Is calibration of a weigh scale required?

SGquality

Quite Involved in Discussions
Is calibration of weigh scale required ? One of our departments strongly believes that daily verification with single weight will suffice. Is that true ?
 

somashekar

Leader
Admin
Re: Is calibration of weigh scale required ?

Is calibration of weigh scale required ? One of our departments strongly believes that daily verification with single weight will suffice. Is that true ?
A regulatory requirement may be applicable for periodic calibration....
 
W

w_grunfeld

Re: Is calibration of weigh scale required ?

In theory not required in most cases, except if used to verify compliance /decide to accept or reject a product. In this case checking with one weight is NOT CALIBRATION.
iN PRACTICE, 9 out of 10 auditors would write up a discrepancy claiming you can't ensure that it won't be used one day by someone for acceptance check.
Just have it calibrated and avoid the hassle:bigwave:
 

Big Jim

Admin
Re: Is calibration of weigh scale required ?

Have a good read of ISO 9001:2008 7.6a, and pay especial attention to the line above a "Where necessary to ensure valid results" as well as "measuring tools shall be . . . calibrated or verified, or both, at specified intervals . . . against measuring standards traceable to national or international standards . . . "

If it is needed for valid results, it needs to be calibrated.

If it needs to be calibrated, it needs to be done often enough that the calibration cannot be called into question.

Common sense should tell you that verifying a scale with one weight, one point within its range, is not calibration, or at most, is limited calibration.

If the weight set used for calibration is not traceable to an international or national standard (in the US, usually NIST) it is of no effect.

But it all hinges on "Where necessary to ensure valid results".
 

dgriffith

Quite Involved in Discussions
You might ask, if I don't care about the scale accuracy and results, why do I even need the make the measurement?
In any case, after the weigh scale is calibrated, and before its next calibration, you can use a single mass, which is also calibrated, as a check standard during the time interval between calibrations. Use an appropriate SPC method to demonstrate no shift in sensitivity, which will give you confidence in weight measurements.
Some labs use one before every measurement, some at the first of every shift, some longer.
The greater the scale range, the heavier the mass needs to be--can't be putting 5 lbs. on a 100 lb. scale to check sensitivity.
 

AndyN

Moved On
Is calibration of weigh scale required ? One of our departments strongly believes that daily verification with single weight will suffice. Is that true ?

Calibration is the science of knowing how far "off" you are when you make a measurement = the variation which is attributable to the measuring equipment. As has been stated, a single point isn't going to give you that - so whomever is calling THAT "calibration", clearly isn't competent in the terminology. You ARE allowed - to get "valid results" - to use a transfer standard (set of known weights in a range of values). This is perfectly acceptable but it's verification, not calibration since no measurement variation is taken into account. You should perform this type of check either side of the value(s) you normally weigh to ensure you understand the characteristics of the scales - are the results linear in nature, for example.
 
Last edited:

gpainter

Quite Involved in Discussions
Most states require that any thing sold by weight or verified by weight require at least yearly calibration. Most counties have a local weight and measures person that will do it on a yearly basis for free (if you sell by weight or measure) or little cost. Shipping scales are just one example in manufacturing, in the test lab- some tests are based on the weight of the product, truck scales., etc
 
Thread starter Similar threads Forum Replies Date
H Specificity in Weigh Scale Calibration Frequencies General Measurement Device and Calibration Topics 13
R Calibration of Scales for Weigh Counting General Measurement Device and Calibration Topics 19
J Should weigh count scales be in the Calibration System? QS-9000 - American Automotive Manufacturers Standard 4
C IR thermometer for refrigerated Centrifuge calibration General Measurement Device and Calibration Topics 2
G ISO 17025.2017 Clause 7.8.4.3 Certificate of Calibration Label ISO 17025 related Discussions 5
M CMM Calibration - What, exactly, is the output? General Measurement Device and Calibration Topics 11
J Using MSA instead of calibration Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
P Hierarchy of Calibration in Metrology General Measurement Device and Calibration Topics 1
J Calibration frequency for glass scale Calibration Frequency (Interval) 2
S Calibration SOP and external vendors General Measurement Device and Calibration Topics 2
P IATF16949 External Lab Calibration -CIPM MRA IATF 16949 - Automotive Quality Systems Standard 1
CaliperJim Acclimation time prior to calibration General Measurement Device and Calibration Topics 8
S How to report a measurement and relate it to calibration certificate General Measurement Device and Calibration Topics 1
R Do I need to get calibration certificate from ISO 17025 for IATF Audit IATF 16949 - Automotive Quality Systems Standard 8
S Netsuite Preventive Maintenance and Calibration Manufacturing and Related Processes 0
D Roughness specimen calibration Calibration and Metrology Software and Hardware 5
dhuot1990 General Calibration SOP Manufacturing and Related Processes 3
B Gage calibration Gonogo (calcul of cgk) General Measurement Device and Calibration Topics 0
Vader22 As-Found FAILED calibration ISO 14001:2015 Specific Discussions 16
M Gravimetric blender for resin in automotive industry - periodical calibration via external lab with certification Reliability Analysis - Predictions, Testing and Standards 0
N Calibration Method for Scales and Flowmeters Reliability Analysis - Predictions, Testing and Standards 7
T Calibration Tracking ISO 13485:2016 - Medical Device Quality Management Systems 4
Q Considering addition of mass calibration to ISO17025 scope ISO 17025 related Discussions 6
Z Resolution in calibration General Measurement Device and Calibration Topics 1
J Run Down Adapters Calibration Processes General Measurement Device and Calibration Topics 4
S Looking for accredited calibration lab - Onsite up to 650A and 1 kV. General Measurement Device and Calibration Topics 3
CaliperJim Screw Thread Micrometer Anvil Calibration General Measurement Device and Calibration Topics 7
Strange Juan HP 85054B Standard Mechanical Calibration Kit Calibration Procedure General Measurement Device and Calibration Topics 1
H Air Pressure Gage Calibration General Measurement Device and Calibration Topics 6
T Calibration on AVL AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
C Determining an as received OOT condition during calibration does not affect measurements taken prior to calibration. General Measurement Device and Calibration Topics 16
Sam.F Do heat treatment ovens need calibration? General Measurement Device and Calibration Topics 25
J Selecting reference standards for calibration General Measurement Device and Calibration Topics 3
C IATF 16949:2016 Major NC pertaining to 7.1.5.2.1 Calibration/verification records IATF 16949 - Automotive Quality Systems Standard 15
E Calibration Records during AS9100 Audit AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
John Predmore Calibration Determination decision Calibration Frequency (Interval) 16
J ISO 17025 Calibration Other Medical Device Related Standards 2
A Electronic Weighing scale calibration Calibration Frequency (Interval) 4
M Bourdon gauge calibration procedure failure ISO 17025 related Discussions 3
Ed Panek Calibration question Manufacturing and Related Processes 4
S Calibration/Worn gages Manufacturing and Related Processes 9
A Settings rings calibration General Measurement Device and Calibration Topics 4
R 7.1.5.2 -- Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
C CALIBRATION CERTIFICATE REQUIREMENTS General Measurement Device and Calibration Topics 11
S Calibration Label issues General Measurement Device and Calibration Topics 21
H Calibration / Capability in a PCP FMEA and Control Plans 1
R First Time Managing Calibration and Measurement System. Need Help. General Measurement Device and Calibration Topics 19
M CE marked medical device - do they need UKAS certified calibration provider ISO 17025 related Discussions 4
JDJohnson Micrometer & Caliper Calibration (ISO-9001) General Measurement Device and Calibration Topics 26
B Precision Current Loop for Current Probe Calibration Reliability Analysis - Predictions, Testing and Standards 2

Similar threads

Top Bottom