So, I got a note from my boss to investigate IEC61508.

IEC61508 would be very useful to us as we do have some electronic safety related systems, however it seems strange to me that I have not encountered any medical device that is IEC61508 certified (at least they do not advertise it on their brochure).

A quick look at the MDD directive in Europe makes no mention of IEC61508 and it does not seem that the standard has been harmonized by the European community.

Have you ever come across a medical device that has the IEC61508 approval ? Is 61508 mandatory for medical devices ?
 
Elsmar Forum Sponsor
#2
Re: Is IEC61508 applicable to Medical Devices

Check the annex C of 62304 where you will find a description of how 61508 relates to this MDD harmonised standard.

Risk is handled differently but 61508 is identified as being useful for recommendations on techniques, tools and methods.
 
K

Krisp

#3
Re: Is IEC61508 applicable to Medical Devices

While 62304 does describe the relationship between 61508 and 62304, it only uses 61508 as a guide for writeup of the standard itself.

Interesting, but I still need to see if there is any regulation somewhere that mandates the use of 61508. Has to be Black on white as they say.
 
#4
Re: Is IEC61508 applicable to Medical Devices

No you do not have to apply this standard.

You might want to choose to apply it if it has been requested by a customer.
 
#6
We were asked by the notified body (TUV Sud) to submit a Functional Safety Report as per 61508 during the review of TCF, for our class 2b medical device. Wondering how to confront them?
 
Thread starter Similar threads Forum Replies Date
H Functional Safety PFH vs PFD according to IEC 61508 Reliability Analysis - Predictions, Testing and Standards 3
sagai Safety Integrity Requirement as per ISO/IEC 61508 Other ISO and International Standards and European Regulations 2
I SIL (Safety Integrity Level) calculation per IEC 61508. Reliability Analysis - Predictions, Testing and Standards 6
N Seeking IEC 61508 Standard Other ISO and International Standards and European Regulations 3
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of amy component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A IEC 62304 safety classification, External Controls and off-label use related risks IEC 62304 - Medical Device Software Life Cycle Processes 5
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C ISO/IEC 17021-1 clause 7.1.2 - Determination of competence criteria Document Control Systems, Procedures, Forms and Templates 2
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C ISO/ IEC 17021 Resource requirement (need help) Document Control Systems, Procedures, Forms and Templates 5
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
S IEC 62304 software costs and time Medical Device and FDA Regulations and Standards News 3
S IEC 62304 - Software verification cost IEC 62304 - Medical Device Software Life Cycle Processes 3
Sravan Manchikanti Software Risk Management & probability of occurrence as per IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
M IEC 62304 Software changes - Minor labeling changes on the GUI IEC 62304 - Medical Device Software Life Cycle Processes 3
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K IEC 62304 - Testing Independance IEC 62304 - Medical Device Software Life Cycle Processes 5
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Exclusions or justification for non-applicability of IEC standards Reliability Analysis - Predictions, Testing and Standards 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
T ISO/IEC 17065 certification scheme Help Other ISO and International Standards and European Regulations 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
K IEC 62304 - Functional and performance requirements for SOUP items IEC 62304 - Medical Device Software Life Cycle Processes 2
K IEC 62304 compliance - Code reviews as part of verification strategy IEC 62304 - Medical Device Software Life Cycle Processes 5
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 62471 vs IEC/TR 62778 CE Marking (Conformité Européene) / CB Scheme 0
B IEC 61010, Touch currents related to burns at higher frequencies (cl. 6.3, A.3) Other Medical Device Related Standards 0
M IEC 61000-4-2: Requirements for coating declared as insulating IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
B IEC 61010-1 - Voltage / Dielectric strength tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M When is EN 62133-1 (new version i.e. IEC 62133:2017) mandatory in EU Other ISO and International Standards and European Regulations 0
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Unused SIP/SOPs - IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1

Similar threads

Top Bottom