Is it mandatory for a Calibration Lab to be ISO 17025 accredited?

SGquality

Quite Involved in Discussions
During one of the customer audits, it was pointed that our calibration service provider is not accredited to ISO/IEC 17025?

Is this a requirement? We have had this service provider who is ISO 9001 certified and has traceability to NIST standards.
 

AndyN

Moved On
During one of the customer audits, it was pointed that our calibration service provider is not accredited to ISO/IEC 17025?

Is this a requirement? We have had this service provider who is ISO 9001 certified and has traceability to NIST standards.

But are they competent? ISO 9001 isn't a check of competency of the lab. Calibration isn't just about having a documented process and traceability to NIST. Do the lab know how measurement uncertainty plays into their results, for example?
 

SGquality

Quite Involved in Discussions
Great responses. I had not considered the measurement uncertainty aspect of the calibration service provider. Thanks Ajit and Andy
 

BradM

Leader
Admin
It's only required if the industry (like stijloor pointed out) requires it, or if there is a customer/contractual obligation.

Is there a problem with this vendor?
 

SGquality

Quite Involved in Discussions
It's only required if the industry (like stijloor pointed out) requires it, or if there is a customer/contractual obligation.

Is there a problem with this vendor?

No issues. It was just an observation. Its not for automotive industry.

Merry Christmas all
 
C

chandaraasekar

No issues. It was just an observation. Its not for automotive industry.

Merry Christmas all
yes , if you are a IATF16949 certified company. Even Auto OEM would ask for a certified lab report.
But ISO 9001 or 13485 ( Medical ) standards doesn't mandate this requirement.
 