Is IEC 60601-1-2 compliance/testing required for battery powered Class I Device

Elsmar Forum Sponsor

Peter Selvey

Staff member
Moderator
#2
In principle yes, but many tests would not be applicable.

It will also depend on what is inside the device, e.g. if it is just very low frequency analogue electronics, no CPU, digital electronics etc, then the tests might be almost non existent. I expect the ESD test is always applicable no matter how simple the device, but an EMC expert would know better.
 
#3
Another factor that would influence which tests are applicable and which are not is what leads are used with the device (if any). The length and function of each lead needs to be considered.
 
Thread starter Similar threads Forum Replies Date
K Is Compliance Testing of IEC 60601-2-63 mandatory for Europe - Dental X-Ray machine CE Marking (Conformité Européene) / CB Scheme 3
K IEC 60601-1-9 Compliance mendatory or Not - CT Scan machine CE Marking (Conformité Européene) / CB Scheme 8
L IEC 60601-1-11 IP22 Compliance - Ingress protection testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E IEC 60601-1-2 compliance for accessories - Electromagnetic compatibility of the ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
H On choosing touchscreen displays and ensuring IEC 60601 compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
K IEC 60601-2-2:2017 Compliance Dates IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
P FDA's deadline for IEC 60601-1 3.1 edition compliance - Aug 1, 2016 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Is IEC 60601 compliance possible when only generic power-supply is specified? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M Routes to IEC 60601 Compliance: Discussion of Merits and Pitfalls IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
K Is compliance to IEC 60601-1 Required for my Product? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
M IEC 60601 Compliance requirements for NON- ME Equpment/Systems IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B When is Compliance required for 4th edition IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
L Terminal Block use in compliance to IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
P Recommended Books on Electrical Safety Compliance for IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
N IEC 60601-1-2 ESD: Lower Compliance Level for ESD Immunity IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K IEC 60601 Compliance Matrix wanted IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601-2-24 Ed. 2.0 - Grace Period for Compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
G 510K for Active Medical Device - IEC 60601-1 or UL 60601-1 Compliance Other US Medical Device Regulations 3
M 60601-1 3rd Edition Compliance with Particular Standards & Equivalency IEC/EN stds EU Medical Device Regulations 6
E Retrofitting Risk Management for IEC EN 60601 3rd Edition Compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
Q Confused: IEC 60601-1-2:YYYY and Required Compliance Dates IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1

Similar threads

Top Bottom