Is TGA approval required for a ARTG listed IVD for a sample type addition?
sreenu927
Quite Involved in Discussions
#1
Hi All,
For a Class 2 and 3 IVDs with Australia's TGA approval, if we include additional sample or specimen type, will it require any updated STED file submission to TGA and seek approval before selling it or just a change notification will do?
Thanks,
Sreenu
Elsmar Forum Sponsor
K
kbryde
#2
Hi there,
For our Class 2 IVDs, our sponsor has informed us that when we add new products with same GMDN code and description, we only have to submit an updated DoC along with a reference to the ARTGC it affects. We have never updated our submission with additional documents.
Best Regards
kbryde
sreenu927
#3
Hi Kbryde,
Thanks for sharing your experience. But here in this case, the intended use statement is changing; however, the description in the ARTG is general. Usually, if intended use is changed, then it requires either a change notification or a new registration.
Regards,
Sreenu
medwise
Involved In Discussions
#5
Hi Sreenu,
ARTG entry is for a "kind of medical device" i.e. -
  • Same manufacturer
  • Same Sponsor
  • Same GMDN and
  • Same Classification
If all the above are same new registration is not required.
Hope this helps.
Romit
Ronen E
Problem Solver
Moderator
#6
Hi Sreenu,
  • Same manufacturer
  • Same Sponsor
  • Same GMDN and
Hope this helps.
Romit
That's quite a liberal interpretation. I would verify with the TGA on an individual case basis. Intended use should also be taken into consideration.
Cheers,
Ronen.
sreenu927
#7
Hi Romit and Ronen,
I have checked with TGA on this. They responded that if there is no impact to the ARTG (register), then no notification is required. If a new model or identified is to be included that requires a change to ARTG, then a notification to TGA is required.
But for intended use change, as Ronen mentioned, it is not clear. However, most of the cases, if impact to intended use change, it will require re-registration.
Thanks,
Sreenu
medwise
#8
Hi Sreenu,
Impact to intended use can be determined by GMDN. If the intended use update necessitates consideration of new GMDN term then a new application will be required. Alternatively, if the scope of the intended use has changed but can still be covered by the same GMDN term then a change application can be made.
Kind regards
Romit
Ronen E
Problem Solver
Moderator
#9
Hi Sreenu,
Kind regards
Romit
Hi,
Has this been stated to you by the TGA or have you seen it in a written guidance? IMO the statement above might be a little far-reaching. Sometimes 2 significantly different intended uses might fall under the same GMDN code and at the same time these would be considered two different kinds of devices (even when the manufacturer and sponsor are the same).
Cheers,
Ronen.
Thread starter Similar threads Forum Replies Date
S How to locate a Sponsor in Australia to submit Product Approval Applications by TGA Other Medical Device Regulations World-Wide 5
T Australia TGA Approval recognized by New Zealand? Other Medical Device Regulations World-Wide 7
R US FDA 510k Australia TGA US Food and Drug Administration (FDA) 0
M EU MDR (GSPR Checklist) consolidated with TGA EP Checklist EU Medical Device Regulations 5
J TGA CER & EU MDR CER Requirements for Australia Other Medical Device Regulations World-Wide 3
J TGA CER Other Medical Device Regulations World-Wide 4
L TGA Requirements Other Medical Device Regulations World-Wide 1
A TGA - Annual Report or Periodic Safety Update Report? Other Medical Device Regulations World-Wide 0
S Australia TGA Clinical Evaluation Report (CER) Other Medical Device Regulations World-Wide 0
E In need of a new TGA sponsor - Small software company Other Medical Device Regulations World-Wide 4
M Informational TGA – Submissions received: Regulation of software, including Software as a Medical Device (SaMD) Medical Device and FDA Regulations and Standards News 1
M Informational TGA presentation: Cybersecurity for medical devices Medical Device and FDA Regulations and Standards News 0
M Informational TGA presentation: How to submit an effective good manufacturing practice clearance application Medical Device and FDA Regulations and Standards News 0
M Informational TGA-led IMDRF Personalised Medical Devices working group meets in Canberra Medical Device and FDA Regulations and Standards News 0
M Informational TGA – Current status of breast implant products in Australia Medical Device and FDA Regulations and Standards News 0
M Informational TGA Consultation: Review of the regulation of certain self-testing IVDs in Australia Medical Device and FDA Regulations and Standards News 0
M Informational Australia – TGA business plan 2019-20 Medical Device and FDA Regulations and Standards News 0
M Informational TGA – Australian regulatory action on breast implants and breast tissue expanders Medical Device and FDA Regulations and Standards News 0
M Informational TGA – Essential principles checklist (medical devices) Medical Device and FDA Regulations and Standards News 2
M Informational TGA – Webinar: The cybersecurity of medical devices Medical Device and FDA Regulations and Standards News 0
M Informational TGA Consultation: Proposed clarification of the regulatory requirements for medical device systems and procedure packs Medical Device and FDA Regulations and Standards News 2
M Informational TGA Consultation: Proposed changes to medical device essential principles for safety and performance Medical Device and FDA Regulations and Standards News 0
M Informational TGA – Medical device incident reporting (MDIR) guide Medical Device and FDA Regulations and Standards News 0
M Informational TGA Webinar: Essential Principles to software Medical Device and FDA Regulations and Standards News 0
M Informational TGA – Medical device patient information leaflets and implant cards Medical Device and FDA Regulations and Standards News 0
M Informational TGA – European implementation of Medical Device and IVD Regulations – Implications for Australia Medical Device and FDA Regulations and Standards News 0
M Informational TGA – Advertising health products: Rules about safety claims in advertising Medical Device and FDA Regulations and Standards News 0
M Informational TGA presentation: Digital Devices Webinar 3, 20 June 2019 Medical Device and FDA Regulations and Standards News 0
M Informational TGA – Medical device cyber security guidance for industry Medical Device and FDA Regulations and Standards News 0
M Informational TGA – Guidance on Therapeutic Goods (Conformity Assessment Standard for Quality Management Systems) Order 2019 Medical Device and FDA Regulations and Standards News 0
M Informational TGA – Medical device annual reporting – Sponsor obligations Medical Device and FDA Regulations and Standards News 0
M Informational TGA – Webinar: An update on the consultation for the Regulation of Software, including Software as a Medical Device (SaMD) Medical Device and FDA Regulations and Standards News 0
J Australian TGA Sponsor costs Other Medical Device Regulations World-Wide 3
M Informational TGA presentation: How the TGA regulates software, 7 March 2019 Medical Device and FDA Regulations and Standards News 0
M Informational TGA – Action Plan for Medical Devices Medical Device and FDA Regulations and Standards News 0
M Informational TGA – Q&A: Use of market authorisation evidence from comparable overseas regulatory bodies for medical devices Medical Device and FDA Regulations and Standards News 0
C Sponsor Agreement Requirements for Australia TGA Other Medical Device Regulations World-Wide 7
Al Rosen Australia TGA Medical Device Sponsor Agreement - Example or Template Other Medical Device Regulations World-Wide 28
M Informational TGA – Several proposed changes to classification to better align with the EU MDR Medical Device and FDA Regulations and Standards News 0
M Informational TGA – Brexit – Implications for therapeutic goods in Australia Medical Device and FDA Regulations and Standards News 0
M Informational TGA Consultation: Proposed regulatory scheme for personalised medical devices, including 3D-printed devices Medical Device and FDA Regulations and Standards News 0
M Informational TGA Consultation: Regulation of software, including Software as a Medical Device (SaMD) Medical Device and FDA Regulations and Standards News 0
M Medical Device News TGA Consultation: Good Clinical Practice Inspections Program Medical Device and FDA Regulations and Standards News 0
M Informational Webinar: The role of the TGA in digital health Medical Device and FDA Regulations and Standards News 0
M Medical Device News TGA Consultation: Potential reclassification of active medical devices for diagnosis and patient therapy Medical Device and FDA Regulations and Standards News 0
M Medical Device News TGA Consultation: Changes to a number of definitions and the scope of the medical device regulatory framework in Australia Medical Device and FDA Regulations and Standards News 0
M Medical Device News TGA – Regulator Performance Framework: Self-assessment Report, July 2017 to June 2018 Medical Device and FDA Regulations and Standards News 0
M Medical Device News TGA – Regulation of Software as a Medical Device Medical Device and FDA Regulations and Standards News 0
M Medical Device News TGA update 26-09-18 - Medical Devices Safety Update, Volume 6 September 2018 Other Medical Device Regulations World-Wide 0
M TGA comparable overseas regulators requirements matrix? Other Medical Device Regulations World-Wide 0
Similar threads
Top Bottom