Is Notified Body (NB) audit required for Class I imaging device ?

pkost

Trusted Information Resource
#11
There was a recent case which progressed through the european courts that confirmed that the manufacturers intended use is the intended use.

Therefore if the manufacturer does not claim measuring, it shoudl not be assessed as such - because it is not!

Guidance on the application of "measuring" functions (MEDDEV 2.1/5) gives 3 criteria for a product to be a measuring device....

1. The device is intended by the manufacturer to measure
2. The result of the measurement is displayed in legal units, or other acceptable units, or is compared to at least one point of reference indicated in legal units etc etc
3. the intended purpose implies accuracy, claimed explicitly or implicity ... ...


If a imaging device by a manufacturer is not intended to meausure and there is no scale to make a measurement, it is not appropriate to declare as a measuring device.
 
Elsmar Forum Sponsor

sagai

Quite Involved in Discussions
#12
I must admit it is a surprise for me that any imaging device could be a Class I kind of thing and I deeply agree with Miregmgr.

Can we ask the OP what does this device actually do for imaging though?

Cheers!
 

sagai

Quite Involved in Discussions
#13
Measuring can be a simple yes/no regardless there is no explicit measuring function for the imaging application.
One obvious case to be mentioned, aneurysm is either on the image or not on the image, simple is that without any explicit measuring function it drives the physician to cut the patient. Moreover measuring can be relative measuring based on reference body particle on the same image.
Cheers!
 
M

MIREGMGR

#14
For discussion:

Those with medical imaging technical knowledge will be aware that anatomical-feature image elements may be displayed at their actual size, or larger or smaller relative to the anatomy around them, according to the incident beam source size, technical characteristics of the detector, and either intentional response nonlinearity of a film/phosphor or fluorophosphor display to provide edge sharpening, or quantitative processing of the image data from an electronic detector prior to display so as to similarly provide edge sharpening.

Edge sharpening processes have long been understood to be medically useful, in greatly easing the image reader's task of discerning the medical significance of minor density gradations...but edge sharpening processes always have the potential to cause anatomical features to be displayed not at true size.

The most important features to be diagnostically evaluated in mammographic imaging...cysts, duct conditions, self contained and invasive carcinomas...present as features on images. A key part of diagnosis is size scoring. The image process itself doesn't measure sizes, but it must present images that display features at close to true size, notwithstanding whatever edge sharpening technology is utilized, so that the diagnostician can correctly evaluate, measure and thus score the observed features.

My understanding is that when FDA evaluates a mammographic imaging system 510(k) for acceptable performance, they give heavy consideration to proof that the resulting images support correct measurement. I don't know if a parallel approach is utilized in the EU, but I would think that would be the case.

It's true that the diagnostician with his or her transparent-Mylar measuring scale is performing the direct measurement, but that measurement process will be useless if the imaging system displays features as larger or smaller than their actual size, because the clinician has no way to separately obtain true-size information. Thus the patient may be mistreated and harmed.

Similar considerations apply, at least with FDA, to CT systems that are used, not to create visual diagnostic displays, but rather to create 3D measured-dimensional data of a tumor location relative to marker features on a patient...that data then later being used to directly control the beam scanning process of an X-ray, E-beam, proton or neutron scanner that destructively irradiates the tumor while sparing the surrounding anatomy.

If the CT scanning and data creation process gets the measurements wrong, the tumor will not be correctly treated, and the patient will be harmed by destruction of healthy anatomy.

My understanding is that in these and other imaging contexts, FDA regards imaging as inherently a measurement process, irrespective of whether there is a calibrated scale in the particular device.
 

Ronen E

Problem Solver
Staff member
Moderator
#15
Measuring can be a simple yes/no regardless there is no explicit measuring function for the imaging application.
One obvious case to be mentioned, aneurysm is either on the image or not on the image, simple is that without any explicit measuring function it drives the physician to cut the patient. Moreover measuring can be relative measuring based on reference body particle on the same image.
Cheers!
I think you're confusing measuring with diagnosing. As far as I understand measuring must yield an absolute numeric result (excluding Boolean 0 & 1).
 

Ronen E

Problem Solver
Staff member
Moderator
#16
...

It's true that the diagnostician with his or her transparent-Mylar measuring scale is performing the direct measurement, but that measurement process will be useless if the imaging system displays features as larger or smaller than their actual size, because the clinician has no way to separately obtain true-size information. Thus the patient may be mistreated and harmed.

Similar considerations apply, at least with FDA, to CT systems that are used, not to create visual diagnostic displays, but rather to create 3D measured-dimensional data of a tumor location relative to marker features on a patient...that data then later being used to directly control the beam scanning process of an X-ray, E-beam, proton or neutron scanner that destructively irradiates the tumor while sparing the surrounding anatomy.

...

My understanding is that in these and other imaging contexts, FDA regards imaging as inherently a measurement process, irrespective of whether there is a calibrated scale in the particular device.
I can see the importance, but aren't these aspects already covered by the requirement that the device will perform it's intended use (e.g. obtaining real-size image) safely and effectively, e.g. within X% of real-size in the examples above?

I'm not going to argue whether or not this is the actual FDA's practice because I have no such specific experience (and to me, this whole discussion is less relevant to the USA (1) because the OP enquired about the EC and NB involvement, and (2) the USA system doesn't have the equivalent of "1m" class); I only express my opinion that this is a too-conservative approach and actually a redundant measure (pun not intended).

Cheers,
Ronen.
 

Ronen E

Problem Solver
Staff member
Moderator
#18
Would you consider a digital thermometer saying "yes" for fever (above 37 Celsius) and "no" for anything under 37 Celsius as a non measuring device?
Anyway, this is off topic already because Class I device can have a measuring function, please find:
http://ec.europa.eu/health/medical-devices/files/meddev/2_1_6_ol_en.pdf

Cheers!
That thermometer actually conveys an absolute numerical measurement result, even though it does not display a numerical figure. Definitely that result is present in the internal workings of that thermometer. I bet that the resulting accuracy and precision of those workings will be scrutinized up close in any regulatory clearance process. It's quite different to the general notion of "providing a true visual represantation of an internal body structure / organ". How would you pin-point the accuracy and precision of that?

Of course there are class 1 devices with a measuring function, I don't think anyone in this thread suggested otherwise. The question is whether the NB has the authority (not just the technical ability) to force an "intended" measuring function when it's actually NOT intended by the manufacturer. It's mostly interesting in the case of class 1 devices, becuase in that case it makes a difference (requiring / not requiring NB involvement). With >1 class devices it makes a lesser difference, at least in the context of the OP's question.

Cheers,
Ronen.
 

sagai

Quite Involved in Discussions
#19
So what is a digital image?
It shows the detected electromagnetic ray intensity as numerical values based on algorithms built in the software and displays this intensity as an image.
The point is regardless we see "only" an image, it is a set of numerical values anyway.
With other words, the equivalent output of the image also could be a matrix of numbers that represents converted numerical values of this projected intensity, it is not a question whether or not it measures something (because of course it does, the intensity itself) rather a question of how does it to be displayed for human.

However I am still curious to know more about this imaging application that is deemed to be class I.

Cheers!
 

Ronen E

Problem Solver
Staff member
Moderator
#20
So what is a digital image?
It shows the detected electromagnetic ray intensity as numerical values based on algorithms built in the software and displays this intensity as an image.
The point is regardless we see "only" an image, it is a set of numerical values anyway.
With other words, the equivalent output of the image also could be a matrix of numbers that represents converted numerical values of this projected intensity, it is not a question whether or not it measures something (because of course it does, the intensity itself) rather a question of how does it to be displayed for human.
This is what I call wordsmithing. I give up.
 
Thread starter Similar threads Forum Replies Date
A NB (Notified Body) Audit of Standards ISO 13485:2016 - Medical Device Quality Management Systems 3
N Customer asking if their notified body can audit us ISO 13485:2016 - Medical Device Quality Management Systems 5
M Measurement Unit Rounding Nonconformity - Notified Body Audit Registrars and Notified Bodies 6
J EU labeling requirement Notified body audit finding EU Medical Device Regulations 3
Q Upcoming ISO13485 Audit from the parent NB (Notified Body) Quality Manager and Management Related Issues 14
C Notified Body MDD Technical Audit of our Technical Construction File EU Medical Device Regulations 2
Ed Panek Notified Body Substantial Change MDD EU Medical Device Regulations 1
M Spain Notified Body Medical Device and FDA Regulations and Standards News 0
M Notified Body Spain Other ISO and International Standards and European Regulations 0
M Spanish Notified Body - Spanish language CE Marking (Conformité Européene) / CB Scheme 2
K MDR Class 1 Self Certification - Competent Authority or Notified Body? EU Medical Device Regulations 6
P Notified Body Approval of Change EU Medical Device Regulations 2
K Adding the Notified Body Number CE Marking (Conformité Européene) / CB Scheme 2
J Conflict of Interest Registrar/Notified Body/Testing House Quality Manager and Management Related Issues 4
R Change Notification - Registrar vs Notified body ISO 13485:2016 - Medical Device Quality Management Systems 1
Watchcat "Disappeared" Linkedin Notified Body Rant EU Medical Device Regulations 2
M Notified Body audits -Class Ia medical device product EU Medical Device Regulations 4
R Notified Body for MDD 1Q20 - Florida Company Registrars and Notified Bodies 4
M Informational EU – New notified body designated under the MDR – NB 2797 BSI Group The Netherlands B.V. Netherlands Medical Device and FDA Regulations and Standards News 0
M Informational EU – New notified body designated under the MDR – NB 1912 – DARE!! Services B.V. – Netherlands Medical Device and FDA Regulations and Standards News 0
N MDR review process by notified body - How many steps exist in the review process EU Medical Device Regulations 0
Ed Panek CAPA against Notified Body? EU Medical Device Regulations 5
M Informational EU – First IVDR designated Notified Body – DEKRA Certification GmbH Medical Device and FDA Regulations and Standards News 0
N Regular meetings with Notified Body EU Medical Device Regulations 13
M Informational EU – New designated Notified Body under the MDR – TÜV Rheinland LGA Products GmbH Medical Device and FDA Regulations and Standards News 0
M Informational New designated Notified Body for Regulation (EU) 2017/745 (MDR) included in Nando – IMQ ISTITUTO ITALIANO DEL MARCHIO DI QUALITÀ S.P.A. Medical Device and FDA Regulations and Standards News 3
M Informational EU – Next Notified Body designated to the MDR 2017/475 is IMQ S.p.A from Italy Medical Device and FDA Regulations and Standards News 0
M Informational EU – DEKRA Certification GmbH is the third Notified Body designated under the MDR Medical Device and FDA Regulations and Standards News 0
JoshuaFroud The withdrawal of Lloyd’s Register MDD & IVDD Notified Body Status CE Marking (Conformité Européene) / CB Scheme 1
M Informational From RAPS – Another Notified Body Bows Out Ahead of EU MDR: ‘Investment Too High’ Medical Device and FDA Regulations and Standards News 2
M Authorized Rep (AR), competent authority, NCA and notified body - Differences EU Medical Device Regulations 9
R How long to wait for Notified Body Response - Potentially significant change to a medical device CE Marking (Conformité Européene) / CB Scheme 6
M Informational New Notified Body designated under the MDR – TÜV SÜD Product Service GmbH Zertifizierstellen Medical Device and FDA Regulations and Standards News 1
shimonv Non-significant notice of change to notified body EU Medical Device Regulations 4
S Difficulty Finding A Notified Body for CE Marking - No Capacity Registrars and Notified Bodies 5
P Notified body located in an other country than manufacturing company ISO 13485:2016 - Medical Device Quality Management Systems 2
M BSI – Want to know more about the Notified Body? Registrars and Notified Bodies 0
M Start-Up Company looking for ISO 13485 Certification Body and Medical Device CE Mark Notified Body Registrars and Notified Bodies 6
M Medical Device News BSI achieves Designation for its Netherlands Medical Devices Notified Body EU Medical Device Regulations 2
C Attestation of QA certificate from foreign (EU) Notified Body Other Medical Device Regulations World-Wide 6
D Notified body - CE Certification Class 2a Products in India CE Marking (Conformité Européene) / CB Scheme 4
M Notified Body Involvement with Self Declared Products EU Medical Device Regulations 7
C QR Code on Medical Device Packaging and Notification to Notified Body EU Medical Device Regulations 1
H Clinical investigation - Role of Notified Body (NB) EU Medical Device Regulations 6
S Certification Body, Registrar, Notified Body - What is the difference? Registrars and Notified Bodies 3
D CE Mark with same NB (Notified Body)? CE Marking (Conformité Européene) / CB Scheme 4
Sam Lazzara Expedited, fast-track Notified Body Technical File reviews EU Medical Device Regulations 3
Sam Lazzara Notified Body CE Project Manager - Worst job ever? EU Medical Device Regulations 6
K Notified Body - Minor Nonconformances - They have withheld all certificates General Auditing Discussions 3
R Non-Conformity Report from Notified Body - MEDDEV Guide is not Legally Binding EU Medical Device Regulations 16

Similar threads

Top Bottom