Is IEC 60601 applicable to our smart bed?

#1
Hello Elsmar,

We are developing smart beds for children with Autism and other special needs. We have determined based on similar products that we're a Class 1 510K exempt product. We're trying to determine what standards are applicable to our bed and specifically if IEC 60601 is.

Our beds have calming sensory inputs (lights, speakers, vibration) as well as a camera for remote monitoring by parents/caregivers. These are the only electrical components. In the future, we may have actuating lifts to raise the head/feet but not yet. All electrical devices are previously UL certified as standalone products.

I understand that the definition of 60601 is:
3 * Terminology and definitions
For the purposes of this document, the following terms and definitions apply.
(…)
3.63
* MEDICAL ELECTRICAL EQUIPMENT
ME EQUIPMENT
electrical equipment having an APPLIED PART or
transferring energy to or from the PATIENT or
detecting such energy transfer to or from the PATIENT
and which is:
a) provided with not more than one connection to a particular SUPPLY MAINS; and
b) intended by its MANUFACTURER to be used:
1) in the diagnosis, treatment, or monitoring of a PATIENT; or
2) for compensation or alleviation of disease, injury or disability
We aren't transferring energy to or from the patient or detecting energy.

I am unsure if we qualify as having "applied parts". Sub-clause 3.8 states that an APPLIED PART is “part of ME EQUIPMENT that in NORMAL USE necessarily comes into physical contact with the PATIENT for ME EQUIPMENT or an ME SYSTEM to perform its function.

Patients within the bed do not necessarily need to come into contact with the electronics for them to function. They could touch the face of the camera/light though. Currently, we do provide the option for the patient to click buttons on the light/speaker to turn them on/off but we could restrict access to this or specify in instructions that only parents/caregivers, not the patient, would be the ones that should access these.

Any help is appreciated.

Thank you,
Caleb
 
Last edited:
Elsmar Forum Sponsor

ECHO

Starting to get Involved
#2
Without knowing more about your design, device description and intended use, it will be hard to say what standards will apply to you device.
However, below are a few that you might want to review.

IEC 60601-1-6
IEC 60606-1-8
IEC 60601-1-11
IEC 60601-2-49
IEC 60601-2-52
IEC 62304

Also, isn't light, sound and vibration different forms of energy?
 

AbelVV

Starting to get Involved
#3
Vibration is definitely energy. There's a clause about it in 60601 (IEC 60601-1 9.6), but that doesn't necessarily make it fit in there. A hand held drill is not under 60601, for example.

The real question is if 1 or 2 are met.

1) in the diagnosis, treatment, or monitoring of a PATIENT; or
2) for compensation or alleviation of disease, injury or disability


Also the use of bedsheets doesn't mean that there is no physical contact between the bed and the "patient", in my opinion. Imagine all the kinds of shenanigans you could pull off doing that.
 

Marcelo

Inactive Registered Visitor
#4
If it's a bed and a patient lies on it, and it's electrical, it has an applied part and IEC 60601 is applicable.
 
Last edited:

AbelVV

Starting to get Involved
#5
If it's a bed and a patient leis on it, and it's electrical, it has an applied part and IEC 60602 is applicable.
But 1) and 2) is what make the distinction "patient" vs regular user. If I have a massaging bed at home it is not a medical device. In this case, does the fact that the intended user has autism preclude that they are a patient?
 

Marcelo

Inactive Registered Visitor
#6
But 1) and 2) is what make the distinction "patient" vs regular user. If I have a massaging bed at home it is not a medical device. In this case, does the fact that the intended user has autism preclude that they are a patient?
The OP does mention that it's a patient, so this distinction does not seem to make a difference here.
 
Thread starter Similar threads Forum Replies Date
G Is impact test according to IEC 60601-1 applicable for HAND-HELD equipment if also classified as BODY-WORN? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
L Pneumatic powertool (Drill) - Is IEC 60601-1 applicable? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 and 60601-1-2 - Not Applicable for My Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 23
P Is IEC 60601-1 applicable to software-only medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Unused SIP/SOPs - IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Proper document of SMPS used in infant warmer for IEC 60601-1 testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Sequence of testing in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Has anyone done an IEC 60601-1 gap analysis to IEC 60335? Medical Device and FDA Regulations and Standards News 4
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
DitchDigger IEC 60601-1 subclause 5.1 - Adequate evaluation in lieu of testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Transformer Requirements - IEC 60601-1 vs. IEC 61558-2-15 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T Clearance and Creepage according to Subclause 8.9.1 for MOPP IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601-1 and IEC 60601-2-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3

Similar threads

Top Bottom