Irritation test - Dermal Filler - 52-week implantation study
E
elleseo
#1
All,
For the Irritation test, it is common to observe the erythema / edema for 3 days after injection of the filler into the rabbit according to the ISO10993-10 guidelines.
In the case of high viscoelastic materials such as our products, the initial irritation is severe and long-term. Therefore, it is known that the period of irritation is measured by adjusting the test period and observing for 2 to 4 weeks.
In the 2017 FDA-approved Teoxane RHA4 filler, no irritation was observed 26 days after injection and a 52-week implantation study confirmed the onset of degradation.
Our nonclinical development department commented that our products could also be requested for data on systemic toxic effects during irritation if the irritation result is similar to RHA4 and lasts for 4 weeks and then decreases It is judged.
The nonclinical development team comments that our products can also be requested for data on local and system effects during irritation if the irritation results are similar to RHA4 for 4 weeks and then decreased.
In addition, from the results of our product's efficacy test, which remained without size change for more than 6 months under the skin, we need an additional 52-week implantation study to confirm the degradation.
We are planning to prepare a dossier for CE certification.
Would it be better to do the 52-week implantation study to moniter the degradation? (Kindly share any regulation or guideline for me)
Thank you for your help in advance.
Elsmar Forum Sponsor
Thread starter Similar threads Forum Replies Date
M ISO 10993 Labs which can test Cytotoxicity, Sensitizaton and Irritation Other Medical Device Related Standards 10
I ISO 10993-10:2010 Guidance Document for Irritation and Sensitization Test - Changes ISO 13485:2016 - Medical Device Quality Management Systems 3
E Test report to certify compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 2
E Tensile test machine, tinius olsen 602 Manufacturing and Related Processes 1
M Test failure Root cause not found Customer Complaints 10
Moumen H Variations between ASTM A29 Standard for steel bars and Mill test certificates specs Manufacturing and Related Processes 1
D Tip test 10° -> 5° IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Sponsor name presented in the test report EU Medical Device Regulations 4
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 0
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Shelf Life Test for IFU and Implant Card Other Medical Device Related Standards 1
S Mechanical Test Under FDA Freedom of Information Act Medical Device and FDA Regulations and Standards News 5
S Critical characteristic on manufacturing operations test procedure Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S Knee Implant (Femoral -Cobalt chrome)-Sub chronic toxicity test (ISO 10993-11)choice of root Medical Device and FDA Regulations and Standards News 2
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Non parametric test for semi-quantitative data. Statistical Analysis Tools, Techniques and SPC 5
Ed Panek Disaster Recovery Test IT (Information Technology) Service Management 3
A How much does a complete biocompatibility test package cost? Other ISO and International Standards and European Regulations 1
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Z Is Op Amp Common-Mode Rejection Ratio (CMRR) test applicable to battery operated devices? Other Medical Device Related Standards 0
E When to perform accelerated aging studies and transport test(OQ or PQ) Qualification and Validation (including 21 CFR Part 11) 4
U Does *anyone* know a lab that will test to EN 455-4 Medical Gloves shelf life determination? EU Medical Device Regulations 1
A Test requirements for marketing in US when RATED values are not of US IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
H Secondary Circuits, Voltage Test and Clearance/Creepage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S Fogging test as per DIN 75201(50ug/g ) VDA Standards - Germany's Automotive Standards 3
R HV Test of HF Surgical Unit IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E 60950-1 resistance to abnormal heat test Various Other Specifications, Standards, and related Requirements 0
S User evaluation for self monitoring blood glucose test systems US Food and Drug Administration (FDA) 4
E MSA for Push and Pull test Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
J Confusing on Harmonic Emissions/Voltage Fluctuations/Flicker test for CISPR 11 Class A Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Routine production tests - dielectric voltage withstand test IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
D Difference between Test Method Validation and Gage R&R Qualification and Validation (including 21 CFR Part 11) 18
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Accelerated Aging - Creating test samples - Implantable medical device Question Other Medical Device Related Standards 4
E Mentor for Test Method Validation (TMV) Design and Development of Products and Processes 2
L ASQ's Biomedical Auditor Course Test ASQ - American Society for Quality 1
W Regarding Requirement for Hi pot & Leakage Current test Station Set-up IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R Regulatory Affairs Certification (RAC) prep - Practice test versus actual exam Professional Certifications and Degrees 1
R SAT (System Accuracy Test) temperature according to AMS2750E AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
R MSA for ATE (Automatic Test Equipment Embedded Software) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
D Test summary report example for design validation wanted - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
S Documenting Design Verification Test Results (ISO 9001) Design and Development of Products and Processes 1
A Earth leakage current test for Permanently installed Large scale MEE IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
Similar threads
Top Bottom