Is Testing Equipment Re-Validation required if the Test Equipment is Relocated?
Tiffany
Hi mates,
One of the re-validation are required for all machines (for production use) if these machine are re-located. Is these also applied for testing equipment?
As this is happened due to the functions re-location from 1st floor to 3rd floor, this testing equipment were re-locate, any re-validation required? Which document in US FDA regulations or other regulations stated this and is there any guidelines can be referring to?
:thanx:
Tiff
Rob Kellock
If you move test equipment it makes sense that it be at least revalidated, in some cases it should be recalibrated. Obviously the lengths you go to will be dependent on equipment and the risk associated with the tests.
Imagine relying on a piece of test equipment for measuring your products. If the test equipment is out, the measurement is no longer valid and therefore you may be supplying products which are wrong. This wouldn't be a good result against any requirements whether they be dictated by the customer, a standard or legislation.
I
isoalchemist
We get into a language issue on what exactly we mean by revalidated and your industry. Should the system be revalidated / recalibrated to ensure it is functioning correctly in its new location Yes, you need to know it was not broken in the move. Should the system be revalidated to prove it is capable of making the measurement required, No the move won't change its capability if it is working properly.
Rob Kellock
And how do you know if it's working properly unless you validate / calibrate it? I'm not sure this is a language issue, unless you mean understanding the term 'validation'.
Ronen E
Problem Solver
Moderator
Hi mates,
:thanx:
Tiff
21 CFR part 820:
Sec. 820.72 Inspection, measuring, and test equipment.
(a)Control of inspection, measuring, and test equipment. Each manufacturer shall ensure that all inspection, measuring, and test equipment, including mechanical, automated, or electronic inspection and test equipment, is suitable for its intended purposes and is capable of producing valid results. Each manufacturer shall establish and maintain procedures to ensure that equipment is routinely calibrated, inspected, checked, and maintained. The procedures shall include provisions for handling, preservation, and storage of equipment, so that its accuracy and fitness for use are maintained. These activities shall be documented.
(emphasis added)
Should you re-validate? I think so. What should be the extent of the re-validation activity? Start off with a risk analysis, and continue from there. For example, let's say that you have a piece of testing equipment that is very vibration-sensitive. Now, on the 1st floor there were no issues because vibration was below any effective threshold, as shown through your original validation. On the 2nd floor you have a machine that causes just enough vibration felt on the 3rd floor to cross the threshold, so now you do have a testing issue (just an imaginary scenario to make a point).
Perhaps the risk analysis will show that very little is required to maintain that the accuracy and fitness are not likely to be significantly affected, but at least you will have it sorted and documented.
Cheers,
Ronen.
Murphys Law
Is sort of depends on your test. Is it a big complex test environment that requires infrastructure or something portable.
Ronen touched upon the vibration example I was going to use. Another that I remember on a site move, was water pressure differences.
E How do you know your test equipment is good? Functional testing and Golden Boards ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
K Is a 3V torch classified as medical electrical equipment needing EMI/EMC testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Testing an equipment for MRI Compatibility Other US Medical Device Regulations 3
E Testing ME equipment powered by Battery EU Medical Device Regulations 0
E Rich O2 Environment Testing Laboratories (Clause 11.2.2 ME EQUIPMENT and ME SYSTEMS) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M ISO17025 Testing Laboratory Equipment Calibration Requirements ISO 17025 related Discussions 4
C Is there a standard for testing equipment at a single point? General Measurement Device and Calibration Topics 13
A Minimum Testing Requirements for a Clear Protective Equipment Barrier (Class II 510K) Other US Medical Device Regulations 1
P Testing to clause 15.4.6 - Actuating parts of controls of Medical Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Testing Medical Equipment used in conjunction with Oxygen Rich Environments IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
F Help with Electrical Safety Testing - Medical Equipment Regulations IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
6 Fuel Testing Laboratory Test Equipment - Methods and Calibration General Measurement Device and Calibration Topics 2
K Where do you draw the line on what fixtures/testing equipment require calibration General Measurement Device and Calibration Topics 36
M GR&R (Gage R&R) for Automated Electrical Testing Equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
S Calibration of equipment in an A2LA accredited testing laboratory General Measurement Device and Calibration Topics 7
D Sealed bag testing equipment Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
D Does waste water equipment need calibration if we also outsource testing? Miscellaneous Environmental Standards and EMS Related Discussions 12
I Procedure for Calibration of my Load Cell Testing Equipment - Needed General Measurement Device and Calibration Topics 5
Y Is NDC needed for gage used for ATE? Automated Test Equipment IC Testing Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
D Looking for 100% sort testing equipment for overall length, flatness, and parallelism General Measurement Device and Calibration Topics 1
A Measurement standards - Testing equipment for lubricant characteristics - Grease, Oil General Measurement Device and Calibration Topics 2
E 60601-1 - Tilt testing - Tensile safety factor IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
V Setup for testing against ISO14708 clause 16 (protection of the patient from herms caused by heat) Other Medical Device Related Standards 0
gramps What do you think about automated QA testing For software app industry? Misc. Quality Assurance and Business Systems Related Topics 5
B In house NIOSH pre Testing accepted by NIOSH? US Food and Drug Administration (FDA) 1
M Bacteriostasis/Fungistasis Testing Other Medical Device and Orthopedic Related Topics 6
P Sample Size for Distribution Simulation Testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
N EN 813, EN 12277, EN 1497 - Testing some harness prototypes to an EN standard Various Other Specifications, Standards, and related Requirements 0
D Essential performance and EMC immunity testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J Biocompatbility testing on Class 1 device requirements Other US Medical Device Regulations 12
Q Summative Usability Evaluation Testing: prior or during Clinical Investigation? Human Factors and Ergonomics in Engineering 6
B NIOSH Approval for Surgical N95 Respirators - Required testing US Food and Drug Administration (FDA) 2
M ECG lead leakage currents - How to specify ECG leads during electrical safety testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C Stress / Challenge Conditions for Design Verification Testing to Reduce Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
D CFR Title 14: Aeronautics and Space PART 120—DRUG AND ALCOHOL TESTING PROGRAM Federal Aviation Administration (FAA) Standards and Requirements 3
lanley liao Purchase Acceptance Criteria - Tensile testing Oil and Gas Industry Standards and Regulations 2
M Device mounted at IV pole - what about mechanical stability testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J Cochlear Implant Testing 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A Class I (exempt) testing requirements Other Medical Device Related Standards 0
JoCam Electrical Testing for Japan, PSE or CB Scheme Other Medical Device Regulations World-Wide 0
M Who are the go to companies for non-destructive hardness testing? General Measurement Device and Calibration Topics 3
P Testing cloud-based backups IT (Information Technology) Service Management 7
I IATF Lab Scope Testing Qualification and Competency Documentation IATF 16949 - Automotive Quality Systems Standard 3
N Chemical Testing on Medical Devices - Solutions in a container closure system (bag) EU Medical Device Regulations 1
M Comparing data from destructive testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
T Flammability testing Reliability Analysis - Predictions, Testing and Standards 0
E Manufacturers should develop a testing device for covid19 Service Industry Specific Topics 0