Is Biocompatibility study performed on medical devices for possible contaminants?

Ajit Basrur

Staff member
Admin
#1
As part of the molding activities, we use Compressed Air that is tested routinely per ISO 8573 requirements.

During a recent customer audit, we were asked if we have done a Biocompatibility study on the devices for the possible compressed air contaminants like Oil, Particles and Water ?

Our reasoning was that since the Compressed Air meets the requirements of ISO 8573, that should be fine but the customer insisted on the Biocompatibility testing.

I havent heard of it and would like to hear from others if this has to be performed ?
 
Elsmar Forum Sponsor
M

MIREGMGR

#2
Re: Is Biocompatibility study performed on devices for possible contaminants ?

I haven't heard of such a requirement, but our devices are all external- or non-patient-contact.

I've never seen such a requirement mentioned in a guidance.

I do understand how your customer, if they're especially rigorous or paranoid, might decide they need that qualification for an implantable or fluid-communicating part. After the metal-on-metal-joint-implant nano-metal-particle fiasco, I've been expecting an industry evolution toward requirements for proof of no nano-metal presence, especially for implantables.
 

Ronen E

Problem Solver
Staff member
Moderator
#3
Re: Is Biocompatibility study performed on devices for possible contaminants ?

Biocompatibility testing is not intended for fault conditions. What if the material itself, for some odd reasons, is not what it is specified to be?... Should one test for biocompatibility of all materials that can remotely be there?... It has no end.

Any risk should be managed. If the client is not happy with the current measures and controls, this is negotiable. However, you should be the one having the final say on how to mitigate. For instance, more filtering and tighter monitoring can be implemented if air contamination is suspected.

My take is that your client's behaviour is non-rational. Now it's your job to get them the ladder to climb off that tree :)

Cheers,
Ronen.
 

Ajit Basrur

Staff member
Admin
#4
Re: Is Biocompatibility study performed on devices for possible contaminants ?

Thanks MIREGMGR and Ronen.

My thinking has been on the same lines as yours. Let me see how far I go with our customer :)
 

kuyakut

Involved In Discussions
#5
During the initial stage of molding for a certain fluid path components, we were advise to do biocompatibility since we are using a anti - static air blower on that certain parts. But after we proved to them that we have upfront measures and controls that the air is not contaminated.The customer agreed to remove biocompatibility test on the molded part.
 
Thread starter Similar threads Forum Replies Date
S Laminaria - Is a Biocompatibility Study Required? EU Medical Device Regulations 6
S Laminaria - Biocompatibility Study? Imported Legacy Blogs 1
F Biocompatibility study conducted in the GLP lab outside Japan Japan Medical Device Regulations 2
A Biocompatibility study necessary? Polycarbonate tip wrench tool Other Medical Device and Orthopedic Related Topics 6
K 510k FDA review, will they accept Biocompatibility result generated using feasibility product lots? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
A How much does a complete biocompatibility test package cost? Other ISO and International Standards and European Regulations 1
M FDA News FDA Releases Draft Guidance Clarifying Application of ISO 10993-1 Biocompatibility Standard Medical Device and FDA Regulations and Standards News 0
F Biocompatibility evaluation for Hardware/Interface Components? Other Medical Device Related Standards 12
L Process changes and biocompatibility (ISO 10993-1) Other Medical Device Related Standards 1
Beliz Biocompatibility Testing for Laser Epilation Device EU Medical Device Regulations 2
L FDA Biocompatibility Requirements - Transitory Contact Other US Medical Device Regulations 1
K Biocompatibility Testing - Multile products of different sizes and shapes US Food and Drug Administration (FDA) 2
K Biocompatibility evaluation of gas pathways Medical Device and FDA Regulations and Standards News 5
K Diagnostic X-ray devices - Applicability of Biocompatibility Testing per ISO 10993-1 Manufacturing and Related Processes 7
Judy Abbott Activation of complement at 90 min - BSI requested for justifications to support the biocompatibility of the device EU Medical Device Regulations 2
Judy Abbott Can any one help me with an urgent BIOCOMPATIBILITY Impact assessment? Internal Auditing 9
S ISO 10993 Biocompatibility Evaluation - Electronic thermometer Other Medical Device Related Standards 3
shimonv Referring to previously submitted biocompatibility results in a new 510(k) submission Other US Medical Device Regulations 1
I ISO 10993 Biocompatibility Requirements - manufacturing process chemicals Other Medical Device Related Standards 8
R 510(k) Biocompatibility vs Cleaning Validation Other Medical Device Related Standards 4
K Biocompatibility classification for Lancet CE Marking (Conformité Européene) / CB Scheme 7
F Sample Size for Biocompatibility Tests Other Medical Device Related Standards 4
F Biocompatibility Device Categorization: Priming with Blood Other Medical Device Related Standards 1
W Biocompatibility Risk Analysis for Clinical Practitioner 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
F Biocompatibility on aged devices Other Medical Device Related Standards 3
Gamula Biocompatibility GLP conform tests outside China China Medical Device Regulations 1
K Biocompatibility Testing - Component Level or Assembled Medical Device? ISO 13485:2016 - Medical Device Quality Management Systems 7
I Biocompatibility Impact on Autoclaved Silicone Other Medical Device Related Standards 3
D Should Biocompatibility Tests be performed on Production Samples? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
R What is the expectation for Biocompatibility Testing when switching Suppliers? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
C Biocompatibility Test for a Non Contact Medical Device US Food and Drug Administration (FDA) 11
R Biocompatibility - ISO 10993 and SVHC Requirements Other Medical Device Related Standards 4
planB Biocompatibility: Blue Book Memorandum #G95-1 superseded by new FDA guidance Other Medical Device Related Standards 1
S ISO 10993-1: External Communicating Device biocompatibility test requirements Other ISO and International Standards and European Regulations 4
M Biocompatibility - why are animal studies necessary? Other Medical Device and Orthopedic Related Topics 11
R Biocompatibility Test of Blue Gown vs Blue Drape 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
R FDA Biocompatibility questions for Surgical Gown 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D Biocompatibility requirements for a Class IIa medical device CE Marking (Conformité Européene) / CB Scheme 7
S Biocompatibility testing for nasal prongs made of medical grade silicone 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
J Biocompatibility tests in AAA LAC Laboratory ISO 13485:2016 - Medical Device Quality Management Systems 6
B Biocompatibility for Packaging Materials Other Medical Device Related Standards 3
J Biocompatibility Evaluation component other than Testing Other Medical Device Regulations World-Wide 1
J Grouping Products and Testing Sterile and Biocompatibility - Validating Products Other Medical Device Regulations World-Wide 3
M Biocompatibility in Respiratory Products - Help interpreting whitepaper IEC 60601 - Medical Electrical Equipment Safety Standards Series 22
M Three final products in one biocompatibility test? Other Medical Device and Orthopedic Related Topics 2
S Selecting materials for implants to comply with ISO 10993 biocompatibility Other Medical Device Related Standards 4
S Laminaria Stick (Cervical Dilator) - Biocompatibility Test required? ISO 13485:2016 - Medical Device Quality Management Systems 1
T Biocompatibility for "Desktop" used Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 4
M Biocompatibility Test requirements for 510k - Orthopaedic Implant Screws US Food and Drug Administration (FDA) 2
A Biocompatibility omission justification for FDA 510(k) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8

Similar threads

Top Bottom