Is Aeration Process in ETO Sterilization Validation Required?

A

ANDY-QA

#1
Dear ALL

Really need your help.

1. Aeration process in ETO sterilization validation it is a must?

2. Can we exclude this aeration process if we can verify by doing ETO residual that our ETO level for product less than 4mg as required by ISO 10993-7?

3. Our ETO chamber having function flushing which is the process purpose likely same as aeration. Can we consider this flushing as aeration?

4. Aeration room or in chamber what is the parameter we need to control to said that this process is validated and can rely on it

TQVM
 
Elsmar Forum Sponsor
T

thnksnow247

#2
My experience is with sterilization of a single use polymer based product.

You need to validate that after a set period of time ETO levels in your product have dropped to acceptable levels. The reality is that you will need some period of time to get samples to the lab and perform the eto residual testing, this time period would define the minimum aeration time, because you haven't verified ETO levels at an earlier time period. You may also need to used the requirement for maximum ETO surface area concentration from 10993-7.

In summary, you want some defined aeration time, this can be short if you are able to validate it.

ETO aeration rate is influenced by temperature and fresh air flush rates. Elevated temperatures are typically used to increase aeration, though a controlled room temperature may work.
 
A

ANDY-QA

#3
Dear thnksnow247

Thank for the response.

1. Actually for aeration process, we already set it 48H but i can't find any document that mention how this days is choose.

2. From your reply you mean by doing the ETO residual we can actually know the minimum time of aeration, right?

3. ...maximum ETO surface area concentration from 10993-7.- This one you mean the calculation in PPM right for ETO residual?
 

planB

Super Moderator
#4
ANDY-QA,

adding to what thnksnow247 wrote:

1. Aeration process in ETO sterilization validation it is a must?
Aeration is an integral part of the overall sterilization process, which has to be validated. So yes, you have to validate except you want to follow ISO 11135:2014, Annex E.2.4 i) and verify each batch on its own merits. However, this is definitely more tedious than validating EO residuals during an initial sterilisation validation.

2. Can we exclude this aeration process if we can verify by doing ETO residual that our ETO level for product less than 4mg as required by ISO 10993-7?
Yes, you can - in case you have a limited-exposure device without any surface contact. Otherwise, additional EO limits apply. Do not forget to also consider the applicable limits for ECH.

3. Our ETO chamber having function flushing which is the process purpose likely same as aeration. Can we consider this flushing as aeration?
Per ISO 11135:2014, section 3.15, flushing typically refers to the active removal of EO from the sterilisation chamber by applying pressure differentials. But yes, flushing can also be used to aerate the product.

4. Aeration room or in chamber what is the parameter we need to control to said that this process is validated and can rely on it
Refer to ISO 11135:2014, sections 9.5.4.g)&h) and 10.2.n)&o)

3. ...maximum ETO surface area concentration from 10993-7.- This one you mean the calculation in PPM right for ETO residual?
Refer to ISO 10993-7:2008, sections 4.3.5.2 and 4.3.5.3

HTH, Gerhard
 
A

ANDY-QA

#5
TQVM for the reply Gerhard,

Aeration is an integral part of the overall sterilization process, which has to be validated. So yes, you have to validate except you want to follow ISO 11135:2014, Annex E.2.4 i) and verify each batch on its own merits. However, this is definitely more tedious than validating EO residuals during an initial sterilisation validation.
- this annex can't find in the standard "ISO 11135:2014, Annex E.2.4 i" it is right?


Yes, you can - in case you have a limited-exposure device without any surface contact. Otherwise, additional EO limits apply. Do not forget to also consider the applicable limits for ECH.

- yes you mean if we can confirm during sterilization validation the EO or ECH residual are below the limit mean aeration can exclude. Or it reflected to your first answer which if we not validate the aeration process we need to verify each batch on its own merits. One to ask, how many day the aeration process need to be done is base one ETO/ECH residual curve decay right?

Per ISO 11135:2014, section 3.15, flushing typically refers to the active removal of EO from the sterilisation chamber by applying pressure differentials. But yes, flushing can also be used to aerate the product.
- if after flushing ETO/ECH residual immediately can confirm below limit during sterilization validation so no need aeration process?


Refer to ISO 11135:2014, sections 9.5.4.g)&h) and 10.2.n)&o)
- Noted but surprisingly no need to monitor ETO level that aerate out during aeration process to confirm ETO level below 5ppm (standard expose limit below 1 h)


Refer to ISO 10993-7:2008, sections 4.3.5.2 and 4.3.5.3
-Noted TQ

HTH, Gerhard
 

somashekar

Staff member
Admin
#6
Keeping the standards aside and looking more into the process, the aeration is a vital process in the ETO cycle.... Why.
Because ethylene oxide (EO) gas is toxic to humans, restrictions have been imposed on its use for sterilization, specifying allowable levels of residual EO remaining in sterilized apparatus and materials.
So what is the material of your device in sterilization ? Does it absorb and leach later like PVC material ?
The aeration time that optimizes the removal of the remaining EO when a rigid sterilizing container is used for a vessel had not been identified.
So aeration process requires validation in ETO cycle.

Some food for thought
 
A

ANDY-QA

#7
Thank Somashekar for the comment

Few series of ETO/ECH residual testing we done, our product was achieve below limit at day 1. So, aeration still required?
 

planB

Super Moderator
#8
ANDY-QA,

in the end you have to demonstrate that your product's residual levels _consistently_ and _reproducibly_ comply to ISO 10993-7 within the (worst-case) tolerances of your process definition. In this respect, I would consider performing "a few series of ETO/ECH residual testing" as exploratory feasibility study, but not a validation per the standard requirements.

Refer also to ISO 11134:2014, section D.6.2.1b), which says
Residues of EO and its reaction products can be hazardous. It is essential for the manufacturer of the product to be sterilized to be aware of the possible occurrence of residues in the product. Temperature, dwell time, forced heated air circulation, load characteristics, product and packaging materials all affect the efficiency of aeration, and the set points and tolerances should be taken into account when evaluating residual levels as outlined in ISO 10993-7. Aeration can be performed within the sterilizer, in a separate area(s), or in a combination of both. [...]. Inadequately aerated items and packaging will release EO, which can injure patients and health care facility personnel.
HTH,

Gerhard
 
Thread starter Similar threads Forum Replies Date
A Validation of Forced Aeration Process ISO 13485:2016 - Medical Device Quality Management Systems 3
D Ethylene Oxide Sterilization aeration Manufacturing and Related Processes 1
A Aeration Room Requirements for EtO Sterilization Other US Medical Device Regulations 3
B How to satisfy clause 5.7.1.5 process validation for valve production API 6D Oil and Gas Industry Standards and Regulations 13
C Change Management Process ISO 14971 - Medical Device Risk Management 8
B Documents for production process IATF 16949 - Automotive Quality Systems Standard 8
J Process Capability - Determining the process capability of certain equipment Lean in Manufacturing and Service Industries 6
Z E-Coating process International Specifications? FMEA and Control Plans 3
W CQI-15 for new welding process IATF 16949 - Automotive Quality Systems Standard 3
D Using electronic lab notebooks in the design control process ISO 13485:2016 - Medical Device Quality Management Systems 3
M Process & Product Characteristic in Control Plan FMEA and Control Plans 17
S Consolidated Analysis Report for Moulding Process Manufacturing and Related Processes 1
P ISO 11607-2 Process Specification Other Medical Device Related Standards 1
Q Best off presentation of process flows Process Maps, Process Mapping and Turtle Diagrams 16
Q Logistics process flow ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
M Process FMEA for Laser Cutting/Press Brake/Welding FMEA and Control Plans 2
H Control of Change i.e. process and WI IATF 16949 - Automotive Quality Systems Standard 9
F Is this misuse of the MRB process? Manufacturing and Related Processes 7
L Product and process Deviation procedure Manufacturing and Related Processes 1
Ch00Ch00 How to evaluate a process ? Document Control Systems, Procedures, Forms and Templates 3
M Implants made by Additive Manufacturing process approved by USFDA Other Medical Device and Orthopedic Related Topics 1
Q Process Key figures and reporting ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
Ch00Ch00 Evaluate a process - Issue number Lean in Manufacturing and Service Industries 7
F Rules of process outsourcing in China China Medical Device Regulations 1
B Process / Procedure - Radiographic (X-Ray) Non-Film Document Control Systems, Procedures, Forms and Templates 0
P Certification process for registered device China Medical Device Regulations 12
J NCR- Failure of contract review process - NADCAP audit AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
Melissa Process Validation of Rotary Heat Sealer Speeds Design and Development of Products and Processes 4
C Validation of process for releasing the UDI EU Medical Device Regulations 4
Melissa Risk Management Process, How far do I need to go? ISO 14971 - Medical Device Risk Management 13
C Validation of process for production and servicing 5.7.1.5 API Spec Q1, 9th Edition Oil and Gas Industry Standards and Regulations 9
Z IA Construction Procurement Process Internal Auditing 3
H Risk Management Plan in agile process ISO 14971 - Medical Device Risk Management 14
DuncanGibbons Process flow & PFMEA for production planning and simulation activities? Process Maps, Process Mapping and Turtle Diagrams 7
N Help with understanding Process Controls Manufacturing and Related Processes 7
M VDA 6.3 – Workshop for Certified Process Auditor VDA Standards - Germany's Automotive Standards 4
C Process Consistency Manufacturing and Related Processes 5
I PFD (Process Flow Diagram) approach. Setup details as part of PFD? APQP and PPAP 5
K Business Process Flowchart Process Maps, Process Mapping and Turtle Diagrams 2
L Looking for Control Plan and FMEA Stamping process FMEA and Control Plans 2
J License renewal process in Iran Other Medical Device Regulations World-Wide 0
J Definition Outsourced process - Clear definition - 13485 Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 5
S IS0 13485 process flows ISO 13485:2016 - Medical Device Quality Management Systems 2
H Verification Process for the existing MDS Product IEC 62304 - Medical Device Software Life Cycle Processes 5
C Revalidation of Process Equipment for Equipment Transfered to New Facility. ISO 13485:2016 - Medical Device Quality Management Systems 5
Q Process map Evaluation and Analysis Method Process Maps, Process Mapping and Turtle Diagrams 5
S In process inspection Manufacturing and Related Processes 2
S Brexit 100% inspection during in process inspection Manufacturing and Related Processes 11
A Complaint review as part of the complaint handling process? ISO 13485:2016 - Medical Device Quality Management Systems 3
mustomutlu Process Validation Final Report Other Medical Device and Orthopedic Related Topics 2

Similar threads

Top Bottom