Is this Calibration or Verification ?
P
PaulJSmith
#11
I have always tended to fall in line with Tara's and Sebastian's ideas of the definitions. However, I see the merit of Jim's explanation as well.
So, who's right?
BradM
#12
For me gauge verification it is activity where we determine gauge error against standard and just leave it, as there is nothing more we can do. Use when error is within expected limit, scrap when outside. E.g. gage block.
For me calibration it is verification followed by adjustment, where we can move gauge error from area outside expected limit to area inside expected limit and then verify again effectiveness of adjustment. E.g. caliper (close jaws and when it shows 0,01 simply press "Cancel" or "Zero", then open and again close jaws and check if it shows 0,00)
I see some similarity to rework of nonconforming product.
And... you're correct too!! :D
But seriously, it doesn't take long to figure out how confusion can abound.
That's why it really important to have the definitions clearly defined within your system.
BradM
#13
So, who's right?
That's a good question.
Like I stated in my post above, it's really important to make sure you either define them, or reference something like the VIM.
I have sat through many audits where there was 'discussion' regarding what is what. However... there were no findings, because it was defined and consistently applied.
Consistency is the key; which means some training. Training can be touchy on them, because people hold their terms near and dear to their heart. :) So it's not a right/wrong; tomato/tomatoe things; its just a 'line in the sand' position and you move forward.
John Broomfield
#14
Adjusting a measuring device so it is calibrated is not repair. It is rework back to the specified requirement. Repair suggests a design change.
Most auto repair shops get this wrong too unless, of course, they are fitting non-OEM parts.
John
#15
John
Let me try and shed some light. First, The "re" in "rework" suggests that some type of work has been done that was not, in some way or another, successful. The term definitely doesn't fit here.
As for general terminology:
  • People can use whatever words they want to use to describe things so long as everyone who needs to understand the meaning does understand the meaning. The possibility of misunderstanding is reduced by common terminology.
  • Move beyond common dictionary denotation with care. If common definitions suffice, leave them alone and resist the temptation to call spades something other than spades. In this case, the common definition of "calibration" is "comparison to a standard." It does not suggest or necessarily entail adjustment. A device that has been compared to a standard and found to be acceptable in accordance with defined criteria, without adjustment, has been calibrated.
  • Most calibration processes require more than just comparison. As Hershal pointed out, there are things like uncertainty and traceability to be concerned about. These things apply to the general process and not the act of calibration itself.
  • "Verification" is an unnecessarily ambiguous term. Calibration is not necessarily different from verification; calibration is a form of verification. If we have to differentiate between the two, let the definitions be distinct and unambiguous.
  • "Validation" happens, in general, to processes and not individual things. Validation is also a form of verification. Validation consists in verifying that given the application of some defined process settings, parameters or requirements, the process output is reasonably predictable.
dgriffith
#16
I think if we all spoke the language of universal-ese, interpreting terms and definitions would be far less entertaining. :)
Even though the VIM specifically defines calibration, specifically defines adjustment, and specifically implies that they are not the same thing, other countries' standards writing bodies have modified these definitions to align them with standard practices in their country.
Even at the lowest organizational level, companies can define, as part of their calibration procedure, what calibration means and what decisions and actions can be taken based on the results.
So, calibration can establish the error and assign a value with uncertainties, or it can include decision parameters such as in/out of tolerance, and an action such as adjustment to align the observed value with the standard. This is the common US definition.
:2cents:
BradM
#17
:2cents:
:agree1:
And.... you're correct too!! :D
I have been surprised at the different definitions I've run across for calibration:
Measurement without Adjustment; measurement with adjustment; adjustment only; no adjustment; etc. etc.
Second only to the heated debates here about calibrating rulers and tapes, :), some of the more spirited (and honestly interesting) discussions have been on what terms mean in this area.
Hershal
Metrologist-Auditor
#18
Re: Calibration or Validation?
I was with you until the last sentence. Calibration is comparison to a standard--nothing more. There might be, and ususally, additional requirements for the calibration process, but a device verified against a standard has been calibrated.
Jim,
You may disagree, but both the VIM and NIST Handbook 150 make it clear that uncertainties are a required part of calibration. Therefore, no uncertainty, not calibration.
Hershal
Metrologist-Auditor
#19
I was always under the impression that it is verification until you make an ADJUSTMENT. When you make an adjustment to a standard, then you are calibrating.
No necessarily so. For some calibrations there are incidental adjustments. Normally if an adjustment is required it is not really a calibration. The calibration is subsequent to the adjustment.
#20
Jim,
Those are standards that have authority only when two or more persons agree to abide by them. Anyone can devise a set of rules for calibration and define terms anyway they see fit. It becomes a standard when another person or entity agrees to its terms. If we are talking strictly about VIM and the NIST handbook, we use the definitions and requirements therein.
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
T Calibration or Verification -> Cm and Cmk, etc. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
M Calibration or Verification? What terminology to use ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
V Calibration certificate verification Qualification and Validation (including 21 CFR Part 11) 3
J Calibration/Verification Records (IATF 16949 7.1.5.2.1) IATF 16949 - Automotive Quality Systems Standard 5
D How to set equipment calibration/verification criteria General Measurement Device and Calibration Topics 1
E IATF 16949 Cl. 7.1.5.2.1 - Calibration and Verification Records Requirements IATF 16949 - Automotive Quality Systems Standard 1
S Calibration and Verification requirement for foreign EASA Part 145 EASA and JAA Aviation Standards and Requirements 7
J IATF 16949 Cl. 7.1.5.2.1 - Gauges (Measuring Jigs) Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
M Clarification on Calibration/Verification Records 7.1.5.2.1d (IATF 16949) General Measurement Device and Calibration Topics 11
B Calibration/Verification Records - IATF 16949 section 7.1.5.2.1 f) IATF 16949 - Automotive Quality Systems Standard 6
P IATF 16949 Cl. 7.1.5.2.1 "Calibration/Verification Records" Interpretation IATF 16949 - Automotive Quality Systems Standard 12
J Pneumatic & Electrical Torque Tool - Calibration/Verification of 'Power Tools' General Measurement Device and Calibration Topics 2
Q Using Caliper A to verify Caliper B - Verification of Calibration ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
K Verification instead of calibration?(ISO 9001:2008) Calibration Frequency (Interval) 10
K "Verification of Calibration" - CMM arm General Measurement Device and Calibration Topics 6
I Verification or calibration Fluke 8508a with 5720a? TUR not 4:1? Measurement Uncertainty (MU) 1
Q Number of Calibration Verification Points - Calipers, height gages, etc General Measurement Device and Calibration Topics 6
A Calibration/Verification of thermo-hygrometer in storage areas ISO 13485:2016 - Medical Device Quality Management Systems 9
D How Working Standard Calibration result use in Verification of Equipment General Measurement Device and Calibration Topics 7
M Calibration Certification Qualification vs. Verification by Uncertified Staff General Measurement Device and Calibration Topics 6
A Torque Wrench Calibration/Verification General Measurement Device and Calibration Topics 24
G BRC (British Retail Consortium) - Calibration vs. Verification Food Safety - ISO 22000, HACCP (21 CFR 120) 4
G TS-16949 7.6.2 ? Calibration/Verification Records - Assessment of the Impact IATF 16949 - Automotive Quality Systems Standard 5
J Balance Calibration Verification question General Measurement Device and Calibration Topics 9
J Log Book for Analytical Balance Calibration Verification General Measurement Device and Calibration Topics 4
Uriel Alejandro How to Record Evidence for Verification/Calibration of Equipment AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
M Establishing Ranges for Weights for Calibration Verification of a Micro-Balance Calibration Frequency (Interval) 1
H Analytical Balance Calibration and Verification General Measurement Device and Calibration Topics 9
D Precision Balance Verification and Calibration Process General Measurement Device and Calibration Topics 3
D Equipment Verification and Calibration Report Standard General Measurement Device and Calibration Topics 9
P Record Retention requirements of Calibration and Verification Service Providers ISO 13485:2016 - Medical Device Quality Management Systems 11
K Calibration and Verification Records - TS 16949 Clause 7.6.2 General Measurement Device and Calibration Topics 8
P Is it compulsory to send new measuring equipment for calibration / verification? Calibration Frequency (Interval) 3
D Calibration vs. Verification and Measurement Device Labeling General Measurement Device and Calibration Topics 10
M Basis used for Calibration or Verification of Measuring Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
D For Reference Only or Calibration/Verification - Tape Measures for cutting bar stock ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 26
M Methods for Verification/Calibration of Video (Optical) Extensometer General Measurement Device and Calibration Topics 2
C Updating our Laboratory Calibration / Verification Procedures IATF 16949 - Automotive Quality Systems Standard 5
R Verification vs. Calibration and what to do with the results of each General Measurement Device and Calibration Topics 3
B Difference between Calibration & Verification and How to calibrate Go/No Go gauge? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
S Audit and Calibration question - does verification alone meet the requirements? General Auditing Discussions 5
D Calibration vs. Verification - Can we calibrate dial indicators in-house? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
P Internal Calibration or Verification - Personnel Requirements IATF 16949 - Automotive Quality Systems Standard 3
S Verification of calibration certificate Miscellaneous Environmental Standards and EMS Related Discussions 1
R Calibration (Verification) of RF cables General Measurement Device and Calibration Topics 2
M Telescoping Gage - Calibration or Verification General Measurement Device and Calibration Topics 5