IQ and OQ for a "Dormitory" Refrigerator - Temperature Stability
Candi1024
We are writing an IQOQ for a standard little college dormitory style refrigerator. We store various materials in this refrigerator that need to be either simply "cool" or approximately 5-10 degrees C.
We seem to be going back and forth on how to verify the temperature is adequate. The main point of contention is wording.
Verify that the temperature is stable at 45degrees +- 5 degrees.
What does stable mean? How long does it need to be stable for? Each time we open the refrigerator to check the temperature, it will be less stable. There are no temperature readings built into this refrigerator, it is set with a dial ranging from 1 to 7.
I'm thinking we should be verifying on a regular basis that the temperature is still adequate. ...
Ninja
Looking for Reality
Re: IQOQ on "Dormitory" refrigerator
If all you need is "cool" or "approximately", why not check it each monday morning>
Is there significant damage caused if it is not "cool" or riding at 15C for the weekend?
A simple indoor - outdoor thermometer (with the sensor on a thin wire) could be hooked up on the outside with the wire run inside. Just monitor the temperature from the outside. Should it (the indoor -outdoor thermometer) need to be 'calibrated' just use a lab thermometer against it for 'calibration'. This type of setup could even be automatically recording if hooked to a digital-analog converter, this will feed through a spreadsheet into excel, we did this with our mixing room back in 2000 and it worked very well.
Ajit Basrur
Re: IQOQ for a "Dormitory" refrigerator - Temperature stability
You are mixing two terms - "Cold", and "Cool"
Cold :
Any temperature not exceeding 8° (46°F) is “cold.” A “refrigerator” is a cold place in which the temperature is maintained thermostatically between 2° and 8° (36° and 46°F).
Cool:
Any temperature between 8° and 15° (46° and 59°F) is “cool.”
v9991
from cGMPs point of view, the 'criticality' of its relevance would determine the extent&depth of controls required. ( with each of the qualification - calibration/performance-verification, routine monitoring etc)
viz., Dormitory refrigerator, used for storing chemicals - standards - materials ., etc
in essence, I would not like to stop at saying, "'risk based approach'.... based on the impact on cGMPs, etc., " but emphasize the -pre-definition of the approach, and then 'reference of data' based on which the relevant decision is arrived & documented.
Following references could provide certain inputs..., in assessing the relevance and requirements.
Elements to Consider When Qualifying Control Environmental Chambers Used for CGMP Applications
Qualification of temperature-controlled storage areas
temperature mapping protocol...
http://lib3.dss.go.th/fulltext/Journal/analyst/Analyst1998/no.9/1988v123n9p1879-1886.pdf
here's quick ref. to definitions of storage conditions.. USP34-NF29 General Notices.pdf
10.30.20. Cold
10.30.30. Cool
Any temperature between 8 ° and 15° (46° and 59°F) is“cool.” An article for which storage in a cool place is directed may, alternatively, be stored and distributed in a refrigerator, unless otherwise specified by the individual monograph.
D Laboratory Refrigerator Validation ISO 13485:2016 - Medical Device Quality Management Systems 2
K ISO 15189, QA & how to calibration for Balance, refrigerator and centrifuge? Qualification and Validation (including 21 CFR Part 11) 5
S Work Instruction template for Automotive Component (Refrigerator) Manufacturing and Related Processes 11
S Temperature Calibration for a Storage Refrigerator General Measurement Device and Calibration Topics 5
F Laboratory Refrigerator Validation and Chamber Temperature Distribution Qualification and Validation (including 21 CFR Part 11) 17
W Drive In / Walk In Freezer and Refrigerator Chamber Validation Qualification and Validation (including 21 CFR Part 11) 1
C Refrigerator/Freezer MU - HOW? Measurement Uncertainty (MU) 2
Marc Shot duck survives 2 days in refrigerator Coffee Break and Water Cooler Discussions 33
S Why does GAGEpack drop the temperature and humidity readings out of the verification event card? Quality Assurance and Compliance Software Tools and Solutions 0
CaliperJim Temperature and Humidity Effects on Measurement Devices General Measurement Device and Calibration Topics 6
H Regarding external surface temperature limit requirement for lab incubators. CE Marking (Conformité Européene) / CB Scheme 2
J Japan registration need manual without temperature reading Japan Medical Device Regulations 0
D Conductivity Water Samples testing temperature Qualification and Validation (including 21 CFR Part 11) 2
J Warehouse temperature and humidity control EU Medical Device Regulations 7
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
C One Time Service Supplier - Temperature and Humidity Testing Service ISO 13485:2016 - Medical Device Quality Management Systems 5
R SAT (System Accuracy Test) temperature according to AMS2750E AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W Insulation requirements of BF type infant skin temperature probe IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
B Temperature loop: partial calibration General Measurement Device and Calibration Topics 10
D ASTM C1064 Temperature Measuring Device General Measurement Device and Calibration Topics 2
C ETO Sterilised Class II Medical Device - Required Temperature Storage ISO 13485:2016 - Medical Device Quality Management Systems 1
M Microscope slide warmer temperature control General Measurement Device and Calibration Topics 4
S ISO 18223-2 Temperature symbol in the Legend table in the Instructions for Use Other ISO and International Standards and European Regulations 4
B ASTM F2924-14 /ASTM F3001 - Room temperature classification criteria Other US Medical Device Regulations 1
O Selection of a calibrator (pressure generator, temperature) for my field devices General Measurement Device and Calibration Topics 2
shimonv Symbol for temperature limit EU Medical Device Regulations 6
M Pharmaceuticals Warehouse temperature control (room temperature) Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
T 5ºC Storage Temperature Validation - Urology disposable devices EU Medical Device Regulations 3
Q Weigh Scale Electronics - Suggested Storage Humidity and Temperature Manufacturing and Related Processes 7
M CQI-9 monitoring of temperature in die Various Other Specifications, Standards, and related Requirements 2
C ISO 11607-1 Packaging System - Temperature and shipping conditioning Other Medical Device Related Standards 4
G Calibration Laboratory Dimensional Temperature Question General Measurement Device and Calibration Topics 2
V Temperature/humidity requirements in internal metrology laboratory General Measurement Device and Calibration Topics 3
P Qualification of a Temperature Controlled Unit have to be empty? Qualification and Validation (including 21 CFR Part 11) 5
D What is the exact Temperature Coefficient Value to be set on a Conductivity Meter? General Measurement Device and Calibration Topics 2
K Temperature Uniformity Survey (TUS) - Vacuum Furnace General Measurement Device and Calibration Topics 1
R Saudi Arabia - Temperature Storage Requirements Other Medical Device Regulations World-Wide 4
H Performing a TUS on a Furnace used at One Temperature - AMS2750-E AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 27
D Sterile Medical Device Temperature and Humidity (Storage) Labeling Requirements EU Medical Device Regulations 5
R Temperature Profile for Transport Simulation (Sterile Eye Drops) Other Medical Device Related Standards 1
N How to Convert Micro Volts into Temperature Uncertainty Measurement Uncertainty (MU) 1
N Calibrate temperature on refrigerated Centrifuges? General Measurement Device and Calibration Topics 2
R Medical Device Storage Temperature Tests and Requirements EU Medical Device Regulations 14
R NADCAP AC7108 - Tank Ambient Room Temperature Requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S What is the service temperature of Polysurlyn Laminated Aluminium Alloy 3105? Manufacturing and Related Processes 2
C Saboraud's Agar Incubation Temperature and Validation Requirements Other Medical Device and Orthopedic Related Topics 10
N Load Thermocouple requirement for TUS (Temperature Uniformity Survey) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
B Do I need to perform Gage R&R on a Digital Temperature & Humidity Gage Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 18