Is EMC + Radio Re-Test Required?

F

Fred1977

#1
Hi,
I have a medical product that incorporates a bluetooth module for comms with a host PC, and a dongle that plugs into the host PC that also incorporates a bluetooth module. Previously we paid $6K for EMC and radio testing (for emc, EN301 489-1 v1.9.2, EN 301 489-17 v2.2.1, EN60606-1-2, and for radio EN301 328 v1.8.1).

I've had to replace the bluetooth module in both unit and dongle with one from a new manufacturer (it's pre-certified), with some very minor changes to the PCB to accomodate the new module - do I need to repeat all the costly testing above, a subset of it, or none of it?

I thank you very much in advance for any input.
 
Last edited by a moderator:
Elsmar Forum Sponsor

Peter Selvey

Staff member
Moderator
#2
It is kind of the wrong way to look at it.

Instead of saying "do I need to retest?" you should be asking "is the old test data representative of the new version?"

Legally, you must always have evidence to support the actual marketed device, including newer versions. It is OK to use or refer to test data on previous versions, but it is not so much a decision about re-test but a decision whether the old data is representative of the new device. If yes, the old data formally becomes the evidence for the new device.

This should be documented, and the person making the decision should be suitably qualified and not subject to undue conflicts of interest (especially, for medium to high risk devices).

Regulatory agencies can of course challenge the decision. But keep in mind they need objective evidence that the original decision was wrong. If the manufacturer wrote up a reasonable rationale, and the person writing was clearly qualified, they need more than just a blind "we don't agree". They will need to point at the technical mistake in the logic or justification.
 
Thread starter Similar threads Forum Replies Date
T GPRS (General Packet Radio Service) and EMC IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601-1-2 EMC Tables IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
L IEC60601-1-11 Home Healthcare + IEC60601-1-2 EMC (Functional Earth allowed?) IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M EMC Directive and product safety standards CE Marking (Conformité Européene) / CB Scheme 1
B Evaluation of Basic Safety during EMC Immunity or Climate Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
B Does EMC Directive 2014/30/EU cease to apply when the Machine Directive 2006/42/EC is applicable? CE Marking (Conformité Européene) / CB Scheme 4
H External power supply - requirements in safety/EMC test? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
F EMC testing for a system that is provided with a computer IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
O EMC - EFT (Electrical Fast Transients) on SIP couples erroneously with other cable CE Marking (Conformité Européene) / CB Scheme 1
K Is a 3V torch classified as medical electrical equipment needing EMI/EMC testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Arguing with the test lab about failing in an EMC test IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
E Airborne Medical Equipment EMI/EMC IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S Portugal ISO 13485 Certification and EMC Test Houses Other Medical Device and Orthopedic Related Topics 3
T Question: Do we consider 60601-1-2 (4th ed) EMC testing a mode for a diagnostic port? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
O Medical Device EMC Risk Management CE Marking (Conformité Européene) / CB Scheme 4
O Guide on EMC for Functional Safety CE Marking (Conformité Européene) / CB Scheme 3
T CE Marking & Compliance to MDD, EMC, RoHS, and GPSD REACH and RoHS Conversations 4
J IEC 60601-2-2 - EMC Requirements - Surgical instrument manufacturer IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
O EMC testing per IEC 60601-1-2 for considering new cables CE Marking (Conformité Européene) / CB Scheme 1
O EMC testing under RED and IEC 60601-1-2 CE Marking (Conformité Européene) / CB Scheme 1
D EMC Testing for a Production Fixture CE Marking (Conformité Européene) / CB Scheme 0
Q IVD vs EMC OJ (Official Journal of the European Union) Harmonised Standard Conflicts EU Medical Device Regulations 1
A EMC Emission Issues - Communication between PC and Monitor IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Particular standards, essential performance in EMC Immunity Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F EMC (Electromagnetic Compatibility) Conformity Charts - IFU Requirements Misc. Quality Assurance and Business Systems Related Topics 3
G CE Marking - Low Voltage Directive and EMC Directive CE Marking (Conformité Européene) / CB Scheme 4
T Firmware version in EMC report / DoC CE Marking (Conformité Européene) / CB Scheme 1
W EMC, ESD for battery operated apparatus Other Medical Device and Orthopedic Related Topics 1
N New EMC IEC60601-1-2 4.0 Ed. Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
B What is our requirement for maintaining the EMC series of standards? IEC 61000 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
P EMC requirements for an accessory to a medical device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T EMC for Medical Devices in Home Environment Other US Medical Device Regulations 4
J EMI/EMC Test differences for Radiography Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M EMC Testing for a US-only, Home-Based Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Safety & EMC certificate updated when MCU (MIcro-Controller) is changed? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
T EMC Tests - While certified power supply is part of device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B EMC Repeat Justification - Reducing NRTL testing required after modification of ME IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C EMC Test Report Requirements to IEC 60601-1-2:2009 ISO 13485:2016 - Medical Device Quality Management Systems 4
O CE TCF (Technical Construction File) Template for LVD/EMC/Pressure Directive CE Marking (Conformité Européene) / CB Scheme 5
P New EMC standard IEC 60601-1-2:2014 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
T IEC 60601-1-2 EMC and WiFi IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
T How to fulfill IEC 60601-1-2 EMC Testing Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S IEC 60601-1-2 EMC Testing Optional in US? IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
S Does Australia/New Zealand require RoHS, WEEE, REACH and EMC? RoHS, REACH, ELV, IMDS and Restricted Substances 4
N Device with Safety & EMC Test Certificates from different Laboratories ISO 13485:2016 - Medical Device Quality Management Systems 22
N User Manual Content Requirements and EMC information ISO 13485:2016 - Medical Device Quality Management Systems 1
B Should EMC (Electromagnetic Compatibility) Testing be performed In-House CE Marking (Conformité Européene) / CB Scheme 10
C An ongoing EMC test in accordance with IEC/EN 60601-2-47 EMC IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
E List of Complete EMC Test Requirements consisting of IEC, EN, Cispr, etc. General Measurement Device and Calibration Topics 2
J EMC (Electromagnetic Compatibility) Requirements - IEC 60601 3rd edition Clause 17 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Similar threads