Is it OK to calibrate a measurement device tighter than manufacturer specifications?
ScottBP
So get this... A customer sends us a dial pressure gauge, 0-3000 PSIG, rated at 1% of full scale by the manufacturer (±30 PSIG.) The customer wanted it calibrated to 0.5% of FS, which is ±15 PSIG. The gage is marked in 50 PSIG increments, with a needle that is maybe 1/4 as wide as the spacing between the markings and no parallax mirror behind the needle. Is the customer fooling themselves by wanting the gauge to be tighter than manufacturer's specs? Is it up to the calibrated eyeball of the technician to guesstimate what the resolution will be?
Even if we can adjust it to be dead on, there's no way we can guarantee it'll meet the customer's required specification after the calibration interval if the manufacturer's spec for that same interval is larger! If we say we calibrate something in accordance with the manufacturer's procedure but then we try to go tighter (as in the example above), then that means we're NOT following the manufacturer's procedure and can get whammied for it on an audit. So how would I word that on a calibration certificate, or do we just tell the customer if they want a tighter specification then go buy a more accurate gauge?
Jerry Eldred
Re: When is it ever OK to calibrate something tighter than manufacturer specs?
It's always okay to calibrate looser than tolerance. But to certify something as tighter than tolerance, you, as a lab (in my opinion) are attesting that the meter is capable of performing better than it's design, and that it will do so for a given interval.
I would think of a polite way to tell them that you can't certify better than it's specifications. It has to have adequate resolution, stability, repeatability, etc. to meet the spec.
This is not to mention that if you attempt to calibrate it and check at various points along the scale, you can't guarantee that it will even test to those specs (even if you optimize it).
My gut feeling is that if you certify in such a manner, you may even potentially incur some liability for it meeting those specs.
bobdoering
Stop X-bar/R Madness!!
You have to help me understand how this works. If you apply a pressure, measured on a high much higher resolution than the gage, and when gage to get calibrated reads 100PSIG, if your gage reads +/-30PSIG of that value its good? If that is the scam, then I guess you could trim the pass/fail tighter to +/- 15 PSIG.... Not sure how you deal with fat needles in your calibration procedure...
Maybe...it might not pass. That is usually the manufacturer's point. And if it doesn't, they don't want to hear about it.
The resolution is still 50 PSIG. The markings don't change.
How could you guarantee it'll meet the customer's required specification after the calibration interval ever? It should be up to their usage to determine the interval. They might play racquetball with the thing...who knows?
If you say you calibrate something in accordance with the manufacturer's procedure but then you try to go tighter, and identify the exception on the report, you are not lying. Now if you didn't mention the exception on the report, then you would be hurting. Does your procedure say you only calibrate something in accordance with the manufacturer's procedure? If so, then simply send it back to the customer uncalibrated. Done.
So how would I word that on a calibration certificate...
"Calibrated per manufacturer's procedure except verified to ±15 PSIG (±30 PSIG manufacturer specification) at the following PSIG: 100, 500, 1000, 1500, 2000, 2500, 2900 PSIG" (or whatever you checked it at...)
...or do we just tell the customer if they want a tighter specification then go buy a more accurate gauge?
Your call...
There a starting point for discussion. :tg:
I don't know about liability. I think either it does or it doesn't pass calibration. If it doesn't, it fails, then they know they have the wrong gage.
I still think it is kooky to have a +/-30 PSIG manufacturer's spec if it reads in 50 PSIG increments...:confused:
Stijloor
Scott,
Do what they want, invoice accordingly ;), and add a disclaimer.
Stijloor.
Hershal
Metrologist-Auditor
Re: Is it OK to calibrate a measurement device tighter than manufacturer specificatio
Short answer, yes you can, and document it under Clause 4.4 Contract Review.
Now the question becomes is there any value in it? That is best left to the customer.
BradM
My answer is no. Every general calibration procedure I have ever seen specifies no tolerance tighter than manufacturer; and no resolution finer than 1/2 the increments. There would have to be interpolation/ interpretation of the actual values due to limited resolution; adding to the uncertainty of the instrument.
Yes, you can hit the cardinal points on the gauge during calibration. But what about reality when the gauge reads processes that are not on cardinal points? There is no way the instrument can display to comply with the accuracy (Is that 17? It looks like 22 to me; it's closer to 100 than 150...:tg:).
It is not advisable to "force" an instrument to comply with process requirements. Instruments must be purchased for their intended use; and calibrated to their requirements/ stated specifications.
I would suggest offering to calibrate to mfg. specifications, and help them spec out a different instrument suitable for that process tolerance.:)
SpongeMouse
i agrre with the cover peeps above.
talk to your client for the reason that it cannot adjust to the specs needed. maybe another gauge in the market can be applicable.
usually, the guage limits from the maufacturer are standardized to all user, so its up to the user to provide means to adjust to its needed limits.
in your case, maybe another type of gauge that can provides more specific limits is applicable.
:notme:
ScottBP
Thanks for the input... In one case, the instrument would have enjoyed about a 6:1 T.U.R. under mfg. specs, but when the customer wanted it tighter, it cut the T.U.R. to 3:1, so I ended up providing an uncertainty statement for all to see... But in another instance where we were barely getting by with a 1:1 calibration by mfg. specs, I had to tell the customer no. :nope:
Jen Kirley
Quality and Auditing Expert
I would never try to calibrate to a tighter tolerance than the gage will read. Half a mark or 1/4 a mark? Phooey. If one wants a tight tolerance, one gets a gage designed and built for accurately measuring to said tolerance. Period. :2cents:
Daniel Walker - 2011
If the customer wants it, and it's properly documented in contract review, I don't see how you could be liable for anything. I would also put a note on the cert stating the customer's requirement.
Now common sense tells you that you should really explain the reasoning to the customer as to why this shouldn't be done.
But, bottom line, if the customer wants it and it's documented, you're operating within the requirements of 17025 as a cal provider.
J Calibrate before use required? Inspection Gages and Measurement Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
M Measurement System Analysis (MSA) - 2500 Instruments - Calibrate every one? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
A Looking for a third party to calibrate our measure equipment - South Carolina General Measurement Device and Calibration Topics 6
I Can I Calibrate On Demand? General Measurement Device and Calibration Topics 9
S Calibrating our own equipment, can we? Micrometers to calibrate vernier calipers General Measurement Device and Calibration Topics 13
F Crimper Calibration - How do you Control & Calibrate your crimpers? Calibration Frequency (Interval) 13
P How to Calibrate a HP 3458A Digital Multimeter with Fluke 5080A calibrator General Measurement Device and Calibration Topics 1
B Choosing not to calibrate (IATF 16949) IATF 16949 - Automotive Quality Systems Standard 6
K New Kilogram - The process we use to calibrate the standards General Measurement Device and Calibration Topics 6
D How to Calibrate any Digital Sphygmomanometer under use for 2 years in a Hospital Forum News and General Information 1
L Re-calibrate after software update? Calibration Frequency (Interval) 3
J To Calibrate or not to calibrate electronics rack components Manufacturing and Related Processes 5
T How to Calibrate Taper Gages General Measurement Device and Calibration Topics 1
N Calibrate temperature on refrigerated Centrifuges? General Measurement Device and Calibration Topics 2
B Can we calibrate a standard pin with a gauge? General Measurement Device and Calibration Topics 2
D Lesser expensive to calibrate various test weights General Measurement Device and Calibration Topics 5
A How to Calibrate Laser Beam Tester & Radiation Monitor General Measurement Device and Calibration Topics 3
D How to Calibrate and Verify a Heating Oven General Measurement Device and Calibration Topics 3
P How do you calibrate a VNA (Vector Network Analyzer) cal kit? General Measurement Device and Calibration Topics 4
V How to Calibrate Ring Force Gauges Service Industry Specific Topics 1
bobdoering Who says you can't calibrate a steel rule! Funny Stuff - Jokes and Humour 14
S How can I Calibrate a Biosafety Cabinet? General Measurement Device and Calibration Topics 1
C How to Calibrate a Viscosity Tester General Measurement Device and Calibration Topics 3
0 Is it possible to use a threaded ring gage to calibrate the M8 gage? Calibration and Metrology Software and Hardware 4
C How to Calibrate a Medical Infrared Thermometer General Measurement Device and Calibration Topics 3
T How do I calibrate NPT Thread Plug Gages? General Measurement Device and Calibration Topics 9
E Do you need a Certification to Calibrate Gauges? General Measurement Device and Calibration Topics 2
B Calibrate Hole Presence Detection Device? General Measurement Device and Calibration Topics 5
J What do you use to Calibrate 40" Calipers General Measurement Device and Calibration Topics 10
A Guideline to Calibrate Chamber Temp. & Pressure of a Medical Steam Sterilizer General Measurement Device and Calibration Topics 8
S How to calibrate AOI (Automated Optical Inspection) for IOPQ Qualification and Validation (including 21 CFR Part 11) 3
S Laboratory to Calibrate Objective Micrometer needed General Measurement Device and Calibration Topics 4
C Tolerance Estimation of a Load Cell I want to Calibrate General Measurement Device and Calibration Topics 9
D Procedure on How to Calibrate a External Micrometer General Measurement Device and Calibration Topics 1
B ISO/TS 16949 Section 7.6.3 - Use of Manufacturer to Calibrate IATF 16949 - Automotive Quality Systems Standard 21
W How to verify/calibrate a Neptune Water Meter General Measurement Device and Calibration Topics 4
N Laboratory that is able to Calibrate Dew Point Meter up to 60?Cdp - Needed General Measurement Device and Calibration Topics 1
C How to Calibrate and Verify Torque Multiplier Gages General Measurement Device and Calibration Topics 6
C How to calibrate or control patterns for fiberglass General Measurement Device and Calibration Topics 4
N Scale with Internal Calibration - Do I need to calibrate? How to document this? Calibration Frequency (Interval) 8
I Need help finding an Accredited Lab that can Calibrate Brunson 76-RH Survey Transits General Measurement Device and Calibration Topics 5
M Controlling Temperature and Time - Do I need to calibrate the thermostat? IATF 16949 - Automotive Quality Systems Standard 4
D How much Tolerance do I use to Calibrate a 150kgs capacity Scale? General Measurement Device and Calibration Topics 5
M To Calibrate or Not to Calibrate various Weight Scales ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 24
N Thermocouple Inhomogeneity - Need to calibrate a roll of 50m thermocouple type K General Measurement Device and Calibration Topics 3
R Traceability for Mercury Argon Calibration Source to Calibrate a Spectrometer? General Measurement Device and Calibration Topics 5
P How to Calibrate Laser Micrometer when no Calibration Company in the Country? General Measurement Device and Calibration Topics 3
S 3 Point Calibration or Calibrate to Extreme High and Low? General Measurement Device and Calibration Topics 9
C Seeking a holder/jig to calibrate large OD micrometer calipers (up to 23") General Measurement Device and Calibration Topics 8
B Difference between Calibration & Verification and How to calibrate Go/No Go gauge? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4