Irritation test - Dermal Filler - 52-week implantation study

#1
All,

For the Irritation test, it is common to observe the erythema / edema for 3 days after injection of the filler into the rabbit according to the ISO10993-10 guidelines.

In the case of high viscoelastic materials such as our products, the initial irritation is severe and long-term. Therefore, it is known that the period of irritation is measured by adjusting the test period and observing for 2 to 4 weeks.

In the 2017 FDA-approved Teoxane RHA4 filler, no irritation was observed 26 days after injection and a 52-week implantation study confirmed the onset of degradation.

Our nonclinical development department commented that our products could also be requested for data on systemic toxic effects during irritation if the irritation result is similar to RHA4 and lasts for 4 weeks and then decreases It is judged.

The nonclinical development team comments that our products can also be requested for data on local and system effects during irritation if the irritation results are similar to RHA4 for 4 weeks and then decreased.

In addition, from the results of our product's efficacy test, which remained without size change for more than 6 months under the skin, we need an additional 52-week implantation study to confirm the degradation.

We are planning to prepare a dossier for CE certification.

Would it be better to do the 52-week implantation study to moniter the degradation? (Kindly share any regulation or guideline for me)

Thank you for your help in advance.
 
Elsmar Forum Sponsor
Thread starter Similar threads Forum Replies Date
M ISO 10993 Labs which can test Cytotoxicity, Sensitizaton and Irritation Other Medical Device Related Standards 10
I ISO 10993-10:2010 Guidance Document for Irritation and Sensitization Test - Changes ISO 13485:2016 - Medical Device Quality Management Systems 3
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Accelerated Aging - Creating test samples - Implantable medical device Question Other Medical Device Related Standards 4
E Mentor for Test Method Validation (TMV) Design and Development of Products and Processes 2
L ASQ's Biomedical Auditor Course Test ASQ - American Society for Quality 1
W Regarding Requirement for Hi pot & Leakage Current test Station Set-up IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R Regulatory Affairs Certification (RAC) prep - Practice test versus actual exam Professional Certifications and Degrees 0
R SAT (System Accuracy Test) temperature according to AMS2750E AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
R MSA for ATE (Automatic Test Equipment Embedded Software) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
D Test summary report example for design validation wanted - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
S Documenting Design Verification Test Results (ISO 9001) Design and Development of Products and Processes 1
A Earth leakage current test for Permanently installed Large scale MEE IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A Bench test report translation to English for 510(k) submission Other US Medical Device Regulations 2
J Production Line Test Brasil - High Voltage Dielectric Strenght Test Other Medical Device Regulations World-Wide 5
W Seeking Guidance Verification Test Strategy for Class B Medical Devices IEC 62304 - Medical Device Software Life Cycle Processes 1
J 510(k) for a control kit for an external IVD test kit 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
K Cord anchorage pull test IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Minitab assistant: 1-Sample % Defective Test Using Minitab Software 1
N Is capability applicable for a destructive test? Capability, Accuracy and Stability - Processes, Machines, etc. 9
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Test method validation - Is MSA (MSA1, MSA2, MSA3 and linearity) a good solution? Medical Device and FDA Regulations and Standards News 1
B Operator protection - When to apply table 7, Dielectric strength test voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Medical Device Airworthiness, DO-160 Test Planner IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J Audit Finding For Not Retaining Test Results ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
W ASTM F1929 dye penetration test - Validation for in-house testing ISO 13485:2016 - Medical Device Quality Management Systems 13
S IVF medical devices standards and test methods Medical Device and FDA Regulations and Standards News 4
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
N Heavy tare to simulate high load in a routine sensitivity test of a floor scale? General Measurement Device and Calibration Topics 3
bryan willemot Looking for a certified test report for aerospace parts AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
A Information about Medical Device Test Laboratories Other Medical Device and Orthopedic Related Topics 4
B Standard for Dielectric Strength Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Definition of customer in ITP (Inspection & Test Plan) Contract Review Process 3
J EN61010 Electrical Withstand Test IEC 62304 - Medical Device Software Life Cycle Processes 0
Z MDD ER 7.5, "phthalate test"? Aspiration tips EU Medical Device Regulations 3
G ISO 17025:2017 Uninterrupted chain of traceability for test laboratories ISO 17025 related Discussions 6
A Single use non-sterile syringe used in the oral cavity - Laboratory test advice US Food and Drug Administration (FDA) 7
M Proper method to test flow of a regulator Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
C CQA Test Questions Bank General Auditing Discussions 1
G Is impact test according to IEC 60601-1 applicable for HAND-HELD equipment if also classified as BODY-WORN? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Watchcat Identification of Test Sample in Test Reports? Design and Development of Products and Processes 22
S Safety Class A with no test? IEC 62304 - Medical Device Software Life Cycle Processes 1
A IEC 60601-2-2 - Defib-proof test amendment to base standard IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K The correct way of doing a 2 Sample T test Reliability Analysis - Predictions, Testing and Standards 7
J Good Documentation Practices Self Test? ISO 13485:2016 - Medical Device Quality Management Systems 6
B EU: Alcohol test kits for breast milk IVDD EU Medical Device Regulations 2
D IEC EN 61326 Dated Standards - Dated standards for test protocols CE Marking (Conformité Européene) / CB Scheme 3
Kuldeep Singh Electromagnetic Compatibility Test IEC 60601-1-2: 2014 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7

Similar threads

Top Bottom