This is the fifth in a series of articles about MSA. The focus of this article will be on measurement repeatability and reproducibility commonly referred to as a gage R&R study.

This article will deal solely with the AIAG MSA methodology. The AIAG methodology is the methodology required by many customers, particularly in the automotive industry. Whether you agree with it or not, it is a standard approach and has widespread acceptance. Most suppliers have no option other than to comply. I will deal with other approaches in a later article.

This is where calibration completely separates from MSA. There is no equivalent in calibration to R&R. Calibration, bias studies, linearity studies and stability studies have all focused on measurement bias. R&R studies focus on measurement variation. Let’s first start with definitions. What is Repeatability? What is Reproducibility? What is an Operator by Part interaction?

Repeatability is the measurement variation observed when a single operator measures one part multiple times.

Reproducibility is the measurement variation observed when multiple operators measure one part multiple times. Depending on the measurement system, AND how the MSA study is designed Reproducibility may also be the measurement variation observed when multiple measurement stations or devices measure one part multiple times. For example, a measurement device may consist of a fully automated measurement device comprised of multiple stations. Each station is dimensionally unique and the difference contributes to measurement variation. Another example is multiple, fully automated measurement devices that measure the same characteristic. Each device has a slightly different measurement bias and contributes to the measurement variation. Yet another example is a semi-automated measurement device that is manually loaded. The resulting measurement is influenced by the manner in which the product is loaded into the fixture. Each operator that loads the product has a slightly different technique for loading that influences the measurement variation.

Operator by Part Interaction is a situation where the result of an operator’s measurement technique is influenced by the part itself. For example, two operators measure a shaft diameter using techniques that are identical in all respects except one. Operator A takes measurements at the midpoint of the shaft length. Operator B measures at one end of the shaft. Two shafts out of ten have burrs on the ends. Operator A’s measurements are not affected by the burr. Operator B’s measurements are affected by the burr. This will result in an interaction between the operator and the part itself.

Part Selection

The first step in an effective R&R study is to determine the use of the gage itself. Will it be used for part inspection to a tolerance, for process control, for statistical studies (e.g., a hypothesis test, capability study, DOE, etc.), or for a combination of these? This is very important because it influences the selection and quantity of parts needed for the R&R study.

If the gage is used solely for part inspection, the selection of parts is not critical because the part variation is not included in the calculation of the R&R metric, %Tolerance (i.e., P/T Ratio). Some will recommend that parts representing the full spread of the tolerance be used. While this does not hurt, it is not really necessary. If a gage linearity study has been performed, the change in bias over the tolerance spread is known. If a gage linearity study has not been performed and there is a linearity issue an R&R study will not detect it.

If the gage is used for process control or for statistical tests, the selection of parts is critical because the part variation is part of the calculation of the R&R metric, % Study Variation (i.e., %GRR). It is vital that the parts selected for the study reflect the actual variation of the process. That is, the StdDev of the parts equals the StdDev of the process. Some statistical packages, such as Minitab, allow the entry of the historical StdDev of the process. If your software has this option, use it, entering the process StdDev from a capability study or calculated from SPC charts. If the software does not have the feature, manual calculations using the historical value are still possible as follows"

% Study Variation = 100 * [StdDevR&R / StdDevTotal Variation]

StdDevTotal Variation = SQRT[StdDevR&R^2 + StdDevPart Variation^2]

Manually substitute the StdDev from a capability study for StdDevPart Variation

How many operators, trials and parts do I use?
The recommended standard is to have three operators measure ten parts three times each. Is this always the best approach? What flexibility do we have in modifying this? To answer this question, we need to look at how the data are used by the ANOVA calculations.

Source of Variation degrees of freedom (n-1)
Reproducibility (3 operators) 2
Parts (10 parts) 9
Pure Error (Repeatability) 78
Total Variation (90 measurements) 89

The 10/3/3 approach provides very good estimates of the total variation and the repeatability. The least reliable estimate of variation will be the Reproducibility because it has the smallest degrees of freedom. If concessions must be made, it is better to run fewer trials in order to maintain or increase the number of operators. The total number of measurements should be maintained near 90. The number of parts may be reduced, if (and only if) an independent estimate of part variation (such as from a capability study) is available and used as described in the previous section.

Selection of Operators
Always use the actual operators that will perform the measurement. Do not use personnel that will not perform the measurement task. Select the operators randomly. Do not handpick the best operators. If only one operator performs the measurement task (e.g., complex analytical equipment), perform the study with that operator only. There is no Reproducibility component in that situation

Measurement of Parts
Parts should be introduced randomly to each operator by an independent party that is not involved in the actual measurements. This is to prevent potential measurement bias caused by an operator remembering a previous measurement and consciously or unconsciously adjusting the next measurement to match.

Parts should be measured using the same method that will normally be used. If Reproducibility is adversely affected by the use of different methods, you need to know that. If there is significant within-part variation in form that adversely affects Repeatability, you need to know that also.

What method do I use?
In the MSA manual, there are two optional methods: the Range method and the ANOVA method. Both methods will provide very similar results. The Range method uses simpler math, but the ANOVA method can detect a potential Operator x Part interaction. If you have software available, use the ANOVA method. It provides additional information. The only compelling reason for using the Range method is if you must perform manual calculations.

continued in next blog entry
 
Last edited:
S

SPC_Newbie

Small error or am I making a bad assumption? - I'm getting ready to run first GR&R and am colating all of my notes so I want to be sure about part selection - "..Will it be used for part inspection to a tolerance, for process control, for statistical studies .."

Then we address the first portion: gages for part inspection: "If the gage is used solely for part inspection, the selection of parts is not critical ..."

Then, it makes sense to me we would address the following two options- gages for process control and/or statistical studies but the next sentence says part inspection whe I expected to see 'process control': "If the gage is used for part inspection or for statistical tests, the selection of parts is critical because..."

Sorry if I'm being nit-picky, this is an honest question, though I'm 'assuming' process control is what was meant there - Thanks!
 

tmayur21

Registered
If the gage is used for part inspection or for statistical tests, the selection of parts is critical because the part variation is part of the calculation of the R&R metric, % Study Variation (i.e., %GRR). It is vital that the parts selected for the study reflect the actual variation of the process. That is, the StdDev of the parts equals the StdDev of the process. Some statistical packages, such as Minitab, allow the entry of the historical StdDev of the process. If your software has this option, use it, entering the process StdDev from a capability study or calculated from SPC charts. If the software does not have the feature, manual calculations using the historical value are still possible as follows"


Hi! I believe this part of the writeup needs to be corrected. Shouldn't it be 'process control' instead of 'part inspection'?
 

Miner

Forum Moderator
Leader
Admin
Hi! I believe this part of the writeup needs to be corrected. Shouldn't it be 'process control' instead of 'part inspection'?
Finally! Someone caught that. I put that in to see if anyone was paying attention. :oops: Not really. Thank you for catching that. You are correct and I fixed it.
 

Wujinn

Registered
Hello,
that's great article!

I have a question related to % of Total Variation and % Study Variation.

I couldn't find %Study Variation in AIAG MSA 4th edition, maybe I'm just blind. Nonetheless, I don't undestand some of these concepts.

1) There are 2 methods that I can use: Tolerance method and Part to Part variation method and it influences my denominator in % of Total variation calcluation. Is % Study Variation the name that minitab always uses and it means that my % will be calculated with Part to Part method? From what I was able to find on minitab website: "The %study variation is calculated as the study variation for each source of variation, divided by the total variation and multiplied by 100.", so they use TV in denominator instead of tolerance and they just called it "%Study Variation"?
2) When I calcuate % Total Variation for EV, AV, GRR, PV and TV there is a thing I don't understand. Why GRR + PV is bigger than TV? For instance, my TV = 0,0575, but GRR + PV = 0,062.

GRR = EV + AV (I omit squre root and ^2)
PV = Rp x K3
TV = GRR + PV (I omit squre root and ^2)

It doesn't add up and I end up with a score greater than 100%, why is it structured this way?
 

Bev D

Heretical Statistician
Leader
Super Moderator
I’m sure MIner will chime in on Monday…I will say that you are doing the math wrong. The formulas are correct so you should follow them. When performing math on standard deviations you must work with the square root of the sum of the squares. This is a geometric property. Think of the measurement error and the part variation as the legs of a right triangle and the total variation as the hypotenuse. The length of the hypotenuse is equal to the square root of the (measurement error squared + the part variation squared).
 

Wujinn

Registered
Thank you. There is one thing that is difficult to understand to me.
When I want to calculate %Contribution I just do simple math and divide one number by another and in this way I'm able to determine %Contribution.
On the other hand %Study Var uses square root on the same formula, so it's not % anymore - name of it is a little bit confusing. It's more like a index. Please correct me if I'm wrong. 1684063620300.png

There is a point that I don't fully understand, why is it caluculated, since we have contribution in %? In other words what is the benefit of calculating it and why is it better than simple % contribution?
I would like to understand this. To be fair, this is not necessary for the study and I am aware of that. My approach to this is: to calculate something is one thing, and to understand why is another story. It's more about statistics at this point, but maybe somebody will be able to answer.
 

Miner

Forum Moderator
Leader
Admin
Bev is correct in her answer to your first question. Regarding your follow-up post, % Contribution is calculated using the Variance. This is mathematically correct to determine percentages that add up to 100% in total. %Study Variation and %Tolerance are calculated using Standard Deviations. This is not mathematically correct but is what is defined and used by AIAG. Therefore, the percentages do not add up to 100 percent for the reasons that Bev gave. Variances, which are the squares of the standard deviation, are additive, but standard deviations are not additive. If you are interested in a more in-depth explanation, see Dr. Donald Wheeler's articles on this topic.

The AIAG MSA reference manual even acknowledges this on the bottom of page 122 where it states, in bold print and all caps, THE SUM OF THE PERCENT CONSUMED BY EACH FACTOR WILL NOT EQUAL 100%, although they do not explain why this is true.
 
Thread starter Similar threads Forum Replies Date
Miner Intro to Measurement System Analysis (MSA) of Continuous Data – Part 5b: R&R Imported Legacy Blogs 14
Miner Intro to Measurement System Analysis (MSA) of Continuous Data – Part 4: Stability Imported Legacy Blogs 2
Miner Intro to Measurement System Analysis (MSA) of Continuous Data – Part 3: Linearity Imported Legacy Blogs 15
Miner Intro to Measurement System Analysis (MSA) of Continuous Data – Part 2: Bias Imported Legacy Blogs 4
Miner Intro to Measurement System Analysis (MSA) of Continuous Data – Part 1: What and Why? Imported Legacy Blogs 22
Miner Intro to MSA of Continuous Data - Part 10: Creating gauge families Imported Legacy Blogs 10
Miner Intro to MSA of Continuous Data - Part 9: %Tolerance for One-sided Specs Imported Legacy Blogs 3
Miner Intro to MSA of Continuous Data – Part 8: Comparison of Two Gages Imported Legacy Blogs 3
Miner Intro to MSA of Continuous Data – Part 7: R&R using Wheeler’s Honest Gage Study Imported Legacy Blogs 17
Miner Intro to MSA of Continuous Data – Part 6: R&R for Non-Replicable Measurements Imported Legacy Blogs 18
T Intro & ISO 9001:2006 Question - Is ISO 9001:2006 Released? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
E French Intro to Quality - Introducing ISO 9001 to New Employees Book, Video, Blog and Web Site Reviews and Recommendations 1
J Laser length measurement system General Measurement Device and Calibration Topics 3
K Alternatives to Metal Foil for touch current measurement of enclosures made of insulating material. IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M What should be measurement method in control plan if you are defining Control method as work instruction. Manufacturing and Related Processes 5
K ISO9001 - Measurement Equipment records ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
adztesla Specification (Significant digits) - Measurement (Medical device) ISO 13485:2016 - Medical Device Quality Management Systems 3
S How to report a measurement and relate it to calibration certificate General Measurement Device and Calibration Topics 1
B IATF 7.1.5.1.1 Measurement system analysis (Visual Inspection) IATF 16949 - Automotive Quality Systems Standard 3
B Leakage current measurement without isolating transformer IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
V Over 60 dimensions CMM measurement in control plan APQP and PPAP 8
C MSA for resistance measurement Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
F Multi-Sensor CMM Measurement General Measurement Device and Calibration Topics 5
P Type of MSA study according to measurement equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
CaliperJim Temperature and Humidity Effects on Measurement Devices General Measurement Device and Calibration Topics 6
E Audit Finding - Measurement of Process - Continuous Improvement - Trend Analysis Oil and Gas Industry Standards and Regulations 22
Rene Minassian Control of (TMME) Testing and Measurement Equipment Oil and Gas Industry Standards and Regulations 2
R DIGITAL MEASUREMENT DEVICE Manufacturing and Related Processes 2
R First Time Managing Calibration and Measurement System. Need Help. General Measurement Device and Calibration Topics 19
J Not able to repeat measurement at the same location General Measurement Device and Calibration Topics 7
Ron Rompen Convert Linear Measurement to Angular Measurement General Measurement Device and Calibration Topics 3
Q Resolution of Measurement Device for a 50' Coil Reliability Analysis - Predictions, Testing and Standards 1
Ron Rompen Assistance with perpendicularity measurement Inspection, Prints (Drawings), Testing, Sampling and Related Topics 10
G QSB+ : capability of measurement equipment IATF 16949 - Automotive Quality Systems Standard 15
B Measurement Calculations vs ASME Y14.5 Various Other Specifications, Standards, and related Requirements 2
S Measurement of threaded holes ISO10578 or ASME standard Measurement Uncertainty (MU) 3
S Comparing measurement results for a part from two ring gage General Measurement Device and Calibration Topics 3
B Building my own measurement equipment General Measurement Device and Calibration Topics 3
P Test Method Validation (TMV) for all Measurement Methods in Rec/Inspection Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
J Keyence Measurement Systems General Measurement Device and Calibration Topics 3
Q Measurement Equipment Revocation - Looking for a Disposal Form with Risk Assessment IATF 16949 - Automotive Quality Systems Standard 10
C IVD Metrological traceability without a reference measurement procedure General Measurement Device and Calibration Topics 11
E Any template/ form of Monitoring and Measurement of Processes and product to ISO 13485? ISO 13485:2016 - Medical Device Quality Management Systems 1
A Flexible "Ring" Roundness Measurement General Measurement Device and Calibration Topics 2
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
I QMS monitoring, measurement, analysis and evaluation requirement - Template ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
F Physical Part Measurement General Measurement Device and Calibration Topics 6
M Measurement Uncertainty in Optical Microscopy Measurement Uncertainty (MU) 1
B ISO 6508 and portable hardness measurement instruments General Measurement Device and Calibration Topics 0
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1

Similar threads

Top Bottom