Is it possible to avoid the OQ?

Kid006

Starting to get Involved
#1
We receive a new CNC machine from a different plant and since we are certified in iso13485 process validation should be performed,
the other plant is not in the medical device branch so a validation was not performed.
The IQ has been completed but specifically this machine only have a speed parameter which does not have a reference unit of measure it is presented as a percentage, also the parameter was locked by the supplier.
am i able to jump from the IQ to the PQ? (avoiding the OQ with a justification)
Our products are Foam Positioners, Medical dev 1
Any guidance regarding this issue will be very helpful
Thanks
 
Elsmar Forum Sponsor

Tidge

Trusted Information Resource
#2
To try to prevent any potential confusion, I will point out that historically "OQ" has meant different things for device and pharmaceutical manufacturers. This thread appears to be (and my response is) relevant to device manufacturing, where OQ is the 'operational qualification' of a process 'at the limits of the process'. The OQ generates evidence that the process makes conforming parts no matter what the machine settings are set at, as long as those settings are within the allowed limits. This data can serve as feedback about the chosen limits for the process.

My opinion:

If the process validation never occurred, then an OQ is called for... even if there is nothing (obvious) that an operator can change.

If the replaced equipment is identical, and the IQ is robust enough (e.g. maintenance and calibration methods) to pin-down potential variables that could exist between the machines; I think it can be rationalized to 'skip' an OQ because you are no longer trying to analyze/establish process limits, provided that the PQ provides the necessary evidence that the operator/material variability still results in a process under control.

Personally, I would still do a 'sanity check OQ' and then rationalize using the data from an (abbreviated) OQ as part of the PQ. It's just my experience, but skipping an OQ tends to raise more eyebrows than a PQ with a slightly wonky study design. I also dislike finding issues in the first step after a 'skipped step'.
 

Kid006

Starting to get Involved
#3
The OQ generates evidence that the process makes conforming parts no matter what the machine settings are set at, as long as those settings are within the allowed limits. This data can serve as feedback about the chosen limits for the process.
I think it can be rationalized to 'skip' an OQ because you are no longer trying to analyze/establish process limits,
IQ is robust enough and the only setting in the machine is the cutting speed therefore the variability between doing parts at 1% or 100% (speed) are very minimal that our measuring instruments cannot detect it

skipping an OQ tends to raise more eyebrows than a PQ with a slightly wonky study design.
I was thinking in running the OQ with several combinations of speeds and the same program to desmotrate that the variability is minimum, but i dont know is that will be enough to conclude the OQ
 

yodon

Staff member
Super Moderator
#4
I agree with @Tidge - especially the raising eyebrows part! First off, though, what do YOUR procedures say? Do your procedures allow OQ to be skipped?

IQ/OQ/PQ is just *an* approach; not THE approach. You just need to be prepared to defend that whatever results you gather demonstrate the equipment can consistently perform to your needs.
 

Bev D

Heretical Statistician
Staff member
Super Moderator
#5
I find it impossible to believe that speed is the only variable setting. Remember that it’s not just about the equipment but the entire process that utilizes the equipment. Are the parts clamped onto a fixture? Is there a torque to that clamping? What are you cutting - whaat type of equipment is this a lathe? A mill? These ar only a few of the questions we could ask...
 

Tidge

Trusted Information Resource
#6
If the machine is a CNC, depending on the actual machine I wouldn't necessarily try to challenge fixture positioning or material holding in an OQ, but rather allow such things to be considered as part of operator/raw material variability in the PQ. If the PQ results indicate potential issues perhaps it would be necessary to implement specific control mechanisms (with limits) that could be challenged as part of an OQ. Even with limited variables in an OQ, I still recommend doing one, partially because within an OQ you can explicitly collect more data that you might otherwise not collect as part of a PQ (which sometimes can default to pass/fail acceptance criteria). For example, within an OQ you can study the variability of certain dimensions across an entire lot, whereas some typical machining operations can only include relatively simple things like examining the first/last/random-middle pieces.

Story time: I was once involved with the review of a manufacturing process with wide variability in the outputs... there was so much variability in the outputs that only one operator could "do it right". It took entirely too long to convince management that in order to reduce the variability in the outputs that it was necessary to enforce tighter controls in the manufacturing process (based on the 'good' employee's choices), and then challenge those limits to guarantee better output.
 

somashekar

Staff member
Super Moderator
#7
We receive a new CNC machine from a different plant and since we are certified in iso13485 process validation should be performed,
the other plant is not in the medical device branch so a validation was not performed.
The IQ has been completed but specifically this machine only have a speed parameter which does not have a reference unit of measure it is presented as a percentage, also the parameter was locked by the supplier.
am i able to jump from the IQ to the PQ? (avoiding the OQ with a justification)
Our products are Foam Positioners, Medical dev 1
Any guidance regarding this issue will be very helpful
Thanks
I have two questions to ask you before going about further...
1. How was your part made earlier.
2. Is the new CNC locked for a specific purpose making it a SPM
 

somashekar

Staff member
Super Moderator
#9
The machine is basically a spinning blade guided by a program to cut foam in certain shapes
From all that what you say, you would not need to validate this process, if you can verify your output part by a valid means of measurement or use the part if it passes a valid gauge that you have specifically made for this and is duly verified.
 
Thread starter Similar threads Forum Replies Date
G Special 510K possible? Only the packaging has changed Other US Medical Device Regulations 0
supadrai Possible to Submit a 510(k) if you have never, do not and cannot produce the device for which you're submitting the application? US Food and Drug Administration (FDA) 4
J Possible to get ISO 13485 certified with only OEM Product? ISO 13485:2016 - Medical Device Quality Management Systems 4
D Where does "as far as possible" stop? FMEA - EN 14971 ISO 14971 - Medical Device Risk Management 29
M Informational EU – Possible EUDAMED delay Medical Device and FDA Regulations and Standards News 3
M Is it possible to get iso 13485:2016 certified as a one man band ISO 13485:2016 - Medical Device Quality Management Systems 1
Ashok sunder Is it possible to reduce Risk likelihood and impact Post control Ranking after corrective action taken for risk? FMEA and Control Plans 1
dhuot1990 USP 41 Hanging Scale - Is it possible to make a hang scale adhere to USP41 Standards? ISO 17025 related Discussions 3
F Apple Watch is not a medical device. How can this be possible? CE Marking (Conformité Européene) / CB Scheme 1
qualprod ISO 9001:2015 Clause 8.3 - Is it possible to exclude one product but include it in other? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
Ashok sunder Is it possible to reduce FMEA Occurrence and Detection Ranking after corrective action taken for customer complaints? FMEA and Control Plans 6
B Integrated management system - ISO 9001 and ISO 14001 - Is possible to have a joint certificate covering both standards? ISO 14001:2015 Specific Discussions 6
E Is it possible to start an ISO 9001 QMS from scratch (Really) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
M Some possible misunderstandings on the application of ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 10
P Is it possible to make an educated decision using a very very small sample size? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
a_bardi Is it possible to exclude internal audit from qms scope? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
G ISO 9001:2015 - Is control possible with Excel adaptable measurement sheets? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 24
S Indications for Use - Is it possible to have a different indication for use EU Medical Device Regulations 1
C %GRR <10% not possible. How to get approval by auditor? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 15
I Is risk acceptability really needed if all risks must be reduced as far as possible? ISO 14971 - Medical Device Risk Management 6
M Is Automated Internal Auditing Possible? Internal Auditing 13
A Integrated Management Systems - ISO 50001 and ISO 9001 possible? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
P Is it possible to receive EASA STC within 48 hours after modification? EASA and JAA Aviation Standards and Requirements 1
3 Israel: 1 Medical Device, Multiple Registrations = Possible? Other Medical Device Regulations World-Wide 3
Marc Yahoo probes possible huge data breach - BBC News - 2016 World News 1
Marc Yahoo probes possible huge data breach World News 0
S Preparing a Root Cause Report - Possible Root Causes Document Control Systems, Procedures, Forms and Templates 11
M Future of ISO 14971 - Possible revision? ISO 14971 - Medical Device Risk Management 9
M Is IEC 60601 compliance possible when only generic power-supply is specified? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M Is it possible to have more than one authorized representative in Saudi Arabia? Other Medical Device Regulations World-Wide 6
Y Is it possible to check the authenticity of a CFDA registration certificate? China Medical Device Regulations 7
T Reduced Audit days when UPGRADING from AS9100 to TS16949 - Is it possible? IATF 16949 - Automotive Quality Systems Standard 2
Ajit Basrur Is Biocompatibility study performed on medical devices for possible contaminants? Other Medical Device Related Standards 4
B Subsidiary - No Managing Director - Is ISO 9001 Certification Possible? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
R Is it possible to randomly select samples from 1 column to many column Using Minitab Software 2
Marc Google Fiber chooses nine metro areas for possible expansion After Work and Weekend Discussion Topics 1
Hershal ILAC is preparing to start the possible revision to ISO/IEC 17025 ISO 17025 related Discussions 2
0 Is it possible to use a threaded ring gage to calibrate the M8 gage? Calibration and Metrology Software and Hardware 4
A Is it possible for STC application before issue of TC ? EASA and JAA Aviation Standards and Requirements 4
B Is it possible that a virus sends spam on Facebook? After Work and Weekend Discussion Topics 6
Ronen E Incretin Mimetic Drugs for Type 2 Diabetes: Possible Increased Risk Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
R Is it possible for medical power supply to declare CE based on MDD? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
S Is it possible to have a General Change Control SOP and Form? Document Control Systems, Procedures, Forms and Templates 6
J Quality Management Process Model Possible Interactions and Effects ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
M Is it possible to have two QMSs? One for each or our two products EU Medical Device Regulations 9
T Is "World Class" Quality possible by Accident? Coffee Break and Water Cooler Discussions 18
J Two Quality Management Systems - ISO 13485 in one company possible? Quality Manager and Management Related Issues 4
Marc Great Lakes - 24 foot waves possible! October 2012 Coffee Break and Water Cooler Discussions 1
R Possible Actions for Nonconforming Product besides issuing NC Nonconformance and Corrective Action 6
M Is it possible to get Natural Tolerance (Tn) with Non Normal Distribution? Statistical Analysis Tools, Techniques and SPC 9

Similar threads

Top Bottom