Is validation required when consumables are changed

#1
Recently going through a regulatory audit, we were told that our validations were incomplete and not to level required as we were not re-validating the equipment when a consumable was changed (example electrode on a welding machine). As you can understand, this brought along some serious concern as these consumables are typically changed each day or two due to wear or use. Initially our machine was qualified and validated testing the extremes of the machine settings and consumables so I am having a difficult time on why you would need to put your equipment through revalidation each time a consumable was changed.

I would appreciate any feedback from the group.
 
Elsmar Forum Sponsor

Ed Panek

QA RA Small Med Dev Company
Trusted Information Resource
#2
I can think of two ideas

1) A risk-based assessment that these changes have no impact
2) A test where you intentionally replace the consumables and document the results to demonstrate that they have no impact.
 

JJ_FDA

Involved In Discussions
#3
Consider, depending on your risk assessment, requiring some sort of requalification test each time the consumable is replaced. Replaced a burned-out lightbulb at home? Flick the switch on and off after you replaced the bulb to make sure it works before you go on with your day. Saves you from discovering, when it's dark out, that your new lighbulb doesn't work.

Not sure why auditor is calling for revalidation of the whole thing. That seems extreme.
 
Thread starter Similar threads Forum Replies Date
M Is Validation of Plating Processes required and who is responsible? Qualification and Validation (including 21 CFR Part 11) 11
J Is Design Validation required for fixtures or test systems used to manufacture a product? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
Ed Panek Validation Required? Disinfection of wearable patches for unbroken skin ISO 13485:2016 - Medical Device Quality Management Systems 6
L Why is initial production units, lots, or batches, or their equivalents required for Design Validation? US Food and Drug Administration (FDA) 3
K Is Mold qualification / process validation always required? Manufacturing and Related Processes 7
S Is Human Factor Testing required as part of Design Control Validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
A Machine Relocation within Facility - Is re-validation required? ISO 13485:2016 - Medical Device Quality Management Systems 8
A Level of details required for Class IIb Product Verification and Validation CE Marking (Conformité Européene) / CB Scheme 1
S Re-validation NCR - In which cases re-validation of processes is required Oil and Gas Industry Standards and Regulations 6
A Is Aeration Process in ETO Sterilization Validation Required? Manufacturing and Related Processes 7
N When is Medical Device Software Validation required? ISO 13485:2016 - Medical Device Quality Management Systems 6
S Introducing a new peripheral for cleared device - Is Validation required? US Food and Drug Administration (FDA) 3
H Is design verification and validation required for samples accompanying the quotes ? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
V Is it required to perform Process Validation for food or dietary supplements? Food Safety - ISO 22000, HACCP (21 CFR 120) 4
S Is a formal review after OQ (in Process Validation) required? Qualification and Validation (including 21 CFR Part 11) 2
F Determining activities required for Validation of Manual Equipment Qualification and Validation (including 21 CFR Part 11) 4
T Is Testing Equipment Re-Validation required if the Test Equipment is Relocated? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
I ERP Software Validation - required or not? ISO 13485:2016 - Medical Device Quality Management Systems 3
H Is Scope Validation required for ISO 9001? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
W Injection Molding Validation Required or Not - FDA QSR 820.75 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
I Where in the TS 16949 standard is Yearly PPAP Re-Validation Required IATF 16949 - Automotive Quality Systems Standard 9
somashekar Application of parts with PSA (Pressure Sensitive Adhesive) - Is Validation required? ISO 13485:2016 - Medical Device Quality Management Systems 8
W Electronic Distribution of Controlled Documents: Validation required? Qualification and Validation (including 21 CFR Part 11) 5
L Validation and documentation activities required for a custom-made software Qualification and Validation (including 21 CFR Part 11) 2
W Is an Oven I/O/Q Validation Required or is Calibration Sufficient? Qualification and Validation (including 21 CFR Part 11) 7
S Does Package Validation required for Service Parts Packaging Carton Box? ISO 13485:2016 - Medical Device Quality Management Systems 1
A Blasting Visual Standard Validation - Site Acceptance Test required Other Medical Device and Orthopedic Related Topics 1
T What is required for APQP step 4.5 Production Validation Testing APQP and PPAP 2
H Required skills for Automotive R&D Electronic / Electrical design validation Design and Development of Products and Processes 1
P SPC Software Validation Required by our Registrar - How do we do this? Statistical Analysis Tools, Techniques and SPC 6
C 8.3.4 Verification and Validation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
W Transport validation Qualification and Validation (including 21 CFR Part 11) 12
mustomutlu Process Validation Final Report Other Medical Device and Orthopedic Related Topics 2
P Software verification and validation procedure IEC 62304 - Medical Device Software Life Cycle Processes 6
Y SaMD Verification and Validation SOP and Plan IEC 62304 - Medical Device Software Life Cycle Processes 6
C Spreadsheet Validation and Release Qualification and Validation (including 21 CFR Part 11) 2
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
M Sterile packing validation tests to be performed and protocol Other Medical Device and Orthopedic Related Topics 1
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 0
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
M ERP / QMS related software standards for Validation IEC 62304 - Medical Device Software Life Cycle Processes 6
C ISO 19227 Validation Cost Other Medical Device Related Standards 2
A CNC Mill and Lathe - Validation Manufacturing and Related Processes 2
N Example for design and development planning,input,output,review,verification,validation and transfer Misc. Quality Assurance and Business Systems Related Topics 4
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
B Oracle Cloud ERP Validation during Quarterly Patch ISO 13485:2016 - Medical Device Quality Management Systems 1
D Software validation team Misc. Quality Assurance and Business Systems Related Topics 3
W LTPD, AQL, Ppk and Cpk validation sampling plan table Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
J Validation Sample Size for Tray Seal Qualification and Validation (including 21 CFR Part 11) 3

Similar threads

Top Bottom