IPCD Sterility test failure during MPQ (EO sterilization validation)

B

brianthermophilus

#1
Hello,

I could use some help in determining how to rationalize a recent IPCD failure during a EO sterilization validation.

Of the 3 half cycles run we had (1) IPCD that showed growth (non-contamination) that was located at the top of the pallet. The IPCD's were doubled up and the one beside it did not show growth. I've investigated the positioning of the IPCDs along with variables of the cycle and nothing seems to be out of the ordinary.

To avoid repeating an additional 3 half cycles, I'm looking to rationalize this (1) IPCD failure but unfortunately since it wasn't a result of contamination or an equipment issue I'm running into road blocks within ISO11135.

Has anyone had experience in IPCD failures during MPQ and talking your way out of it? Maybe I'm missing something?

Thank you!
 
Elsmar Forum Sponsor
M

MIREGMGR

#3
I don't think it's a good idea to write a rationale for why an issue is not a problem, when you don't understand what happened.

If it takes more testing, maybe with some variable exploration, to figure that out, you'll end up with a better-understood process in return for your additional investment of cycle costs and time.
 
Thread starter Similar threads Forum Replies Date
D Sterility and BioBurden Testing for Plastics ISO 13485:2016 - Medical Device Quality Management Systems 5
B Determining sample size for device sterility Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
C Sterility and Bioburden Testing - LAL passed under 0.5EU/mL Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
H EO Sterilization Validation - Sterility Testing and Load Configuration Other Medical Device Related Standards 1
B Sterility test on gloves ISO 13485:2016 - Medical Device Quality Management Systems 3
M Medical Devices Sterility - Question from A Regulator Other Medical Device Regulations World-Wide 3
O Freeze-Dried Powder Sterility Test using the Membrane Filtration Technique Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
M How many Pieces to Test - Sterility Testing USP EU Medical Device Regulations 2
N Sterility Testing - USP Chapter 71 and ISO 11737-2 ISO 13485:2016 - Medical Device Quality Management Systems 5
H ISO 13485 for Sterility Only? ISO 13485:2016 - Medical Device Quality Management Systems 2
S Shelf Life Determination - Sterility test (Stasis test) ISO 13485:2016 - Medical Device Quality Management Systems 12
M Sterility Testing vs Bioburden Testing - Differences Qualification and Validation (including 21 CFR Part 11) 7
N Consultant Obligations, CFR 803 and Sterility Quality Manager and Management Related Issues 4
A Shelf Life Testing - We would like to extend the sterility expiration date Misc. Quality Assurance and Business Systems Related Topics 1
A Sterility Assurance Level? Special 510k US Food and Drug Administration (FDA) 5
D ETO Sterility Requirements for FIM Clinical Study in Europe Other Medical Device Regulations World-Wide 3
M GLP (Good Laboratory Practices) Requirement for Sterility Validations Qualification and Validation (including 21 CFR Part 11) 1
J Container Closure Integrity & Sterility Test - The Dye Immersion Method Quality Assurance and Compliance Software Tools and Solutions 16
B SAL (Sterility Assurance Level) for High Level Disinfection vs. Sterilization US Food and Drug Administration (FDA) 7
S Sterility - Class I or Class II Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
S Suitable Sterilization Cycle to achieve Sterility Assurance Level of <10-6 ISO 13485:2016 - Medical Device Quality Management Systems 1
D Sterility (bacteriostasis/fungistasis) Testing by Membrane Filtration validation Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
M Gamma Sterility Validation - ISO 11137-2, 4.2 vs. FDA Requirements ISO 13485:2016 - Medical Device Quality Management Systems 6
S Sterility Assurance Level of Liquids (SAL) - Special Culture Medium ISO 13485:2016 - Medical Device Quality Management Systems 3
A Is calibration of test weight required General Measurement Device and Calibration Topics 1
S Non parametric test for semi-quantitative data. Statistical Analysis Tools, Techniques and SPC 5
Ed Panek Disaster Recovery Test IT (Information Technology) Service Management 1
A How much does a complete biocompatibility test package cost? Other ISO and International Standards and European Regulations 1
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Z Is Op Amp Common-Mode Rejection Ratio (CMRR) test applicable to battery operated devices? Other Medical Device Related Standards 0
E When to perform accelerated aging studies and transport test(OQ or PQ) Qualification and Validation (including 21 CFR Part 11) 4
U Does *anyone* know a lab that will test to EN 455-4 Medical Gloves shelf life determination? EU Medical Device Regulations 1
A Test requirements for marketing in US when RATED values are not of US IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
H Secondary Circuits, Voltage Test and Clearance/Creepage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S Fogging test as per DIN 75201(50ug/g ) VDA Standards - Germany's Automotive Standards 3
R HV Test of HF Surgical Unit IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E 60950-1 resistance to abnormal heat test Various Other Specifications, Standards, and related Requirements 0
S User evaluation for self monitoring blood glucose test systems US Food and Drug Administration (FDA) 4
E MSA for Push and Pull test Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
J Confusing on Harmonic Emissions/Voltage Fluctuations/Flicker test for CISPR 11 Class A Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Routine production tests - dielectric voltage withstand test IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
D Test Method Validation Qualification and Validation (including 21 CFR Part 11) 4
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Accelerated Aging - Creating test samples - Implantable medical device Question Other Medical Device Related Standards 4
E Mentor for Test Method Validation (TMV) Design and Development of Products and Processes 2
L ASQ's Biomedical Auditor Course Test ASQ - American Society for Quality 1
W Regarding Requirement for Hi pot & Leakage Current test Station Set-up IEC 60601 - Medical Electrical Equipment Safety Standards Series 6

Similar threads

Top Bottom