IPCD Sterility test failure during MPQ (EO sterilization validation)

B

brianthermophilus

#1
Hello,

I could use some help in determining how to rationalize a recent IPCD failure during a EO sterilization validation.

Of the 3 half cycles run we had (1) IPCD that showed growth (non-contamination) that was located at the top of the pallet. The IPCD's were doubled up and the one beside it did not show growth. I've investigated the positioning of the IPCDs along with variables of the cycle and nothing seems to be out of the ordinary.

To avoid repeating an additional 3 half cycles, I'm looking to rationalize this (1) IPCD failure but unfortunately since it wasn't a result of contamination or an equipment issue I'm running into road blocks within ISO11135.

Has anyone had experience in IPCD failures during MPQ and talking your way out of it? Maybe I'm missing something?

Thank you!
 
Elsmar Forum Sponsor
M

MIREGMGR

#3
I don't think it's a good idea to write a rationale for why an issue is not a problem, when you don't understand what happened.

If it takes more testing, maybe with some variable exploration, to figure that out, you'll end up with a better-understood process in return for your additional investment of cycle costs and time.
 
Thread starter Similar threads Forum Replies Date
S Orthopedic implant sterility testing Medical Device and FDA Regulations and Standards News 1
D Sterility and BioBurden Testing for Plastics ISO 13485:2016 - Medical Device Quality Management Systems 5
B Determining sample size for device sterility Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
C Sterility and Bioburden Testing - LAL passed under 0.5EU/mL Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
H EO Sterilization Validation - Sterility Testing and Load Configuration Other Medical Device Related Standards 1
B Sterility test on gloves ISO 13485:2016 - Medical Device Quality Management Systems 3
M Medical Devices Sterility - Question from A Regulator Other Medical Device Regulations World-Wide 3
O Freeze-Dried Powder Sterility Test using the Membrane Filtration Technique Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
M How many Pieces to Test - Sterility Testing USP EU Medical Device Regulations 2
N Sterility Testing - USP Chapter 71 and ISO 11737-2 ISO 13485:2016 - Medical Device Quality Management Systems 5
H ISO 13485 for Sterility Only? ISO 13485:2016 - Medical Device Quality Management Systems 2
S Shelf Life Determination - Sterility test (Stasis test) ISO 13485:2016 - Medical Device Quality Management Systems 12
M Sterility Testing vs Bioburden Testing - Differences Qualification and Validation (including 21 CFR Part 11) 7
N Consultant Obligations, CFR 803 and Sterility Quality Manager and Management Related Issues 4
A Shelf Life Testing - We would like to extend the sterility expiration date Misc. Quality Assurance and Business Systems Related Topics 1
A Sterility Assurance Level? Special 510k US Food and Drug Administration (FDA) 5
D ETO Sterility Requirements for FIM Clinical Study in Europe Other Medical Device Regulations World-Wide 3
M GLP (Good Laboratory Practices) Requirement for Sterility Validations Qualification and Validation (including 21 CFR Part 11) 1
J Container Closure Integrity & Sterility Test - The Dye Immersion Method Quality Assurance and Compliance Software Tools and Solutions 16
B SAL (Sterility Assurance Level) for High Level Disinfection vs. Sterilization US Food and Drug Administration (FDA) 7
S Sterility - Class I or Class II Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
S Suitable Sterilization Cycle to achieve Sterility Assurance Level of <10-6 ISO 13485:2016 - Medical Device Quality Management Systems 1
D Sterility (bacteriostasis/fungistasis) Testing by Membrane Filtration validation Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
M Gamma Sterility Validation - ISO 11137-2, 4.2 vs. FDA Requirements ISO 13485:2016 - Medical Device Quality Management Systems 6
S Sterility Assurance Level of Liquids (SAL) - Special Culture Medium ISO 13485:2016 - Medical Device Quality Management Systems 3
shimonv Test Method Validation ISO 13485:2016 - Medical Device Quality Management Systems 10
C Test compulsory for Sutures? Manufacturing and Related Processes 6
dgrainger Informational DRAFT: The Medical Devices (Coronavirus Test Device Approvals) (Amendment) Regulations 2021 UK Medical Device Regulations 1
K Identification test in ph.eur. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 1
AllTheThings ISO-17025 accredited TEST lab performing internal calibrations (Yes: Here is how) ISO 17025 related Discussions 6
R Cost per test tool/formula Oil and Gas Industry Standards and Regulations 1
R Cost per test tool/formula Oil and Gas Industry Standards and Regulations 1
A R chart from ANOVA gage R&R test Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Burst Test Acceptability Criteria ISO 13485:2016 - Medical Device Quality Management Systems 2
A 60601-1 clause 8.11.3.6 Test Clarification IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E Test report to certify compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3
E Tensile test machine, tinius olsen 602 Manufacturing and Related Processes 1
M Test failure Root cause not found Customer Complaints 10
Moumen H Variations between ASTM A29 Standard for steel bars and Mill test certificates specs Manufacturing and Related Processes 1
D Tip test 10° -> 5° IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Sponsor name presented in the test report EU Medical Device Regulations 4
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 1
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Shelf Life Test for IFU and Implant Card Other Medical Device Related Standards 1
S Mechanical Test Under FDA Freedom of Information Act Medical Device and FDA Regulations and Standards News 5
S Critical characteristic on manufacturing operations test procedure Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S Knee Implant (Femoral -Cobalt chrome)-Sub chronic toxicity test (ISO 10993-11)choice of root Medical Device and FDA Regulations and Standards News 2

Similar threads

Top Bottom