SBS - The best value in QMS software

Is there an FDA requirement for a Supplier Audit Schedule

Elsmar Forum Sponsor
B

Bunny

#12
Yes, it is based upon risk; risk of the device and risk of the components. You can not rely on parts inspection to qualify your vendors.
 
M

MIREGMGR

#14
We (my employer) normally do an on-site periodic audit only of those suppliers that provide us with sterilization services, and/or that do contract/component manufacture of products that have sterile barrier functionality and/or that they sterile barrier package.

We also do site visits by qualified specialist engineers, but not formal on-site audits, for suppliers of sterile barrier packaging materials.

Other suppliers are periodically re-qualified by review of quality records, reference to updated certifications, questionaire responses and the like.

We, like many other companies, were burned by the Pharmaceutical Innovations fiasco...in our case, to a very limited extent because we'd already cut them off due to uneasiness plus the difficulty of working with their owner, but still it was a sobering lesson. We had not done a full audit of them, including review of validations. Part of that was due to their being ISO 13485 certified, and our believing their certificate, and part of it was that what they supplied us was under their name, and we just kit-packed it. And, we had some "shortage of resources, have to prioritize" thinking. We also don't have an in-house deep knowledge of the oddball dry-heat-through-foil sterilization approach they were using, and had been using forever including through previous FDA reviews. Big mistake. We won't do that again.
 
Last edited by a moderator:

TWA - not the airline

Trusted Information Resource
#15
What we do is to have a component control plan for each device. Based on the risk of a component failure for the safety and performance of the device we define what control is needed. Typically class III devices have more critical components than class I, but there are also some class III device components that have very low risk (sometimes lower than the more critical ones of a class I device). Depending on the risk class of the component we may e.g. need to use only suppliers with a quality systems in place, a signed quality agreement and we may have to audit them regularly. Component risk assessment also governs extend and frequency of the incoming inspection.
On the supplier side we have an initial supplier qualification based on a questionaire, general reputation and experience (this may also mean an initial audit) and the quality of the first shipment. Once this is done we regularly assess our supplier (mostly based on the quality they deliver) and based on the results we may need to adapt incoming inspections or the frequency of supplier audits (maybe even down to zero) or eventually change the supplier and qualify a new one.
Generally I think that most of the time it makes little sense to regularly audit suppliers that are 13485 certified; the beef is in the processes and if you are shown a validation you can only start nitpicking on formalities but never really assess whether they have the important aspects under control or not. For essential or critical components however it makes sense to do an initial audit and check whether the cert is just a piece of paper or if your supplier really has a system and knows quality. Once you are sure of that, you can leave the rest to the regular re-certification audits. My :2cents:
 

v9991

Trusted Information Resource
#16
here's another pertinent reference...

Contract Manufacturing Arrangements for Drugs: Quality Agreements
this document provides wider clarity (especially the 4-case studies illustrate)on the binding/ownership of compliance on both owner and contract organization;
4Quality Agreements described in this guidance should also be used by Owners of combination products as they are subject to requirements under 21 CFR 211, 21 CFR 820, or both (see 21 CFR 4.3). In addition to facilitating compliance with requirements under 21 CFR 211, Quality Agreements with Contracted Facilities would also be appropriate for demonstrating compliance, in part, with 21 CFR 820.50 (Purchasing Controls) and with 21 CFR 820.80(b) (Receiving Acceptance Activities), for the combination product..
 

Attachments

Last edited:

Ronen E

Problem Solver
Staff member
Moderator
#17
here's another pertinent reference...

Contract Manufacturing Arrangements for Drugs: Quality Agreements
this document provides wider clarity (especially the 4-case studies illustrate)on the binding/ownership of compliance on both owner and contract organization;
Just to clarify, on p. 1 of the guidance it says:

This guidance does not cover the following types of products: Type A medicated articles and medicated feed, medical devices, dietary supplements, or human tissues intended for transplantation regulated under section 361 of the Public Health Service Act.
(Emphasis added)
 
S

ssarangapani

#18
Re: Supplier Audit Schedule

I don't be believe you have to audit all your suppliers in the sense of "Hey, we're here to audit you". Do you inspect parts or what ever the suppiler supply as is arrives? If so, are you not doing a sort of "audit" of the suppiler that way?
Does not seem to be enough for FDA. Recently they have been strict about supplier audits. Excuses like, we are too small to conduct audit of big companies do not seem to go well with the FDA auditor. However, you can put in place an audit assessment form, that relies on the incoming inspection and say that a particular supplier meets or exceeds your standards.
 

Ronen E

Problem Solver
Staff member
Moderator
#19
Re: Supplier Audit Schedule

Does not seem to be enough for FDA. Recently they have been strict about supplier audits. Excuses like, we are too small to conduct audit of big companies do not seem to go well with the FDA auditor. However, you can put in place an audit assessment form, that relies on the incoming inspection and say that a particular supplier meets or exceeds your standards.
I can't really say how concerned FDA is with the consequences of this specific approach (provided that it really is their policy) for small manufacturers, but they ARE definitely concerned with the well-being of such manufacturers. FDASIA actually made them accountable to that, refer the attachment.
 

Attachments

M

MIREGMGR

#20
Re: Supplier Audit Schedule

(...) small manufacturers, but they ARE definitely concerned with the well-being of such manufacturers.
...right up to the point where the regulated small business offers their small size as an excuse for why they didn't conform to the small-business-accommodating device safety and effectiveness requirements...at which point, the basic requirement comes to the fore: being too small/understaffed/not having the money for the costs of following the rules is not an acceptable excuse.
 
Thread starter Similar threads Forum Replies Date
C Non-sterile reusable surgical instruments - FDA sterilization requirement Other Medical Device Related Standards 2
B FDA requirement for CAPA Signoff ISO 13485:2016 - Medical Device Quality Management Systems 6
J Is Device Tracking Card an actual requirement under FDA regulation? Other US Medical Device Regulations 0
D FDA Date Format Requirement US Food and Drug Administration (FDA) 0
I FDA requirement for 0.5 and 5 micron particle monitoring US Food and Drug Administration (FDA) 0
chris1price Is there a regulatory (MDD, MDR, FDA, ISO, etc) requirement to perform a mock recall? CE Marking (Conformité Européene) / CB Scheme 4
N Cry for help - FDA calibration requirement for accuracy & precision General Measurement Device and Calibration Topics 2
R Is it a requirement to notify my customer if I receive a 483 from the FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
K 510 (k) - Requirement to notify the FDA or is only the Notified Bodies required? Registrars and Notified Bodies 1
V FDA regulatory requirement on the lot number convention for medical devices US Food and Drug Administration (FDA) 2
R ISO 13485 vs FDA Requirement question about when ISO 13485 Certification is Required ISO 13485:2016 - Medical Device Quality Management Systems 4
R Is a Quality Manual a requirement for FDA registration? US Food and Drug Administration (FDA) 5
K Is there a requirement for monitoring the FDA's Medwatch and/or Medline US Food and Drug Administration (FDA) 9
M IEC 62366:2007 - Is FDA Usability Engineering Standard a Requirement in Canada? Canada Medical Device Regulations 5
G FDA requirement that 'sister divisions' to be treated as suppliers in 21 CFR 820 ISO 13485:2016 - Medical Device Quality Management Systems 2
G How to determine FDA Product Performance Requirement US Food and Drug Administration (FDA) 5
L FDA GMP requirement: Avoid Non contact with floor? 21 CFR Part 110 US Food and Drug Administration (FDA) 1
C Software Validation Procedure per FDA requirement Qualification and Validation (including 21 CFR Part 11) 3
amjadrana FDA requirement for Minimum Font or Type Size for Labelling 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
L FDA CAPA Requirement for Contract Med. Dev. Manufacturers Other US Medical Device Regulations 4
H EU Validation Requirement vs. FDA Validation Requirement - EN 62304 IEC 62304 - Medical Device Software Life Cycle Processes 16
M FDA Compliant Company - which procedure addresses Customer Order Changes/Requirement? Misc. Quality Assurance and Business Systems Related Topics 5
M CE requirement for software importer / reseller - Class II for FDA and Health Canada EU Medical Device Regulations 1
R System Requirement - FDA - Software Validation on various OSs ISO 13485:2016 - Medical Device Quality Management Systems 13
L Are Fireproof Cabinets a Regulatory Requirement for record storage? FDA or EU Records and Data - Quality, Legal and Other Evidence 23
D FDA Guidance on Computer Software Assurance versus 21 CFR Part 11 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
G Does FDA allows remote approvals of quality documentation. Is there any specific guidance on signing any quality records remotely? Document Control Systems, Procedures, Forms and Templates 1
B Does FDA Registration QSR need to cover non-medical devices for contract repackager? US Food and Drug Administration (FDA) 1
D FDA Information - Revising the Instructions for Use US Food and Drug Administration (FDA) 0
S Transitional Adolescent A and B - "CDRH PREMARKET REVIEW SUBMISSION COVER SHEET FORM FDA 3514" Medical Device and FDA Regulations and Standards News 1
P MSDS for MVQ FDA White, Vinyl Methyl Silicone Rubber EU Medical Device Regulations 4
S Manufacturing Process FDA FOIA Medical Device and FDA Regulations and Standards News 3
S Manufacturing Process FDA FOIA US Food and Drug Administration (FDA) 4
S Mechanical Test Under FDA Freedom of Information Act Medical Device and FDA Regulations and Standards News 5
A Medical Device Contract Manufacturer - Does the CM need to register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Supplier requirements - Major supplier is a Non-Profit registered with ICCBBA (FDA UDI) Supply Chain Security Management Systems 12
C RA (Regulatory Assurance) Training (FDA) looking for resources Training - Internal, External, Online and Distance Learning 5
E FDA 513(g) Cover Letter US Food and Drug Administration (FDA) 5
S Records - Do's and don't' of record entries (FDA - 21 CFR 820) Records and Data - Quality, Legal and Other Evidence 13
B New Facility register with FDA Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 4
K 510k FDA review, will they accept Biocompatibility result generated using feasibility product lots? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
J FDA regulation on decorative contact lenses Medical Device and FDA Regulations and Standards News 5
P Anyone have an Idea on UAE Medical device registeration- Class B with FDA only Other Medical Device Regulations World-Wide 0
F FDA classification for a mobile app Medical Information Technology, Medical Software and Health Informatics 3
Ed Panek IFU Contact Requirements - FDA and MDD/MDR US Food and Drug Administration (FDA) 1
O Any info on release date of FDA “Computer Software Assurance for Manufacturing and Quality System Software” document? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
P Writing a presubmission to the FDA prior to the De novo submission US Food and Drug Administration (FDA) 4
K FDA - Can we have more than 1 Initial Importer Medical Device and FDA Regulations and Standards News 4
K Two FDA questions regarding UDI (and potentially 803.52 MDR) Other US Medical Device Regulations 0
R Importers - For the FDA Registration/ Listing, is the CMO the importer? Medical Device and FDA Regulations and Standards News 6

Similar threads