Is IEC 60601-2-37 suitable for Fetal Doppler Detector?
Roland chung
I guess someone here has experience in ultrasound.
I would like to konw if IEC 60601-2-37 is suitable for fetal doppler detector which mainly used to detect the fetal heartbeat.
The scope of standard does not exclude the fetal doppler. But if IEC 60601-2-37 applies, many clauses seems not applicable because of the ultrasound power is sufficiently low.
Please kindly advise it.
Thank you very much!
Roland
You could try this one -
IEC 61266
Ultrasonics - Hand-held probe Doppler foetal heartbeat detectors - Performance requirements and methods of measurement and reporting
and there is IEC 61157 - Standard means for the reporting of the acoustic output of medical diagnostic ultrasonic equipment.
Search the IEC website (www.iec.ch) for anything from technical committee TC87
raymond02302
IEC 60601-2-37 is applicable to such equipment. This standard is for all ultrasonic diagnostic and monitoring equipment. Regarding fetal doppler detector (heart rate ?), its acoustic output generally needn't to be declared (TI or MI less than limit required to be declared)
R How to decide a suitable contact time from table 23 of IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Transformer Requirements - IEC 60601-1 vs. IEC 61558-2-15 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T Clearance and Creepage according to Subclause 8.9.1 for MOPP IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4