Is it acceptable it is to list new products under an existing 510(k)?
#11
I can understand both points of view to be honest, so I'm still not sure which thought process is best to go with.
For example we could manufacture a device and sell it to another company in America. Therefore, the labelling would be branded with their logos etc...
If we were to sell this same device to a second company, it would involve a non-significant change to the labelling to update the branding on the labelling.
However, following the train of thought that if the 510(k) was cleared for one device (and one device only), would that mean that we would have to submit a new 510(k) even though the only difference between the two devices would be the artwork on the packaging?
Bugeater55
#12
If you are rebranding the same product for a different client then a note-to-file is acceptable as FDA does not approve labeling for 510(k)-cleared products.
MIREGMGR
#13
I agree. In the case of two private label customers selling what otherwise is an identical or not-significantly-different device product, for 510(k) purposes there is only one product.
FDA cares about regulatorily significant changes to labeling content. The name of the private label customer is not regulatorily significant.
#14
One more thing worth to notice...
As far as I know, if your product is a IVD device, you may have to submit another CLIA (add-to-file) when you change device name or distributor.
Julie O
#15
Consult the various 510(k) related guidances. FDA's language is significant; when they write "device" (singular), that's what they mean.
I would say that's a very long way from explicitly saying anything.
If you go through the flowchart and determine that no new submission is required, what you have determined is that A, B, C, D, and E, are, from FDA's perspective, all the same device, cleared under the same 510(k). The differences among them may be meaningful from a manufacturing or marketing perspective, but what the flowchart tells you is whether those differences are meaningful from a regulatory perspective.
W Misinterpretation of requirement acceptable as root cause? Problem Solving, Root Cause Fault and Failure Analysis 19
M Are Fungal counts acceptable in class 1000 clean rooms? Other ISO and International Standards and European Regulations 2
T ISO 14971-2019 doubt - Evaluate if estimated risks are acceptable ISO 14971 - Medical Device Risk Management 9
samer Acceptable limits for Spills - Tracking Hydraulic Spills ISO 14001:2015 Specific Discussions 3
V Acceptable maximum RSD (relative standard deviation) for an sample size Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
B Gage R&R Acceptable (10-30%), deduct Total Variation from Tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
Q Acceptable calibration accuracy of a 60" linear measuring device General Measurement Device and Calibration Topics 16
F It is acceptable moving remote locations staff to manufacturing plant for auditing? IATF 16949 - Automotive Quality Systems Standard 3
Q Is it acceptable to mix components from two different lots into an assembly? ISO 13485:2016 - Medical Device Quality Management Systems 1
C Acceptable NDC for %GR&R part inspection to Tolerance (%Tolerance) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
D Risk Register - have we considered enough and is the format acceptable? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
K Operator Checks - How to show that they were completed on a checklist as acceptable IATF 16949 - Automotive Quality Systems Standard 2
C Example Work Instruction/Procedure for AQL (Acceptable Quality Level) AQL - Acceptable Quality Level 4
S Dates on Labels acceptable to the USA - GS1 General Specification 3.4.4 Other US Medical Device Regulations 3
E Root Cause Analysis - Is Insufficient Understanding an acceptable Root Cause? General Auditing Discussions 9
W Unattached (stand-alone) Forms acceptable for AS9100C? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
V Why Gauge R&R 10% is acceptable for variable instruments ? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
X Acceptable methods to store and archive records ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Is it acceptable to audit to the Nuclear Principles? Quality Manager and Management Related Issues 9
S If this EN ISO 11137 certificate acceptable for Contract Sterilization? ISO 13485:2016 - Medical Device Quality Management Systems 2
D What is satisfactory & acceptable for Coliform - Staphylococcus - Hands and Surfaces Food Safety - ISO 22000, HACCP (21 CFR 120) 3
S Is a Dynamic Controlled Form acceptable per 21CFR820 Document Control Systems, Procedures, Forms and Templates 3
C Lowest Acceptable Grade for Master Gage Set Calibration and Metrology Software and Hardware 1
B Digital Signature Solutions acceptable to use in an FDA Medical Device Environment 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
J Is a ndc less than five is also acceptable? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
T Technical regulations that determine the acceptable limits of cigarettes General Measurement Device and Calibration Topics 4
L IEC 60601-1 3rd edition Clause 9.2.2 Trapping Zone (Acceptable Gaps) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
N Is it acceptable to sterilize products without claiming and labeling "sterile" ? ISO 13485:2016 - Medical Device Quality Management Systems 6
M Quality Objectives - Acceptable Levels and Functions ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 37
C What is an acceptable bioburden level or count when working in cleanroom ISO class7 ISO 13485:2016 - Medical Device Quality Management Systems 1
D Is embedding documents within a Contract an acceptable practice ? Contract Review Process 9
B Non Acceptable % Tolerance Result - Decrease Sorting Limits Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
F ANSI ASQ Z1.4 - Is the following process acceptable? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
C Maximum acceptable Tolerance - Uncertainty Ratio (TUR) General Measurement Device and Calibration Topics 2
S Is it acceptable to have an ASL for Critical suppliers alone? Supplier Quality Assurance and other Supplier Issues 4
S Is this Quality Policy acceptable in the context of ISO 13485? ISO 13485:2016 - Medical Device Quality Management Systems 24
A Help with acceptable screw gage values per standard BS 919 for Go/ No Go gage General Measurement Device and Calibration Topics 4
S What are typical Acceptable Quality Levels (AQL) adopted in Food industry? Food Safety - ISO 22000, HACCP (21 CFR 120) 1
somashekar Who can provide a WHO GMP certificate in India acceptable to the CDSCO Other Medical Device Regulations World-Wide 6
D Root Cause and Corrective Action - Is our Practice Acceptable? ISO 13485:2016 - Medical Device Quality Management Systems 5
W Application of AQL (Acceptable Quality Level) Concept AQL - Acceptable Quality Level 15
H Acceptable Time to provide proof of CAR (Corrective Action Report) Nonconformance and Corrective Action 4
A Sample Size for Mathematical Model with acceptable Confidence Level Using Minitab Software 7
M AQL (Acceptable Quality Level) for Mixed Part Defect IATF 16949 - Automotive Quality Systems Standard 1
C Comparing Two Test Methods - Acceptable Difference(s)? Statistical Analysis Tools, Techniques and SPC 3
S AQL (Acceptable Quality Level) Report Form example wanted AQL - Acceptable Quality Level 1
M What is an acceptable PPM for large volume fasteners? APQP and PPAP 6
K How do you determine what is acceptable base on an FMEA RPN FMEA and Control Plans 6