someone help me with this equipment specification. The equipment is a medical electrical safety analyzer. The brand is Gossen Metrawatt and the model is Seculife ST.
I am trying to find the accuracy of the equipment but I cannot find it in the manual. However, the equipment specifies intrinsic uncertainty in its specification. I enclosed the specification of the equipment. Is intrinsic uncertainty similar to accuracy?
Elsmar Forum Sponsor
dgriffith
Quite Involved in Discussions
#2
Re: Intrinsic uncertainty and accuracy
Just love the way language is different around the world--even scientific/technical terms . . .
I'm hazarding a guess that they mean relative accuracy, which does not account for the propagated accuracy/uncertainty of the standards in the traceability chain.
In other words, the accuracy the instrument is capable of if all standards used to calibrate it were perfect.
After re-reading the specifications, if it were me, I would call the MFR. or whoever sold it to you and ask to speak with the sales engineer. Ask them to explain what they actually mean when they say 'uncertainty'. These look more like accuracy specs than uncertainty.
My answer to your question, "is accuracy similar to uncertainty", is that they are related but are not the same. Accuracy reflects the expected closeness of agreement to the true value. Since the true value can never be known, the error in the measurement can also never be known. Uncertainty is an estimate of the range of values within which the true error lies. It can be shown that the uncertainty of the error is also, then, the uncertainty of the measurement. But it is not the accuracy.
Thread starter Similar threads Forum Replies Date
M Measurement Uncertainty in Optical Microscopy Measurement Uncertainty (MU) 1
L Uncertainty calculation for an average Measurement Uncertainty (MU) 1
R Determining Uncertainty from Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
F Missing value of Uncertainty B Measurement Uncertainty (MU) 10
F Issues in Uncertainty Estimation Measurement Uncertainty (MU) 5
F Operator Manual and Type B Uncertainty Measurement Uncertainty (MU) 3
F Learning Uncertainty Budget for Novice - Fluke 5520A Measurement Uncertainty (MU) 1
G Does pitch/increment/resolution of a ruled scale apply to measurement uncertainty as line item? Measurement Uncertainty (MU) 10
R Uncertainty in measurement larger than tolerance Measurement Uncertainty (MU) 2
S Uncertainty Budgets - Phase Noise (SSB), Harmonics Measurement Uncertainty (MU) 2
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
M Informational US FDA Final Guidance – Consideration of Uncertainty in Making Benefit-Risk Determinations in Medical Device Premarket Approvals, De Novo Classificati Medical Device and FDA Regulations and Standards News 0
A Uncertainty Budget - Externally produced sensor module Measurement Uncertainty (MU) 1
E Discussion between co-worker on tolerance and uncertainty and how to apply it. Thoughts? 17025 ISO 17025 related Discussions 1
G Assigning a calibration tolerance - An x-y coordinate machine - Uncertainty as my verification tolerance General Measurement Device and Calibration Topics 4
G 3 Operator Std. Dev. and uncertainty calculation Measurement Uncertainty (MU) 2
G Reporting measurement uncertainty for custom items Measurement Uncertainty (MU) 2
L Gage R&R for test precision or uncertainty prediction? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
G Uncertainty of staging a short line scale standard on longer measuring machine? Measurement Uncertainty (MU) 4
K Measurement Uncertainty Budget (MU) of for Ultrasonic Thickness gauge Measurement Uncertainty (MU) 1
G NIST SOP 29 - Assignment of Uncertainty - Question General Measurement Device and Calibration Topics 0
M Medical Device News FDA news - 05-09-18 - Draft - Uncertainty in Benefit-Risk Determinations Other US Medical Device Regulations 0
G Can Measurement Machine Bias be part of Uncertainty? General Measurement Device and Calibration Topics 1
G How to include machine error in uncertainty calculations? Measurement Uncertainty (MU) 5
D Difference between uncertainty and expanded uncertainty of measurement General Measurement Device and Calibration Topics 1
M Calibration certificates avowing accuracy beyond the uncertainty limits? Measurement Uncertainty (MU) 7
G ANAB ISO 17025 Accreditation - Uncertainty and systematic error? ISO 17025 related Discussions 6
G How many Uncertainty Measurements for Micrometers and Calipers? Measurement Uncertainty (MU) 3
G What degrees of freedom means in uncertainty budget and how do I set the value Measurement Uncertainty (MU) 2
G Uncertainty Experts - ISO 17025 - If the Distribution is Rectangular Measurement Uncertainty (MU) 2
G Uncertainty Budget Examples for Caliper, Micrometer and Dial Gauge Measurement Uncertainty (MU) 3
P Uncertainty and Significantly Different Means - Buffer Solutions Measurement Uncertainty (MU) 4
S Pressure Gauge Calibration Uncertainty Calibration and Metrology Software and Hardware 5
E Uncertainty of Mean - Test rig that measures water temperatures Measurement Uncertainty (MU) 3
J Internal Laboratory Scope and Measurement Uncertainty (MU) IATF 16949 - Automotive Quality Systems Standard 1
G Manual Procedure with Uncertainty Budgeting for either a Caliper or Micrometer General Measurement Device and Calibration Topics 6
C MU (Measurement Uncertainty) in an Internal Calibration Laboratory Measurement Uncertainty (MU) 1
N How to Convert Micro Volts into Temperature Uncertainty Measurement Uncertainty (MU) 1
N Include homogeneity uncertainty into budget table for thermocouples Measurement Uncertainty (MU) 1
J CMM Uncertainty Calculation Question - Deviations Measurement Uncertainty (MU) 9
M Uncertainty Budget for a Micrometer Measurement Uncertainty (MU) 5
C Uncertainty Trainer in Texas? Measurement Uncertainty (MU) 3
P Purpose of calculating Uncertainty value in calibration study Measurement Uncertainty (MU) 5
C Uncertainty Estimation for Spark Optical Emission Device Measurement Uncertainty (MU) 1
W Dimensional Measurement of 10% accuracy and 25% uncertainty standard General Measurement Device and Calibration Topics 6
B Measurement Uncertainty using Minitab Measurement Uncertainty (MU) 2
J Uncertainty budget for IR thermometry at high T - 1000 - 1500?C Measurement Uncertainty (MU) 1
N Understanding the absolute uncertainty specification for a Fluke 5500A Measurement Uncertainty (MU) 3
M Help with uncertainty calculation in the calibration of a Pyrogen Test Processor Measurement Uncertainty (MU) 7
Similar threads
Top Bottom