Is IEC 60601-1-2 required by FDA for all electronic medical devices?

#1
I have a question regarding a Class I electronic hardware device we’re currently developing. We asked a reputable lab to safety test it for us and they sent back a quote specifying IEC 60601-1-2 and FCC CFR 47 Part 15, Subpart B. The timeline they gave us for the IEC 60601-1-2 was not what our CEO was expecting, so he asked the lab – without talking to his QA/RA guy (me) – what was the minimum we needed to put the product in the hospital. The lab said just the FCC testing. They inferred that IEC 60601-1-2 was something the hospital expected, but that we were OK with FDA not getting the IEC 60601-1-2 testing done.

Does this sound right?
 
Elsmar Forum Sponsor

Peter Selvey

Staff member
Super Moderator
#2
Strictly speaking, yes. In the US medical device regulations, for the most part standards don't exist in the law, they are there to help smooth the FDA's clearance for sale process such as a 510(k). If your device does not need to go through pre-market clearance then technically standards are not relevant. However if you do need a 510(k) then FDA will expect to see IEC 60601-1-2, or at least will require a good explanation if not applied.
 
Thread starter Similar threads Forum Replies Date
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
N ECoG/EEG electrodes - Required IEC 60601-1 tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K Why two fuses are required - IEC 60601-1 cl. 8.11.5 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
G What edition of the IEC 60601-1 standard is required? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
L IEC/EN 60601-1 Certificate required before CE Marking? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K When is IEC 60601 required? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W Is IEC 60601-1-2 compliance/testing required for battery powered Class I Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Is compliance to IEC 60601-1 Required for my Product? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
B When is Compliance required for 4th edition IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S Separation Devices as required in 16.5 of IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
K Identifying Required Testing to comply with IEC 60601 EU Medical Device Regulations 4
D Battery Powered Disposable Pump - What IEC 60601 testing is required, if any? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S IEC 60601-1-4 - Required for Data Management System that just collects Data IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
Q Confused: IEC 60601-1-2:YYYY and Required Compliance Dates IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 60601 - List of required tests - Medical Device Class IB IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
M IEC 60601-1 Testing required for low-voltage devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-2-57 light source classification IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
L Top opening of enclsoure against fire in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C EN 45502 vs IEC 60601-1-8 Other Medical Device Related Standards 0
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1

Similar threads

Top Bottom