Hi everyone. I want to have some insights related to the section Impartiality which has been added to the revised guidance document IS/ISO/IEC 17025:2017. Under this section they have mentioned requirements for impartiality in four scentences which do have a vast meaning which can be interpreted and applied in various ways. What i want to know ours is a testing laboratory which has been recognized by NABL, now how can we implement imaprtiality?? Does it require to establish a new procedure?? or else we can justify that the requirements stated under impartiality section are getting indirectly reflected in our existing procedures?? If it requires to establish a new procedure for impartiality, then should we reflect impartiality for the technical sytem or process or management system??
Thanks in advance.
Benjamin Weber
The requirements regarding impartiality are not completely new in 17025:2017. Similar requirements were already in the 2005 edition, see subclauses 4.1.4, 4.1.5 b) and 4.1.5 d) (here impartiality is explicitely mentioned!).
So actually, when your lab was already accredited to the 2005 edition, there shouldn't be much to do for compliance with the 2017 edition.
17025:2017 goes a little more into detail when it comes to impartiality and gives a few examples (e.g. in the note to subclause 4.1.4).
What you should now do is:
(1) Have a procedure to identify the possible risks, that could lead to a loss of impartiality. Since there is no requirement, how this should be done, you are free to choose a way, that fits you. This might be a formal risk analyis or just a meeting, where you discuss these possible risks.
(2) Record the identified risks and the evaluate, if you have to take any precautions to minimize the risks.
(3) In case you identified risks, where you have to act you should record the implementations.
(4) Re-identify and re-evaluate the risks regularly, e.g. once a year and when certain triggers might occur (complaints, audits, lab acqusisiton by another economic stakeholder, management review)
There are some general aspects that might affect impartiality:
- There is only one client you work for (or one client that generate the majority of your sales volume): You maybe are willing to deliver "client-friendly" test results, because the dissatified client may go to another lab and as result you get into financial problems.
- There is a relation between the lab staff and the client, e.g. one tester owns a lot of shares of a client company, is the brother-in-law of the owner of the client company or similar. These are typical conflicts of interests, that should at least be disclosed and recorded.
I hope this helps a littel to clarify.
Thank you so much Benajmin for the clarification, information you have provided.
Tallinec
Venkatesh Swamy,
In you procedure or in other document: 1)you shall list up available and possible risks of impartiality; 2)to provide actions to be done or actions in progress to eliminate/minimize identified risks of impartiality (see 1).
Examples:
Available risk of impartiality: salaries of technicians depend on number of issued testing reports.
Action to be done to eliminate the risk: to add in labor contracts of technicians a provision that their salary does not depend on number of issued testing reports.
or:
Possible risk of impartiality: owner of lab presses technicians to use obscure wording in testing reports.
Action to be done to minimize the risk: to add in labor contracts of technicians a provision that in professional activities they take orders only from the senior technician of lab.
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
D Laboratory Manual ISO/IEC 17025 Example wanted ISO 17025 related Discussions 2
Douglas E. Purdy ISO/IEC 17025:2017 3rd Ed. Changes from 2nd Ed. ISO 17025 related Discussions 6
Douglas E. Purdy ISO/IEC 17025:2017 Clause 8 & Annex B ISO 17025 related Discussions 9
Ajit Basrur Informational ISO/IEC 17025:2017 Published - November 2017 ISO 17025 related Discussions 8
M Does Calibration to ISO/IEC 17025 conform to Z540.3? ISO 17025 related Discussions 1
D ISO/IEC 17025 Implementation Workshop Ideas ISO 17025 related Discussions 2
Hershal ILAC is preparing to start the possible revision to ISO/IEC 17025 ISO 17025 related Discussions 2
J ISO/IEC 17025 - Which External Documents require control? ISO 17025 related Discussions 6
Z Force & Strain Calibration Uncertainty for ISO/IEC 17025 ISO 17025 related Discussions 2
R Implementation of ISO/IEC 17025 Requirement for our Test Laboratory ISO 17025 related Discussions 2
K Where to buy ISO/IEC 17025:2005 ISO 17025 related Discussions 4
G A2LA's ISO/IEC 17025 and Accreditation Training - Opinions from previous attendees ISO 17025 related Discussions 2
M Does ISO/IEC 17025:2005 require the Laboratory to have separate ISO 9001 Procedures ISO 17025 related Discussions 4
P NABL ISO/IEC 17025 Manual Requirements ISO 17025 related Discussions 3
QMMike TS 16949 - A2LA or ISO/IEC 17025 required for material testing? ISO 17025 related Discussions 6
N Measurement Uncertainty Explanation per ISO/IEC 17025:2005 clause 5.10.4.2 ISO 17025 related Discussions 3
G Is Agilent?s automated network analyzer verification ISO/IEC-17025 compliant? ISO 17025 related Discussions 4
R Example of ISO/IEC 17025 Implementation Project Plan ISO 17025 related Discussions 8
I Unapproved Accreditation Bodies offer ISO/IEC 17025 Accreditations ISO 17025 related Discussions 3
suildur Assuring the Quality of Test and Calibration Results - ISO IEC 17025 Clause 5.9 ISO 17025 related Discussions 13
Hershal Training in ISO/IEC 17025:2005 - May 2010 Washington, DC ISO 17025 related Discussions 3
A Compliance with ISO/ IEC 17025:2005 Clause 5.9 for Calibration Laboratories ISO 17025 related Discussions 6
A How to comply with ISO/IEC 17025 Laboratory Requirements ISO 17025 related Discussions 7
D Calibration Frequency of CMM as per ISO/IEC 17025 ISO 17025 related Discussions 11
Hershal Training in ISO/IEC 17025 and 17020 ISO 17025 related Discussions 0
V ISO/IEC 17025 Clause 4.4 - Review of requests, tenders and contracts ISO 17025 related Discussions 4
T Calibration Procedures - ISO/IEC 17025, Clause 5.5.9 - Outsourced Calibration ISO 17025 related Discussions 10
Hershal Update from MSC........ ISO/IEC 17025 - Challenges and Solutions ISO 17025 related Discussions 1
Hershal Reviving the discussion on ISO/IEC 17025 ISO 17025 related Discussions 4
Hershal Introduction to ANS/ISO/IEC 17025 and Z540 ISO 17025 related Discussions 2
Hershal ISO/IEC 17025:2005 Clause 4.1.5 and 4.1.6 ISO 17025 related Discussions 5
S Project Kick-Off - First visit agenda to kick off a ISO/IEC 17025:2005 ISO 17025 related Discussions 1
D Do Supplier's Laboratories Need to be Accredited to ISO/IEC 17025 ISO 17025 related Discussions 15
S Laboratory is accredited to ISO/IEC 17025 but that is not stated on the Cal Cert ISO 17025 related Discussions 7
B What Laboratories are the 'National Equivalent' of Accredited to ISO/IEC 17025? ISO 17025 related Discussions 3
Hershal ISO/IEC 17025 and ISO/IEC 17020 training ISO 17025 related Discussions 2
D Appropriate Equipment for ISO/IEC 17025:2005 Clause 5.6.3.3 - Intermediate Check ISO 17025 related Discussions 5
A Linearity and Eccentricity Point Definitions - Ref: ISO/IEC 17025:2005 ISO 17025 related Discussions 1
J ANSI/NCSL Z540-1-1994 vs. ANSI/ISO/IEC 17025:2000 ISO 17025 related Discussions 4
Hershal For accredited labs that expect more - Certification of ISO/IEC 17025 assessors ISO 17025 related Discussions 4
WCHorn ISO/IEC 17025 Registrars - I'm confused about "accredited" registrars for ISO 17025 ISO 17025 related Discussions 18
A Sampling - What does it mean by Sampling in ISO:IEC 17025 ISO 17025 related Discussions 1
Hershal Measurement Uncertainty - Need help with ANS/ISO/IEC 17025 and MU ISO 17025 related Discussions 0
Hershal Measurement Uncertainty Training - ANS/ISO/IEC 17025, ISO/IEC 17020 ISO 17025 related Discussions 0
A ISO/IEC 17025 Training in Pakistan, or any other country in that region ISO 17025 related Discussions 4