Is there an FDA requirement for a Supplier Audit Schedule
Bunny
Is there an FDA requirement for a medical device manufacturer to have a supplier audit schedule in place? If yes, please point me to the requirement.
mdurivage
Re: Supplier Audit Schedule
There is not a direct requirement, but it does demonstrate that there is a plan. Additionally, it would be difficult to explain to an inspector/auditor that you do not have an audit schedule in place.
mdurivage
If you have an annual supplier assessment program in place, the assessment process should/could lead to the development of a supplier audit schedule.
Bunny
What would occur in the event that many of the suppliers are not audited as indicated by the schedule? I foresee that resource constraints would not allow some audits to take place. Would that cause a problem with the investigator?
mdurivage
Lack of resources is not an acceptable excuse and should never be verbalized to an inspector. If you use a risk based approached and use supplier quality metrics to review/grade your suppliers as a method to develop you audit schedule you should be fine. If you miss an audit due to a scheduling conflict (on your end or the suppliers) document the reason and reschedule.
SuperGirl
I don't be believe you have to audit all your suppliers in the sense of "Hey, we're here to audit you". Do you inspect parts or what ever the suppiler supply as is arrives? If so, are you not doing a sort of "audit" of the suppiler that way?
SuperGirl
Not necessarily, Section 820.50 (Purchasing controls) states
Each manufacturer shall establish and maintain procedures to ensure that all purchased or otherwise received product and services conform to specified requirements.
(a)Evaluation of suppliers, contractors, and consultants. Each manufacturer shall establish and maintain the requirements, including quality requirements, that must be met by suppliers, contractors, and consultants. Each manufacturer shall:
(1) Evaluate and select potential suppliers, contractors, and consultants on the basis of their ability to meet specified requirements, including quality requirements. The evaluation shall be documented.
(2) Define the type and extent of control to be exercised over the product, services, suppliers, contractors, and consultants, based on the evaluation results.
(3) Establish and maintain records of acceptable suppliers, contractors, and consultants
So, you set up your inspection to ensure the supplier is supplying the correct parts and you have defined this to be the extent of your controls (and can justify based on risk), that is all you have to do.
We have many suppliers that we do not “audit”, but we evaluate and have controls set up to satisfy the requirement. It's all about making the system work for your company.
Ed Panek Does this FDA Requirement Apply to international (not USA) distributors for USA based manufacturing companies? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
J Is Device Tracking Card an actual requirement under FDA regulation? Other US Medical Device Regulations 0
D FDA Date Format Requirement US Food and Drug Administration (FDA) 0
I FDA requirement for 0.5 and 5 micron particle monitoring US Food and Drug Administration (FDA) 0
chris1price Is there a regulatory (MDD, MDR, FDA, ISO, etc) requirement to perform a mock recall? CE Marking (Conformité Européene) / CB Scheme 4
N Cry for help - FDA calibration requirement for accuracy & precision General Measurement Device and Calibration Topics 2
R Is it a requirement to notify my customer if I receive a 483 from the FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
K 510 (k) - Requirement to notify the FDA or is only the Notified Bodies required? Registrars and Notified Bodies 1
V FDA regulatory requirement on the lot number convention for medical devices US Food and Drug Administration (FDA) 2
R ISO 13485 vs FDA Requirement question about when ISO 13485 Certification is Required ISO 13485:2016 - Medical Device Quality Management Systems 4
R Is a Quality Manual a requirement for FDA registration? US Food and Drug Administration (FDA) 5
K Is there a requirement for monitoring the FDA's Medwatch and/or Medline US Food and Drug Administration (FDA) 9
M IEC 62366:2007 - Is FDA Usability Engineering Standard a Requirement in Canada? Canada Medical Device Regulations 5
G FDA requirement that 'sister divisions' to be treated as suppliers in 21 CFR 820 ISO 13485:2016 - Medical Device Quality Management Systems 2
G How to determine FDA Product Performance Requirement US Food and Drug Administration (FDA) 5
L FDA GMP requirement: Avoid Non contact with floor? 21 CFR Part 110 US Food and Drug Administration (FDA) 1
C Software Validation Procedure per FDA requirement Qualification and Validation (including 21 CFR Part 11) 3
amjadrana FDA requirement for Minimum Font or Type Size for Labelling 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
L FDA CAPA Requirement for Contract Med. Dev. Manufacturers Other US Medical Device Regulations 4
H EU Validation Requirement vs. FDA Validation Requirement - EN 62304 IEC 62304 - Medical Device Software Life Cycle Processes 16
M FDA Compliant Company - which procedure addresses Customer Order Changes/Requirement? Misc. Quality Assurance and Business Systems Related Topics 5
M CE requirement for software importer / reseller - Class II for FDA and Health Canada EU Medical Device Regulations 1
R System Requirement - FDA - Software Validation on various OSs ISO 13485:2016 - Medical Device Quality Management Systems 13
L Are Fireproof Cabinets a Regulatory Requirement for record storage? FDA or EU Records and Data - Quality, Legal and Other Evidence 23
B What is the difference btw RUO vs IUO for IVD in FDA guidance ? US Food and Drug Administration (FDA) 3
B How to submit Pre-submission to FDA? US Food and Drug Administration (FDA) 4
L FDA & 21 CFR Part 11 Medical Device and FDA Regulations and Standards News 19
B A.I diagnostic software is considered as medical device in FDA? US Food and Drug Administration (FDA) 5
Ed Panek 2020 FDA Top Ten Observations 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M FDA Requirements for Investigational Devices - Clinical Investigation & Shipping Medical Device and FDA Regulations and Standards News 0
B FDA Breakthrough Device can be overlapped with a designated device? US Food and Drug Administration (FDA) 6
pbojsen ISO 13485 Requirements versus FDA product classification and GMP exemptions - Audits ISO 13485:2016 - Medical Device Quality Management Systems 3
S Examples of FDA acceptable Software Design Specification (SDS) Medical Device and FDA Regulations and Standards News 6
F Labelling to comply with both FDA and MDR US Food and Drug Administration (FDA) 6
Watchcat FDA vs NB Fees? Other US Medical Device Regulations 7
K FDA Registration and listing weird situation Medical Device and FDA Regulations and Standards News 4
D FDA Guidance on Computer Software Assurance versus 21 CFR Part 11 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
G Does FDA allows remote approvals of quality documentation. Is there any specific guidance on signing any quality records remotely? Document Control Systems, Procedures, Forms and Templates 1
B Does FDA Registration QSR need to cover non-medical devices for contract repackager? US Food and Drug Administration (FDA) 1
S Transitional Adolescent A and B - "CDRH PREMARKET REVIEW SUBMISSION COVER SHEET FORM FDA 3514" Medical Device and FDA Regulations and Standards News 1
P MSDS for MVQ FDA White, Vinyl Methyl Silicone Rubber EU Medical Device Regulations 4
C RA (Regulatory Assurance) Training (FDA) looking for resources Training - Internal, External, Online and Distance Learning 5