Is it possible to obtain FDA clearance (510k) without meeting QSR?

#1
If someone wanted to design a device and obtain clearance only to immediately sell it to another legal manufacturer. Is it possible to obtain FDA 510k clearance on a medical device product following FDA 820.30 and not have a quality system in place (meeting the rest of the QMS)? The DHF would be a paper-based binder with wet signatures.

What if the classification was de novo?
 
Elsmar Forum Sponsor

shimonv

Trusted Information Resource
#2
It doesn't work. No serious organization will buy a DHF from an organization without QMS. Most likely, you will end up building and backdating QMS and missing records. Backdating is illegal.
Do it right from the beginning and develop your QMS in parallel to product development.

Shimon
 
#3
It doesn't work. No serious organization will buy a DHF from an organization without QMS. Most likely, you will end up building and backdating QMS and missing records. Backdating is illegal.
Do it right from the beginning and develop your QMS in parallel to product development.

Shimon

This reads like an opinion. Can you please cite some supporting regulation? The FDA's 510k checklist does not indicate you must have a QMS.

I do believe one should have a QMS in place, however, this appears to only be an opinion / best practice, not a requirement from what I can find. I'm trying to find where the regs explicitly state you must have a QMS.

There is no need to post date or do anything wrong. You write the DDP, sign and date it, then execute it; all under 820.30. The DDP will need to contain the design control procedure content versus having QMS procedures.

I also disagree that no organization would buy it. Are you of the age that can remember paper based binder DHFs?
 

shimonv

Trusted Information Resource
#4
scda0505,
I am older than I look :) and I've done such projects in the past, and I'm no stranger to paper based DHF.
A typical 510k clearance letter says "[...] You must comply with all the Act's requirements, including, but not limited to: registration and listing (21 CFR Part 807); labeling (21 CFR Part 801); medical device reporting (reporting of medical device-related adverse events) (21 CFR 803) for devices or postmarketing safety reporting (21 CFR 4, Subpart B) for combination products; good manufacturing practice requirements as set forth in the quality systems (QS) regulation (21 CFR Part 820) for devices or current good manufacturing practices (21 CFR 4, Subpart A) for combination products; and, if applicable, the electronic product radiation control provisions (Sections 531- 542 of the Act); 21 CFR 1000-1050."

Design controls is not a miniature self-contained QMS; it interacts with the other sub-system.

Theoretically it's possible, but practically - not recommend.

Shimon
 

mihzago

Trusted Information Resource
#5
If a company is new and has no products on the market, then requirements from parts 803, 806, 807 and some areas of the QSR don't really apply because the company has not marketed any products yet and has very limited operations.

While important and certainly a good idea, QMS compliance is far from a show-stopper for investors, unless it's a higher risk product. And, for a 510(k) or De Novo submission, you certify compliance with Design Controls, not the entire QMS.

There are many start-up that obtained a 510(k) without a QMS in place. For some it's a business strategy to obtain clearance with no plans to market the device.
 

Orca1

Involved In Discussions
#6
If someone wanted to design a device and obtain clearance only to immediately sell it to another legal manufacturer. Is it possible to obtain FDA 510k clearance on a medical device product following FDA 820.30 and not have a quality system in place (meeting the rest of the QMS)? The DHF would be a paper-based binder with wet signatures.

What if the classification was de novo?
It is not possible to obtain FDA 510(k) clearance on a medical device product following FDA 820.30 without having a quality system in place that meets the rest of the Quality Management System (QMS) requirements. The Quality System (QS) Regulation requires manufacturers to establish and follow quality systems to ensure that their products consistently meet applicable requirements and specifications (FDA CP 7382.845, A. THE QUALITY SYSTEM (QS) REGULATION, 3).

For a de novo classification, FDA may inspect relevant facilities to determine if the implementation of Quality System Regulation (part 820 of this chapter) requirements, in addition to other general controls and any specified special controls, provide adequate assurance that critical and/or novel manufacturing processes produce devices that meet specifications necessary to ensure reasonable assurance of safety and effectiveness (21 CFR 860.240(c)).

Regarding the Design History File (DHF), it should document or reference the documentation of the activities carried out to meet the design plan and requirements of Sec. 820.30. A DHF is necessary for each type of device developed (Medical Devices; Current Good Manufacturing Practice (CGMP) Final Rule; Quality System Regulation, 61 FR 52602 (V)(C)(91)). The complete history of the design process should be documented in the DHF (Medical Devices; Current Good Manufacturing Practice (CGMP) Final Rule; Quality System Regulation, 61 FR 52602 (V)(C)(92)).

If you want to sell the device to another legal manufacturer after obtaining clearance, you must ensure that the FDA-assigned premarket submission number of the cleared premarket notification is included in the device listing (21 CFR 807.25(g)(4)).
 
Thread starter Similar threads Forum Replies Date
S Possible termination of MDD Certificate CE Marking (Conformité Européene) / CB Scheme 4
M It is Possible to apply a wall plug power adapter to provide power from the actuators of a dialysis chair or hospital bed? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
MaHoDie Is it possible to assign medical software to security class A (according to IEC 62304)? EU Medical Device Regulations 8
B SAE J 524 Pressure Test/Possible conflicting requirements? Manufacturing and Related Processes 0
M Possible topic for Master thesis topic - MDR and QMS ISO 13485:2016 - Medical Device Quality Management Systems 4
M Is complete testing required as per ISO 10993 for materials used in orthopedic implants or is literature review route possible Other Medical Device Related Standards 3
S Operation of possible rework in control plan FMEA and Control Plans 8
J Impact Testing--Is conversion between Izod and Sharpy Possible? General Measurement Device and Calibration Topics 3
G One-person QMS - possible? ISO 13485:2016 - Medical Device Quality Management Systems 18
L Significant subcontractor update, still under MDD Certification, possible? CE Marking (Conformité Européene) / CB Scheme 3
C Narrow definition of IMTE or calibration possible? ISO 13485:2016 - Medical Device Quality Management Systems 8
L Reduce risks as far as possible - Quartz Crystal ISO 14971 - Medical Device Risk Management 11
J EC Certification transfer from CMO to Legal Manufacturer Possible? CE Marking (Conformité Européene) / CB Scheme 0
K Silver Plating corroding possible during SMT Manufacturing and Related Processes 10
K Is it possible to avoid the OQ? ISO 13485:2016 - Medical Device Quality Management Systems 20
G Special 510K possible? Only the packaging has changed Other US Medical Device Regulations 0
supadrai Possible to Submit a 510(k) if you have never, do not and cannot produce the device for which you're submitting the application? US Food and Drug Administration (FDA) 4
J Possible to get ISO 13485 certified with only OEM Product? ISO 13485:2016 - Medical Device Quality Management Systems 4
D Where does "as far as possible" stop? FMEA - EN 14971 ISO 14971 - Medical Device Risk Management 29
M Informational EU – Possible EUDAMED delay Medical Device and FDA Regulations and Standards News 3
M Is it possible to get iso 13485:2016 certified as a one man band ISO 13485:2016 - Medical Device Quality Management Systems 1
Ashok sunder Is it possible to reduce Risk likelihood and impact Post control Ranking after corrective action taken for risk? FMEA and Control Plans 1
dhuot1990 USP 41 Hanging Scale - Is it possible to make a hang scale adhere to USP41 Standards? ISO 17025 related Discussions 3
F Apple Watch is not a medical device. How can this be possible? CE Marking (Conformité Européene) / CB Scheme 1
qualprod ISO 9001:2015 Clause 8.3 - Is it possible to exclude one product but include it in other? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
Ashok sunder Is it possible to reduce FMEA Occurrence and Detection Ranking after corrective action taken for customer complaints? FMEA and Control Plans 6
B Integrated management system - ISO 9001 and ISO 14001 - Is possible to have a joint certificate covering both standards? ISO 14001:2015 Specific Discussions 6
E Is it possible to start an ISO 9001 QMS from scratch (Really) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
M Some possible misunderstandings on the application of ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 10
P Is it possible to make an educated decision using a very very small sample size? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
a_bardi Is it possible to exclude internal audit from qms scope? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
G ISO 9001:2015 - Is control possible with Excel adaptable measurement sheets? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 24
S Indications for Use - Is it possible to have a different indication for use EU Medical Device Regulations 3
C %GRR <10% not possible. How to get approval by auditor? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 15
I Is risk acceptability really needed if all risks must be reduced as far as possible? ISO 14971 - Medical Device Risk Management 6
M Is Automated Internal Auditing Possible? Internal Auditing 13
A Integrated Management Systems - ISO 50001 and ISO 9001 possible? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
P Is it possible to receive EASA STC within 48 hours after modification? EASA and JAA Aviation Standards and Requirements 1
3 Israel: 1 Medical Device, Multiple Registrations = Possible? Other Medical Device Regulations World-Wide 3
Marc Yahoo probes possible huge data breach - BBC News - 2016 World News 1
Marc Yahoo probes possible huge data breach World News 0
S Preparing a Root Cause Report - Possible Root Causes Document Control Systems, Procedures, Forms and Templates 11
M Future of ISO 14971 - Possible revision? ISO 14971 - Medical Device Risk Management 9
M Is IEC 60601 compliance possible when only generic power-supply is specified? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M Is it possible to have more than one authorized representative in Saudi Arabia? Other Medical Device Regulations World-Wide 6
Y Is it possible to check the authenticity of a CFDA registration certificate? China Medical Device Regulations 7
T Reduced Audit days when UPGRADING from AS9100 to TS16949 - Is it possible? IATF 16949 - Automotive Quality Systems Standard 2
Ajit Basrur Is Biocompatibility study performed on medical devices for possible contaminants? Other Medical Device Related Standards 4
B Subsidiary - No Managing Director - Is ISO 9001 Certification Possible? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
R Is it possible to randomly select samples from 1 column to many column Using Minitab Software 2

Similar threads

Top Bottom