ISO 10993-7:2008/ Amd.1:2019 - Classification of special populations
akyt
Hi everyone! Can you help me?
I have some doubts related to the implementation of Amendment 1 of the ISO 10993-7:2008 standard.
The amendment talks about new EO and ECH residual limits for products for special populations, such as premature neonates, neonates and children.
For example: for products with limited exposure (less than 24 hours), the limit for EO is 0.3 mg/Kg/day, and the limit for ECH is 0.64 mg/Kg/day.
We have in our portfolio medical devices classified as neonates, pediatric and infants.
So, how can I define a median body mass to establish the new limits of EO and ECH?
The amendment also establishes that the body mass used for each category of special patient population shall be justified and documented.
Another doubt: this amendment was published on 2019-12. What is the maximum period to adapt the company after this publication?
Thank you all.
planB
akyt,
2 options for body-mass references for your patient population:
  • ISO 10993-17:2002, section 6.2.2, Devices specifically intended for use in neonates and children - quote:
TE calculations should be performed using body mass 3,5 kg for neonates and 10 kg for children as the human body mass for that device.
  • Statistical data, either from national institutions or from the WHO
Unless there is a published transition period (which is not for this amendment), there is no fixed time frame, until when regulators expect compliance. Typically, this may be 1/2 - 1 year after publication. For this amendment, at least in the EU, the contents of this amendment have practically already been expected by notified bodies since 2016, after the French authorities raised serious concerns about this standard in 2014.
HTH,
Gerhard
I'm having difficulty parsing what text replaces what with this amendment.
Just focusing on section 4.3.2 for example.
The amendment reads,"4.3.2, last paragraph Replace the paragraph with the following: When the device is intended to be used in special populations, the appropriate patient body mass shall be used for the derivation of the allowable limits. For example, if the device is intended to be used in premature neonates, neonates or children, the allowable limits shall be derived using the TI value of 0,02 mg/kg/day for EO and 0,029 mg/kg/day for ECH as established in G.6.4 and H.4.1.3, respectively. The appropriate default body mass used for each category of special patient population shall be justified and documented."
Does that replace the entire section on ECH limits including the bulleted 24h and 30d limits? If so, it seems to replace what was a fairly parallel structure between EO and ECH limits with something very much not parallel. It seems to replace those limits for ECH with the 0.029mg/kg/day limit, and yet not do the same for EO.
What actually is the last paragraph? What am I missing here? Is there someplace I can just get the amended standard?
Thanks in advance,
TMM
planB
TMM,
Consolidated section 4.3.2 would read like this (amendments in red):
4.3.2 Permanent contact devices
In the case of a device used in an adult of body mass mb = 70 kg, and with CEF = 0,2 and PEF = 1,0 (default factors),
the average daily dose of EO to patient shall not exceed 0,1 mg/d. In addition, the maximum EO dose shall not exceed:
⎯ 4 mg in the first 24 h;
⎯ 60 mg in the first 30 d;
⎯ 2.5 g in a lifetime.
the average daily dose of ECH to patient shall not exceed 0.4 mg/d. In addition, the maximum ECH dose shall not exceed:
⎯ 9 mg in the first 24 h;
⎯ 10 g in a lifetime.
To my knowledge, there is unfortunately no consolidated edition of ISO 10993-7 including this latest amendment.
HTH,
I EO (Ethylene Oxide) Residuals - EN ISO 10993-7:2008 EU Medical Device Regulations 5
M EO (Ethylene Oxide) and ECH (Ethylene Chlorohydrin) Residual as per ISO 10993-7:2008 ISO 13485:2016 - Medical Device Quality Management Systems 15
B Implementation Schedule in EU for ISO 10993-7:2008 - Transition period? EU Medical Device Regulations 9
S Knee Implant (Femoral -Cobalt chrome)-Sub chronic toxicity test (ISO 10993-11)choice of root Medical Device and FDA Regulations and Standards News 2
R ISO 10993-18 Medical Device and FDA Regulations and Standards News 1
M ISO 10993-17 - Ancillary Medicinal substance leachables Medical Device and FDA Regulations and Standards News 1
M Medical Devices - Ancillary - ISO 10993-17 Medical Device and FDA Regulations and Standards News 6
M FDA News FDA Releases Draft Guidance Clarifying Application of ISO 10993-1 Biocompatibility Standard Medical Device and FDA Regulations and Standards News 0
L Biological Assessment (ISO 10993-1) Other Medical Device Related Standards 1
L Process changes and biocompatibility (ISO 10993-1) Other Medical Device Related Standards 1
S ECG Cable Banana to Snap or Tab electrode Adapters -- ISO 10993 Requirement Other Medical Device Related Standards 0
K Diagnostic X-ray devices - Applicability of Biocompatibility Testing per ISO 10993-1 Manufacturing and Related Processes 7
P European versions of standards - ISO 10993-1 and ISO 11607-1 EU Medical Device Regulations 6
S ISO 10993 Biocompatibility Evaluation - Electronic thermometer Other Medical Device Related Standards 3
P ISO 10993-1:2018 - When will the European equivalent become a Harmonized Standard? Other Medical Device Related Standards 13
I ISO 10993 Biocompatibility Requirements - manufacturing process chemicals Other Medical Device Related Standards 8
N Technical File Reviewer has requested more testing to ISO 10993 Other Medical Device Related Standards 10
T ISO 10993-7 - Ethylene oxide sterilization residues Other Medical Device Related Standards 1
planB ISO 10993-1:2018 has just been published Other Medical Device Related Standards 2
M Does ISO 10993 Apply to Packaging? (e.g. Travel Case) Other Medical Device Related Standards 4
K ISO 10993 Requirements For 510k - Substantial Equivalence Other US Medical Device Regulations 11
T Using Risk Management in ISO 10993 - Medical Device Accessory 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
P ISO 10993 - Sterile (Irradiation) Polypropylene Syringe Question Other Medical Device Related Standards 6
D Actual Definition of Non-Contact for ISO 10993 Other Medical Device Related Standards 1
R Biocompatibility - ISO 10993 and SVHC Requirements Other Medical Device Related Standards 4
S ISO 10993-1: External Communicating Device biocompatibility test requirements Other ISO and International Standards and European Regulations 4
S Selecting materials for implants to comply with ISO 10993 biocompatibility Other Medical Device Related Standards 4
T Plastic of an infusion pump enclosure and ISO 10993 ISO 13485:2016 - Medical Device Quality Management Systems 2
E Upcoming changes for ISO 10993 and IEC 60601 in South Korea Other Medical Device Regulations World-Wide 2
M FDA issues draft guidance to replace G95-1 (Application of ISO 10993) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
B Verify GB/T 16886 is the same as ISO 10993 US Food and Drug Administration (FDA) 5
B New ISO 10993-12 - Consistency of Extraction Conditions through different Assays Other ISO and International Standards and European Regulations 1
S Suppliers of ISO 10993 compliant Material for Skin Contact Other ISO and International Standards and European Regulations 3
M ISO 10993 - Repeat tests necessary if colour changes slightly? Other Medical Device Related Standards 11
M Pre-Selecting Materials for ISO 10993-1 Biocompatibility Compliance Other ISO and International Standards and European Regulations 5
F ISO 10993-10 - Removed from List of Harmonized Standards Other Medical Device Related Standards 4
S ISO 10993 Cytotox Testing - Direct Contact rather than Extractables Other Medical Device Related Standards 4
E Meeting ISO 10993-1 2009 Material Risk Assessment Requirements Other Medical Device Related Standards 13
M ISO 10993 Labs which can test Cytotoxicity, Sensitizaton and Irritation Other Medical Device Related Standards 10
M ISO 10993 for Fabric based Electrode - Finding already FDA Approved Materials Other Medical Device Related Standards 11
J Biological Evaluation requirements per ISO 10993 - Medical Device Testing ISO 13485:2016 - Medical Device Quality Management Systems 2
A Claiming Biocompatibility without ISO 10993-1 Other Medical Device Related Standards 19
J ISO 10993 (Biological Compatibility) Approved Materials US Food and Drug Administration (FDA) 23
T ISO 10993-10: 2002 vs 2010 version - What are the Major Changes? ISO 13485:2016 - Medical Device Quality Management Systems 2
W ISO 10993 (Biological Evaluation of Medical Devices) for Suppliers Other ISO and International Standards and European Regulations 8
D Does anyone know of a biocompatible ABS tested & certified to ISO 10993-1 Other Medical Device Related Standards 5
W ISO 10993 - Replacing Sensitization Test with Systemic Toxicity Test? Other Medical Device Related Standards 4
I ISO 10993-10:2010 Guidance Document for Irritation and Sensitization Test - Changes ISO 13485:2016 - Medical Device Quality Management Systems 3
I ISO 10993-1:2009 Sensitization (Initial Evaluation Test) vs. USP Other Medical Device Related Standards 3
B What ISO 10993 considers Non-Invasive vs. Invasive MDD 93/42/EEC EU Medical Device Regulations 3